Clause 7.1.5 Monitoring and measuring resources

eule del ayre

Involved In Discussions
#1
Good Evening from the Philippines.

Hoping anyone will help me on this problem.

Can anyone help me answering to my NCCAR given by our Certifying Body.

Regarding on 7.1.5 Monitoring and Measuring Resources.

This is the statement given by our Certifying Body
Control of Monitoring and Measuring Equipment Procedure (LPC P-24) was established but procedure only talks about measuring equipment. The only activity identified to maintain measuring equipment is verification. It is not clearly evident if difference of Monitoring Equipment vs. Measuring equipment is clear to the organization. This is also true for Verification vs. Calibration

LPC-P-24 is our Procedure, yet I am new to the process and this was the Procedure when I was transferred to Calibration.
The Procedure meant only to Measuring and did not discuss any Monitoring in it.
It only tackle verification only and no Calibration in it.

Can you guys help me answering the NCCAR I'm having a trouble answering it by myself.

Please and thank you, any root cause will help, I will make any 5 whys to it after.
 
Elsmar Forum Sponsor

eule del ayre

Involved In Discussions
#3
Your request is the same as saying "I'm hungry" and expecting satisfaction from someone else eating a meal for you.

Sorry for being greedy but yes, that's the same thing for the NCCAR should be send within 7 working days.
what I'm trying to say is, IS THERE ANY EXCUSES THAT THOSE LISTED ABOVE WAS NOT ON THE PROCEDURE?
I mean I can't think of any, because I was just transferred to Calibration, and yet my supervisor is giving me the responsibility to answer the NCCAR
 

Mike S.

Happy to be Alive
Trusted Information Resource
#4
All I can do is speculate, and if the person who wrote the procedure is unavailable for you to ask, maybe all you can do is speculate as well. Is that person unavailable (i.e. no longer with the company)?

If that is the case, I would state in my CAR response that since the original author of the procedure is unavailable, all I can do is speculate on the root cause, and that the most likely cause is that the author did not thoroughly understand the requirement of the standard.
 

eule del ayre

Involved In Discussions
#5
If that is the case, I would state in my CAR response that since the original author of the procedure is unavailable, all I can do is speculate on the root cause, and that the most likely cause is that the author did not thoroughly understand the requirement of the standard.

Well this is one of my so many root causes sir, but hear me out, before the certification audit (surveillance audit) we had the so-called internal audit right. Before the internal audit I was transferred to calibration, which little did I know the procedure was lacking. But didn't really study well the procedure because the one handling the calibration got no endorsement.

But if I will do that, didn't the internal audit was lacking? not my procedure? that's why I needed you guys to give me insights.

Before Calibration, I handle Chemicals under QA (MSDS, GHS, PCO stuffs)

But I really appreciate your answer, the problem is if I do that, wouldn't that reflect first to Internal audit, that's what I'm thinking sir.
 

Mike S.

Happy to be Alive
Trusted Information Resource
#6
Personally, I think you're over-thinking this, making it harder than it needs to be.

Just because an external auditor finds a NC doesn't mean the internal audit process failed. All auditors are just sampling, not auditing everything. And internal auditors generally audit to the company's procedures and documents, not to the standard itself.

Unless this was the very first audit by the "certifying body" didn't they, too, fail to detect this as a NC in the past?

If it were me, I would just move on. I'd say since the original author of the procedure is unavailable, all I can do is speculate on the root cause, and that the most likely cause is that the author did not thoroughly understand the requirement of the standard. The corrective action is that you are the new person in charge of the area, you now understand the requirements, and you have done x, y, and z to fix the problem.
 

eule del ayre

Involved In Discussions
#7
If it were me, I would just move on. I'd say since the original author of the procedure is unavailable, all I can do is speculate on the root cause, and that the most likely cause is that the author did not thoroughly understand the requirement of the standard. The corrective action is that you are the new person in charge of the area, you now understand the requirements, and you have done x, y, and z to fix the problem.[/QUOTE]

- Nicely done sir, you're right I am over-thinking.

Unless this was the very first audit by the "certifying body" didn't they, too, fail to detect this as a NC in the past?
- Yes, after my supervisor said in the meeting this morning (Philippine time) they fail to detect this in the past, so they now give us an NC (minor) over it. Since we have the procedure yet focusing on Measuring tools only neglecting the monitoring tools, and No Calibration statement whatsoever on the procedure just verification.
For what I remember in the morning meeting, my supervisor said, "The Auditor are leveling up and need thorough auditing, reflecting Standards to our Procedures"

But thank you sir for helping me enlighten and understand what I will gonna do to this NCCAR.

Well Appreciated.
 

BradM

Leader
Admin
#8
Hello there!!

I'm confused. Are you saying you were issued a CAR because your procedure didn't have the word "calibration" in it?
Your instrumentation... are you performing... checks/ calibrations/ verifications with traceable standards?

What exactly is the deviation?
 

eule del ayre

Involved In Discussions
#9
I'm confused. Are you saying you were issued a CAR because your procedure didn't have the word "calibration" in it?
Your instrumentation... are you performing... checks/ calibrations/ verifications with traceable standards?

What exactly is the deviation?
Good Morning From Philippines sir,

Sorry about the confusion sir, well the procedure indicates just verification only, but what the supervisor told me to do was to initiate the calibration.
On the day of Third Party Audit, when they review the procedure, the procedure indicates verification only, and tried questioning me, "So you're in-charge in the calibration, let me see your procedure and understand it".
What we really do, technically was calibration (external service provider) for small tool and measuring equipment, and they started to question me about, if I understand the word Calibration vs. Verification.
To make it short Sir, we perform calibration (external service provider) but on the procedure it says verification only, we didn't include external service provider.

Another Problem is that the our procedure LPC-P-24 called Control of Monitoring and Measuring Equipment Procedure, does not include the Monitoring Equipment.
 
Last edited by a moderator:

BradM

Leader
Admin
#10
I fixed your quote in the last post; it makes is so much easier to read. :)

Good Morning From Philippines sir,

Sorry about the confusion sir, well the procedure indicates just verification only, but what the supervisor told me to do was to initiate the calibration.
On the day of Third Party Audit, when they review the procedure, the procedure indicates verification only, and tried questioning me, "So you're in-charge in the calibration, let me see your procedure and understand it".
What we really do, technically was calibration (external service provider) for small tool and measuring equipment, and they started to question me about, if I understand the word Calibration vs. Verification.
To make it short Sir, we perform calibration (external service provider) but on the procedure it says verification only, we didn't include external service provider.

Another Problem is that the our procedure LPC-P-24 called Control of Monitoring and Measuring Equipment Procedure, does not include the Monitoring Equipment.
Ok... for the moment.... set aside your procedure and the CAR.

You have a facility with instruments that are used that need to have their accuracy verified. So someone uses a traceable standard, makes checks on them against some procedure/ work instruction, and documents that somehow.

Regardless of this main procedure or the CAR, is that currently happening?
 
Thread starter Similar threads Forum Replies Date
J Compliance with ISO 9001 Clause 7.6 Control of Monitoring & Measuring Devices ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
R ISO 13485 Clause 7.6 Control of Monitoring and measuring devices - Requirements ISO 13485:2016 - Medical Device Quality Management Systems 10
W Computer Software - ISO 9001 Clause 7.6 Control of Monitoring and Measuring equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
L How to comply with AS9100 Clause 7.6 - Monitoring and measuring devices to be used AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 31
L Control of Monitoring and Measuring Devices - Clause 7.6 Exclusion? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
J How to accomplish ISO 13485 Clause 8.2.3 Monitoring and Measuring? ISO 13485:2016 - Medical Device Quality Management Systems 5
3 Calibration Frequency - Control of Monitoring and Measuring Devices - Clause 7.6 Calibration Frequency (Interval) 14
R TS 16949 Clause 7.6 Control of Monitoring and Measuring Devices IATF 16949 - Automotive Quality Systems Standard 12
M Calibrate Computers - ISO 9001 Clause 7.6. Control of Monitoring & Measuring Devices ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
qualprod ISO 9001 Clause 9.1 - Monitoring measurement analysis and evaluation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
A Records required for ISO 9001 clause 7.1.5.1 - Fitness for purpose of the monitoring and measurement resources ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
B IATF 16949 Section Clause 8.3.4.1 - Monitoring - Design and Development Input(s) IATF 16949 - Automotive Quality Systems Standard 3
M Quality Objectives and Monitoring & Measurement of Processes - Clause 8.2.3 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D AS9100C Clause 8.2.4 Monitoring and Measurement of Product AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
S ISO 9001:2008 Clause 8.2.3 Monitoring and Measurement of Processes Help ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
H ISO 9001:2008 Clause 7.6 Control of Monitoring and Measurement (Computer Software) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
W AS9100 Clause 8.2.3.C Monitoring and Measurement of Processes AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
S Process Monitoring and Measurement - AS9100 Clause 8.2.3 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
E ISO 9001:2008 Clause 8.2.4 - Monitoring and Measurement of Product ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
M TS 16949 Clause 8.2.4 Monitoring and Measurement of Product - Customer Scope? IATF 16949 - Automotive Quality Systems Standard 3
G Monitoring and Measurement of Manufacturing Processes - TS 16949 Clause 8.2.3.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100 Clause 8.2.3.C Monitoring and Measurement and 8.5.2 Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
P Control of Monitoring and Measurement Devices - ISO 13485 Clause 7.6 says ... ISO 13485:2016 - Medical Device Quality Management Systems 3
C ISO 14971 Clause 9 Requirements - Post-Production Monitoring and Risk Management ISO 14971 - Medical Device Risk Management 7
R Interpretation of TS 16949 Clause 7.4.3.2 Supplier Monitoring IATF 16949 - Automotive Quality Systems Standard 7
R Control of Monitoring & Measurement Equipment - ISO 9001:2008, Clause 7.6 c) General Measurement Device and Calibration Topics 9
Q Suppliers Monitoring their Processes - How to comply with Clause 7.4.3.2 IATF 16949 - Automotive Quality Systems Standard 6
A How to carry out 'Monitoring and measurement of processes' - Clause 8.2.3. Nonconformance and Corrective Action 12
M Supplier Monitoring - Do we need to do Supplier Audits? - TS16949 Clause 7.4.3.2 IATF 16949 - Automotive Quality Systems Standard 7
A TS 16949 Clause 7.4.3.2 Promotion of supplier monitoring - What is meant by promote? Supplier Quality Assurance and other Supplier Issues 6
A Supplier Monitoring - Meeting the requirements of TS 16949 Clause 7.4.3.2 Supplier Quality Assurance and other Supplier Issues 6
S TS 16949 Clause 7.4.3.2 - Supplier Monitoring - How to comply Supplier Quality Assurance and other Supplier Issues 4
Sidney Vianna Monitoring and measurement of processes - How do you comply? ISO 9001 Clause 8.2.3 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 30
E What are monitoring devices? ISO 9001 Clause 7.6 - Control of ... monitoring devices ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
G Clause 8.5.1 -- Is non Identification evidence of ineffective control of production ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 20
G Clause 7.5.3 and Required Work Instructions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
LincolnQA AS9100D, Clause 8.4.1.1d - actions to take when external providers do not meet requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
H IEC 61010-1:2010/A1:2016 - Clause 10.5.3 - (b) - Interpretation CE Marking (Conformité Européene) / CB Scheme 6
E Clause 15.4.2.1 d) - Loss of ESSENTIAL PERFORMANCE IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Clause 7.5.2 of ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
S Is design and development clause 8.3 applicable for laboratory ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Clause for use of expired quality alerts AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
H DIN 12880:2007-05 Clause 7.2 Test Equipment Description CE Marking (Conformité Européene) / CB Scheme 0
D API Q1 clause (1.0) Scope Oil and Gas Industry Standards and Regulations 1
D API Q1 clause (5.6.3) Verification of Purchased Products or Activities ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
R 9001 and 9004 clause misalignment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
L 60601-2-10 clause 201.12.4.103 Output Indicator (itself in SFC?) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B IATF 16949 clause 10.2 - requirements following a customer complaint. IATF 16949 - Automotive Quality Systems Standard 9
R Clause 8.5.1.2 - Y1 Capacitors in Series IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
P AS9100D clause 8.6 - Documentation required to show evidence of conformity AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5

Similar threads

Top Bottom