Clause 7.6 - "Shall have Identification in order to determine its calibration status"

#11
Re: Clause 7.6 - "Shall have Identification in order to determine its calibration sta

Jim, I think you're not seeing the fallacy of 'the next due' date. Real, effective calibration is based on use, not dates!

I was burned (in the distant past) where a cal. system was run on a time based recall. Trouble was, the gauge usage sky rocketed (how would the gauge lab find out?) and the first 3 turns of a bunch of male thread gauges wore badly (used in stainless steel). Those cal systems I've seen, rarely rely on cal. data (as found, as left) or similar. They are based on some arbitrary, wet finger in the air guestimate (the gauge supplier's answer?) - we read about them here all the time!

An effective cal system takes that data and trends it, keeping in mind use, to determine recall, while still within spec. Real world...Trouble is, most people don't know/want to get into that (it's resource consuming) and external suppliers of cal services don't go that far with their services...

If I was running a shop, I'd be all over an operator being paid to wander around trying to find out if a $50 gauge was calibrated - when I pay someone to know how to calibrate/control it! Double whammy!:mg:
 
Last edited:
Elsmar Forum Sponsor

Big Jim

Super Moderator
#12
Re: Clause 7.6 - "Shall have Identification in order to determine its calibration sta

Jim, I think you're not seeing the fallacy of 'the next due' date. Real, effective calibration is based on use, not dates!

I was burned (in the distant past) where a cal. system was run on a time based recall. Trouble was, the gauge usage sky rocketed (how would the gauge lab find out?) and the first 3 turns of a bunch of male thread gauges wore badly (used in stainless steel). Those cal systems I've seen, rarely rely on cal. data (as found, as left) or similar. They are based on some arbitrary, wet finger in the air guestimate (the gauge supplier's answer?) - we read about them here all the time!

An effective cal system takes that data and trends it, keeping in mind use, to determine recall, while still within spec. Real world...Trouble is, most people don't know/want to get into that (it's resource consuming) and external suppliers of cal services don't go that far with their services...

If I was running a shop, I'd be all over an operator being paid to wander around trying to find out if a $50 gauge was calibrated - when I pay someone to know how to calibrate/control it! Double whammy!:mg:
If the company has a system that is able to determine calibration intervals based on usage, not on time (although I've heard of them in theory, I've never seen one) then my comments may not apply.

But I wonder how it would be tracked. If it is by the user maintaining a log, then the user still would be the first to know when the usage limit is nearing or hit.
 

Jim Wynne

Staff member
Admin
#13
Re: Identification for 7.6

I can see that I'm standing against two giants in the group, but I still feel that it should be the responsibility of the user to check the status before using a tool. I completely agree that it is the primary duty of management to recall them and get them calibrated, but a back stop is needed for those that may fall through the cracks.

Very frequently on audits I find expired tools that management missed. Sometimes they were missed because they could not find it, and it was recorded as missing.

In a perfect world, that would never happen, but we do not live in a perfect world. A backstop is need for those that fall through the cracks. Saying that no one ever looks is a training issue and a cop-out. They should look.
A label that says that the calibration date for a device has passed doesn't mean that the device isn't accurate. Calibration status should have little to do with labels and a lot to do with what's being measured. If a situation occurs where measurements are falling close enough to specification limits that gage error is likely to cause a problem, the device should be verified (e.g., calibrated against a standard) before decisions are made. Likewise, when measuring things with tight tolerances, devices should be verified, regardless of what a sticker might indicate.

I think due-date labels are more likely to create a false sense of security than they are to help anyone to make accurate measurements.
 
#14
Re: Identification for 7.6

A label that says that the calibration date for a device has passed doesn't mean that the device isn't accurate. Calibration status should have little to do with labels and a lot to do with what's being measured. If a situation occurs where measurements are falling close enough to specification limits that gage error is likely to cause a problem, the device should be verified (e.g., calibrated against a standard) before decisions are made. Likewise, when measuring things with tight tolerances, devices should be verified, regardless of what a sticker might indicate.

I think due-date labels are more likely to create a false sense of security than they are to help anyone to make accurate measurements.
Excellent point, Jim W! I had had the same thought, but forgot to mention the point about accuracy after the specified date, in my last post.

It's always been a bone of contention for me that auditors would write up an nc; "Gauges X, Y and Z were not recalled after the date". They can be perfectly accurate the day after, and possibly for months later, under normal usage. But the auditor didn't probe into the calibration data/trends to see what the risk is! So, a wild goose chase ensues, equipment purged only to discover the 'as found' conditions indicate it is still accurate for use! Value of audit finding? 20% of what it SHOULD be...

Big, Jim - it's a sad indictment of calibration system that such a comment is true - rarely are recalls etc based on use, so they tend to be the default of a very basic calendar which is fairly arbitrarily selected, what's more.
 

Big Jim

Super Moderator
#15
Re: Identification for 7.6

Excellent point, Jim W! I had had the same thought, but forgot to mention the point about accuracy after the specified date, in my last post.

It's always been a bone of contention for me that auditors would write up an nc; "Gauges X, Y and Z were not recalled after the date". They can be perfectly accurate the day after, and possibly for months later, under normal usage. But the auditor didn't probe into the calibration data/trends to see what the risk is! So, a wild goose chase ensues, equipment purged only to discover the 'as found' conditions indicate it is still accurate for use! Value of audit finding? 20% of what it SHOULD be...

Big, Jim - it's a sad indictment of calibration system that such a comment is true - rarely are recalls etc based on use, so they tend to be the default of a very basic calendar which is fairly arbitrarily selected, what's more.
Both of you are forgetting the requirements of the standard. Don't ding your auditor for doing what he should.

7.6 a speaks specifically to what you both are ignoring.

"Where necessary to ensure valid results, measuring equipment shall . . . be calibrated or verified, or both, AT SPECIFIED INTERVALS . . . "

Emphasis added.

If the specified interval is time with a defined due date, IT DOESN'T MATTER IF THE TOOL IS STILL ACCURATE. If it is still accurate then you can avoid grief over the paragraph after the letters in 7.6.

"In addition, the organization shall assess and record the validity of the previous measuring results when the equipment is found not to conform to requirements. The organization shall take appropriate action on the equipment and any product affected."

Maybe you guys are using a different standard.
 
#16
Re: Clause 7.6 - "Shall have Identification in order to determine its calibration sta

No Jim, not ignoring, not using a different standard - just looking at it from another perspective, not the conventional 'wisdom'...:notme:
 

Jim Wynne

Staff member
Admin
#18
Re: Identification for 7.6

Both of you are forgetting the requirements of the standard. Don't ding your auditor for doing what he should.

7.6 a speaks specifically to what you both are ignoring.

"Where necessary to ensure valid results, measuring equipment shall . . . be calibrated or verified, or both, AT SPECIFIED INTERVALS . . . "

Emphasis added.

If the specified interval is time with a defined due date, IT DOESN'T MATTER IF THE TOOL IS STILL ACCURATE. If it is still accurate then you can avoid grief over the paragraph after the letters in 7.6. I don't know who "both of you" refers to; I didn't say anything about specified intervals not being required.

"In addition, the organization shall assess and record the validity of the previous measuring results when the equipment is found not to conform to requirements. The organization shall take appropriate action on the equipment and any product affected."

Maybe you guys are using a different standard.
I don't know who "both of you" and "you guys" refers to. I haven't ignored anything and haven't said anything about "specified intervals" not being required. It has nothing to do with the subject at hand, which is whether labels are required. They're not required, and the argument can and has been made that labels might not even be useful or prudent. If one does find them useful, they should be used, but I think they're used mainly because people think they have to use them.
 
Thread starter Similar threads Forum Replies Date
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 11
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
J AS9104-1 Clause 18.1 (a) (b) - Organizations (contractor) shall be required to allow access to Tier 2 data AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
B IATF 16949 clause 7.1.5.1.1 - Statistical studies shall be conducted IATF 16949 - Automotive Quality Systems Standard 3
M AS9100 Clause 4.2.3 requirements - Documents required by the QMS shall be Controlled AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 13
T Quality Manual shall Outline Structure of Documentation - Clause 4.2.2 ISO 13485:2016 - Medical Device Quality Management Systems 10
P Specified stages of design and development shall be defined - TS 16949 clause 7.3.4.1 IATF 16949 - Automotive Quality Systems Standard 4
R TS16949 Clause 7.6.3.1 - Internal laboratory - Facility shall have a defined scope IATF 16949 - Automotive Quality Systems Standard 5
K OHSAS 18001 - Clause 4.4.3 Communication "...shall be documented..." Occupational Health & Safety Management Standards 5
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
R AS5553 Clause 3.1.7 f - "Implement a returns process....." AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
V Setup for testing against ISO14708 clause 16 (protection of the patient from herms caused by heat) Other Medical Device Related Standards 0
M IATF16949 Clause 9.1.2.1e - Customer notification related IATF 16949 - Automotive Quality Systems Standard 4
N 60601-2-2 Ed 6 Clause 201.8.8.3.102 - HF instrument leakage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
A AS9100D - Clause 8.1 Operation - Coating service company AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
T AS9100D Clause 10.2.1g Supplier Corrective Action for each and every nonconformity? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
H AS9100D clause 8.5.1 f) & 8.5.1 g) - Special Processes AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S ISO 9001 Clause 8.2.3 - Review of the requirements for products and services in a Cafe ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
S API Spec. Q1 clause 5.6.1.2 On site evaluation Oil and Gas Industry Standards and Regulations 10
lanley liao How to understand the clause 6 Planning of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
lanley liao How to understand the term 3.1.21 'Servicing' and the clause 5.7.1.2 'Servicing' of API Spec Q1? Oil and Gas Industry Standards and Regulations 19
earl62 IATF 16949 Clause 9.1.1.1 - What is the batch conformance to specification method? IATF 16949 - Automotive Quality Systems Standard 3
D Layered Process Audits - FCA 9.2.2 - Exemption Clause? IATF 16949 - Automotive Quality Systems Standard 5
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J Scope of ISO 9001 clause 10.2 in the product life cycle ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S MDR GSPR Clause 17 - Software Requirements EU Medical Device Regulations 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
R Clause 7.7 Replicate, Recalibration and Intermediate checks using Artifact ISO 17025 related Discussions 1
A ISO 13485 - Clause 8.3.3 appropriate actions in response to nonconforming product ISO 13485:2016 - Medical Device Quality Management Systems 2
M Description of the requirements of clause 9.2.2.3 manufacturing process audit- needs your feedback IATF 16949 - Automotive Quality Systems Standard 0
A ISO 41001:2018 - Clause No.8 Operations Part Quality Management System (QMS) Manuals 2
L AS5553 Clause 3.1.7 e "Control packaging material ..................reused" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
R What is meant by Use of COMPONENTS WITH HIGH-INTEGRITY CHARACTERISTICS in ME EQUIPMENT (clause no 4.9) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P Definition of production tooling in IATF16949 clause 8.5.1.6 IATF 16949 - Automotive Quality Systems Standard 4
K ISO 13485:2016, Clause 4.2.3 Medical Device File ISO 13485:2016 - Medical Device Quality Management Systems 4
I Master Document Access - ISO 9001:2015 clause 7.5.3 Document Control Systems, Procedures, Forms and Templates 5
qualprod Why new clause (7.1.6) in ISO 9001:2015? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
F ISO 13485 8.2.3 Reporting to regulatory authorities: Question regarding a procedure for this clause. ISO 13485:2016 - Medical Device Quality Management Systems 4
R Process control documents - Context of API Spec Q1, Clause 5.7.1.3 Other ISO and International Standards and European Regulations 0
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
O How can I justify excluding the R&D group and the design and development clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1

Similar threads

Top Bottom