Clause 7.6 - "Shall have Identification in order to determine its calibration status"

somashekar

Staff member
Super Moderator
#21
Re: Clause 7.6 - "Shall have Identification in order to determine its calibration sta

It could turn out to be both and more .. :notme:
Standard is so general in its saying that any method of identification could work for the organization managing a calibration system. This also has got to do in the maturity and knowledge level of the user of the instrument. With nothing done, the unique Sl. # of the instrument can be the identification of its calibration status, if this Sl. # is captured into the calibration system.
There are situations where some model of instruments do not have Sl. # and you could be using many of them. There could be situations where the instrument is so mounted that it is not easy to access and see the Sl. # of the instruments perhaps on its rear panel. These are situations when the organization plan to identify by some numbering method (label / sticker) and across the board apply to all instruments. We had an other thread about POST-IT note being used to mark identification.
So coming back to the OP's point, he is OK in using his expensive s/w package and its issued label for identification.
 
Last edited:
Elsmar Forum Sponsor

howste

Thaumaturge
Super Moderator
#22
Re: Identification for 7.6

It's always been a bone of contention for me that auditors would write up an nc; "Gauges X, Y and Z were not recalled after the date". They can be perfectly accurate the day after, and possibly for months later, under normal usage. But the auditor didn't probe into the calibration data/trends to see what the risk is! So, a wild goose chase ensues, equipment purged only to discover the 'as found' conditions indicate it is still accurate for use! Value of audit finding? 20% of what it SHOULD be...
I'm going to have to disagree with you on this one. If the organization specified the calibration/verification intervals based on risk then there's a good reason to write a nonconformity if the gages aren't recalled by their due date. It's exceeded the organization's own risk criteria. The organization determined that to go beyond that date was a risk they were unwilling to accept.

Now if the organization specified the intervals based on nonsensical reasons, then it's not the auditor's fault that there's no value in the finding. You suggest that the auditor's job is to identify the risks to the organization. I say that it's the organization's responsibility to identify their risks. If there's no value, the responsibility rests squarely on the organization.
 

somashekar

Staff member
Super Moderator
#23
Re: Identification for 7.6

But the auditor didn't probe into the calibration data/trends to see what the risk is! So, a wild goose chase ensues, equipment purged only to discover the 'as found' conditions indicate it is still accurate for use! Value of audit finding? 20% of what it SHOULD be...
Perhaps audit purpose gets diverted to analysis, investigation and advisory...leading to consultancy.
 

harry

Super Moderator
#24
Re: Identification for 7.6

............... The question you should want to find out is, what's best for the employees and your system? I personally prefer to have something to reference on the equipment itself, such as a label, so that I know it's in cycle before I use it. Not everyone has that preference though.
I thought howste sums it up very nicely in this post with regards to 'label' - it is individual preference. Not having one doesn't improve the situation of having one.

What should had been mentioned (and for the benefit of those who are new) is that the calibration process had to be actively managed just like any other processes. You start from the risk angle. The people who calibrate your equipments often do not know how you use it. They just affix a date of calibration and a date for next calibration - commonly and arbitrarily fixed as 1 year. The person who managed the calibration program should assess whether this 1 year period is acceptable from the risk angle. If not, they should fix a date/period and monitor it by some means. At other times, you may find that you can extend this 1 year period - for which you will need all the justification and then document monitor and recall the equipment when the date is due.

The same applies to situations where you choose not to affix a sticker - the argument of whether an affixed sticker causes confusion is ongoing. My :2cents: is that in an industrially less matured country like my, having a sticker is more common.
 

Big Jim

Super Moderator
#25
Re: Identification for 7.6

I don't know who "both of you" and "you guys" refers to. I haven't ignored anything and haven't said anything about "specified intervals" not being required. It has nothing to do with the subject at hand, which is whether labels are required. They're not required, and the argument can and has been made that labels might not even be useful or prudent. If one does find them useful, they should be used, but I think they're used mainly because people think they have to use them.
I apologize Jim. I made a bad assumption that the two of you were standing together.

I agree that there are other ways to meet the requirement for 7.6c other than using labels. It is up to the organization involved to determine what is effective for them.

I like the idea of using stickers or tags, and if it isn't practical to apply a sticker or tag on the instrument itself, to apply it to the storage container and keep the container in close proximity. I like it, because it can be an easy way to provide a back stop for watching for expiration. I also believe it makes it easier for management to keep track of them too.
 

Big Jim

Super Moderator
#26
Re: Clause 7.6 - "Shall have Identification in order to determine its calibration sta

So by implication, then, you have the only correct interpretation and the position Jim W and I take is, therefore, incorrect? Is that what you mean, Big Jim?:popcorn:
I don't know what it means for you. Only you can tell. Maybe it means you have forgotten to read the requirement. No interpretation needed. It is very clear.
 

Big Jim

Super Moderator
#27
Re: Clause 7.6 - "Shall have Identification in order to determine its calibration sta

Jim, I think you're not seeing the fallacy of 'the next due' date. Real, effective calibration is based on use, not dates!

I was burned (in the distant past) where a cal. system was run on a time based recall. Trouble was, the gauge usage sky rocketed (how would the gauge lab find out?) and the first 3 turns of a bunch of male thread gauges wore badly (used in stainless steel). Those cal systems I've seen, rarely rely on cal. data (as found, as left) or similar. They are based on some arbitrary, wet finger in the air guestimate (the gauge supplier's answer?) - we read about them here all the time!

An effective cal system takes that data and trends it, keeping in mind use, to determine recall, while still within spec. Real world...Trouble is, most people don't know/want to get into that (it's resource consuming) and external suppliers of cal services don't go that far with their services...

If I was running a shop, I'd be all over an operator being paid to wander around trying to find out if a $50 gauge was calibrated - when I pay someone to know how to calibrate/control it! Double whammy!:mg:
Where's the "double whammy?

How is a user "running around trying to find out if a $50 gauge was calibrated" if all he needs to do is look at the sticker when he uses the tool.

If you wanted the user to be the back stop and there were no stickers, then, yes, he could be wandering around, but that was not my suggestion was it?
 
M

Mr.Happy

#28
Re: Identification for 7.6

Very frequently on audits I find expired tools that management missed. Sometimes they were missed because they could not find it, and it was recorded as missing.
How do you explain that you can find "missed equipment" when they are with a sticker? That means that the users do not look at all to the expired date on it!
Me myself choose to not sticker any measuring equipment except maybe those that are calibrated on location by a external lab e.a. CMM.
Try to sticker an M2 Gage :confused:
Our calibration due times are central controlled by our self made Excel file and except from the Excel program from MS it is far from expensive :lmao:
Cheers
Mr.Happy
 
Last edited by a moderator:

Big Jim

Super Moderator
#29
Re: Identification for 7.6

How do you explain that you can find "missed equipment" when they are with a sticker? That means that the users do not look at all to the expired date on it!
Me myself choose to not sticker any measuring equipment except maybe those that are calibrated on location by a external lab e.a. CMM.
Try to sticker an M2 Gage :confused:
Our calibration due times are central controlled by our self made Excel file and except from the Excel program from MS it is far from expensive :lmao:
Cheers
Mr.Happy
Management could not find it and users provided no insight as to where it was so management figured that the tool had grown legs. As an auditor, I found it in use with an expired sticker. The records also showed that the calibration update had not been performed.

It's not up the the auditor to determine why it was missed. Auditors write nonconformances based on what they find. It's up to the organization to figure out what happened and apply a cure. The nonconformance response on one of those occasions found that an employee had taken the tool home so it was missing when it was due for recall and didn't tell anyone when it was returned to use.

I'm sure there are other things that happen to cause the same situation, such as the tool being put away where it isn't expected, such as leaving with a bin of tooling for a particular project.

Your system sounds appropriate.
 

BradM

Staff member
Admin
#30
Re: Clause 7.6 - "Shall have Identification in order to determine its calibration sta

Wow! This popped up to be an interesting discussion. :agree1:

I guess for me I have seen and heard the calibration sticker debate, and see merit on both sides and such. I guess overall I find it easier to have the calibration stickers with:

  • ID number
  • Calibration date
  • Due Date
  • Name doing the calibration
  • Use range/limits/ etc.
Now... some questions come to mind about the overall process, including

  • Size of instrument program
  • Accessibility of information in calibration program
  • Overall training of individuals
  • Level of management concern in the instrument program
  • General control of the instruments (are they "hid"; do they end up in drawers; etc.)
It really doesn't matter what system you develop, if people are too lazy or don't care enough to assess the current status, nothing will fix that. Too, if your calibration status database/program takes a while to access, requires special passwords, not basic in its operation, etc., the chances of people checking it are slim.

Overall... I have found that calibration stickers work out best, as it gives an 'in the field' assessment of a piece of equipment. The odds of getting people to "assess" the equipment they are using is a bit slim anyway. :D But tends to be slightly better if there is an indicator on the instrument.
 
Thread starter Similar threads Forum Replies Date
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 11
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
J AS9104-1 Clause 18.1 (a) (b) - Organizations (contractor) shall be required to allow access to Tier 2 data AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
B IATF 16949 clause 7.1.5.1.1 - Statistical studies shall be conducted IATF 16949 - Automotive Quality Systems Standard 3
M AS9100 Clause 4.2.3 requirements - Documents required by the QMS shall be Controlled AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 13
T Quality Manual shall Outline Structure of Documentation - Clause 4.2.2 ISO 13485:2016 - Medical Device Quality Management Systems 10
P Specified stages of design and development shall be defined - TS 16949 clause 7.3.4.1 IATF 16949 - Automotive Quality Systems Standard 4
R TS16949 Clause 7.6.3.1 - Internal laboratory - Facility shall have a defined scope IATF 16949 - Automotive Quality Systems Standard 5
K OHSAS 18001 - Clause 4.4.3 Communication "...shall be documented..." Occupational Health & Safety Management Standards 5
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
R AS5553 Clause 3.1.7 f - "Implement a returns process....." AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
V Setup for testing against ISO14708 clause 16 (protection of the patient from herms caused by heat) Other Medical Device Related Standards 0
M IATF16949 Clause 9.1.2.1e - Customer notification related IATF 16949 - Automotive Quality Systems Standard 4
N 60601-2-2 Ed 6 Clause 201.8.8.3.102 - HF instrument leakage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
A AS9100D - Clause 8.1 Operation - Coating service company AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
T AS9100D Clause 10.2.1g Supplier Corrective Action for each and every nonconformity? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
H AS9100D clause 8.5.1 f) & 8.5.1 g) - Special Processes AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S ISO 9001 Clause 8.2.3 - Review of the requirements for products and services in a Cafe ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
S API Spec. Q1 clause 5.6.1.2 On site evaluation Oil and Gas Industry Standards and Regulations 10
lanley liao How to understand the clause 6 Planning of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
lanley liao How to understand the term 3.1.21 'Servicing' and the clause 5.7.1.2 'Servicing' of API Spec Q1? Oil and Gas Industry Standards and Regulations 19
earl62 IATF 16949 Clause 9.1.1.1 - What is the batch conformance to specification method? IATF 16949 - Automotive Quality Systems Standard 3
D Layered Process Audits - FCA 9.2.2 - Exemption Clause? IATF 16949 - Automotive Quality Systems Standard 5
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J Scope of ISO 9001 clause 10.2 in the product life cycle ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S MDR GSPR Clause 17 - Software Requirements EU Medical Device Regulations 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
R Clause 7.7 Replicate, Recalibration and Intermediate checks using Artifact ISO 17025 related Discussions 1
A ISO 13485 - Clause 8.3.3 appropriate actions in response to nonconforming product ISO 13485:2016 - Medical Device Quality Management Systems 2
M Description of the requirements of clause 9.2.2.3 manufacturing process audit- needs your feedback IATF 16949 - Automotive Quality Systems Standard 0
A ISO 41001:2018 - Clause No.8 Operations Part Quality Management System (QMS) Manuals 2
L AS5553 Clause 3.1.7 e "Control packaging material ..................reused" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
R What is meant by Use of COMPONENTS WITH HIGH-INTEGRITY CHARACTERISTICS in ME EQUIPMENT (clause no 4.9) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P Definition of production tooling in IATF16949 clause 8.5.1.6 IATF 16949 - Automotive Quality Systems Standard 4
K ISO 13485:2016, Clause 4.2.3 Medical Device File ISO 13485:2016 - Medical Device Quality Management Systems 4
I Master Document Access - ISO 9001:2015 clause 7.5.3 Document Control Systems, Procedures, Forms and Templates 5
qualprod Why new clause (7.1.6) in ISO 9001:2015? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
F ISO 13485 8.2.3 Reporting to regulatory authorities: Question regarding a procedure for this clause. ISO 13485:2016 - Medical Device Quality Management Systems 4
R Process control documents - Context of API Spec Q1, Clause 5.7.1.3 Other ISO and International Standards and European Regulations 0
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
O How can I justify excluding the R&D group and the design and development clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1

Similar threads

Top Bottom