Clause 7.6 - "Shall have Identification in order to determine its calibration status"

Jim Wynne

Staff member
Admin
#31
Re: Identification for 7.6

I'm going to have to disagree with you on this one. If the organization specified the calibration/verification intervals based on risk then there's a good reason to write a nonconformity if the gages aren't recalled by their due date. It's exceeded the organization's own risk criteria. The organization determined that to go beyond that date was a risk they were unwilling to accept.

Now if the organization specified the intervals based on nonsensical reasons, then it's not the auditor's fault that there's no value in the finding. You suggest that the auditor's job is to identify the risks to the organization. I say that it's the organization's responsibility to identify their risks. If there's no value, the responsibility rests squarely on the organization.
I agree in principle that there's no way around the "specified intervals" requirement, and that auditors have to call 'em as they see 'em. The problem with the whole thing is that the calibration interval, regardless of accounting for risk, will always be arbitrary at some level. No one knows when or if anything will need to be calibrated or adjusted and when something does, the need almost never has anything to do with chronology. For this reason, too much emphasis is placed on dates and not enough on actual measurement issues, and too many meaningless CAs are done for things being beyond their dates when being beyond the date hasn't been shown to affect anything. This is not the fault of auditors, though--if the organization has broken its own rule, it's a nonconformity that should be documented.

As I suggested earlier, devices should always be verified when measurements are close to spec limits or when spec limits are very tight. If anyone thinks that stickers are helpful, stickers should be used.
 
Elsmar Forum Sponsor

Ajit Basrur

Staff member
Admin
#32
Re: Identification for 7.6

I agree in principle that there's no way around the "specified intervals" requirement, and that auditors have to call 'em as they see 'em. The problem with the whole thing is that the calibration interval, regardless of accounting for risk, will always be arbitrary at some level. No one knows when or if anything will need to be calibrated or adjusted and when something does, the need almost never has anything to do with chronology. For this reason, too much emphasis is placed on dates and not enough on actual measurement issues, and too many meaningless CAs are done for things being beyond their dates when being beyond the date hasn't been shown to affect anything. This is not the fault of auditors, though--if the organization has broken its own rule, it's a nonconformity that should be documented.

As I suggested earlier, devices should always be verified when measurements are close to spec limits or when spec limits are very tight. If anyone thinks that stickers are helpful, stickers should be used.
Grat points, Jim :agree1:
 
#33
Re: Identification for 7.6

I'm going to have to disagree with you on this one. If the organization specified the calibration/verification intervals based on risk then there's a good reason to write a nonconformity if the gages aren't recalled by their due date. It's exceeded the organization's own risk criteria. The organization determined that to go beyond that date was a risk they were unwilling to accept.

Now if the organization specified the intervals based on nonsensical reasons, then it's not the auditor's fault that there's no value in the finding. You suggest that the auditor's job is to identify the risks to the organization. I say that it's the organization's responsibility to identify their risks. If there's no value, the responsibility rests squarely on the organization.
That's the difference between compliance auditing/auditors and value added auditing...
 
S

simonyeeklang

#35
Re: Clause 7.6 - "Shall have Identification in order to determine its calibration sta

Hi,
If you have a piece of expensive software that can export the information out maybe you can do labels using excel.
all you have to do is to copy and paste the information transposed.
and print to sticker then stick to equipment:bigwave:
 
Thread starter Similar threads Forum Replies Date
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 11
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
J AS9104-1 Clause 18.1 (a) (b) - Organizations (contractor) shall be required to allow access to Tier 2 data AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
B IATF 16949 clause 7.1.5.1.1 - Statistical studies shall be conducted IATF 16949 - Automotive Quality Systems Standard 3
M AS9100 Clause 4.2.3 requirements - Documents required by the QMS shall be Controlled AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 13
T Quality Manual shall Outline Structure of Documentation - Clause 4.2.2 ISO 13485:2016 - Medical Device Quality Management Systems 10
P Specified stages of design and development shall be defined - TS 16949 clause 7.3.4.1 IATF 16949 - Automotive Quality Systems Standard 4
R TS16949 Clause 7.6.3.1 - Internal laboratory - Facility shall have a defined scope IATF 16949 - Automotive Quality Systems Standard 5
K OHSAS 18001 - Clause 4.4.3 Communication "...shall be documented..." Occupational Health & Safety Management Standards 5
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
R AS5553 Clause 3.1.7 f - "Implement a returns process....." AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
V Setup for testing against ISO14708 clause 16 (protection of the patient from herms caused by heat) Other Medical Device Related Standards 0
M IATF16949 Clause 9.1.2.1e - Customer notification related IATF 16949 - Automotive Quality Systems Standard 4
N 60601-2-2 Ed 6 Clause 201.8.8.3.102 - HF instrument leakage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
A AS9100D - Clause 8.1 Operation - Coating service company AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
T AS9100D Clause 10.2.1g Supplier Corrective Action for each and every nonconformity? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
H AS9100D clause 8.5.1 f) & 8.5.1 g) - Special Processes AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S ISO 9001 Clause 8.2.3 - Review of the requirements for products and services in a Cafe ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
S API Spec. Q1 clause 5.6.1.2 On site evaluation Oil and Gas Industry Standards and Regulations 10
lanley liao How to understand the clause 6 Planning of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
lanley liao How to understand the term 3.1.21 'Servicing' and the clause 5.7.1.2 'Servicing' of API Spec Q1? Oil and Gas Industry Standards and Regulations 19
earl62 IATF 16949 Clause 9.1.1.1 - What is the batch conformance to specification method? IATF 16949 - Automotive Quality Systems Standard 3
D Layered Process Audits - FCA 9.2.2 - Exemption Clause? IATF 16949 - Automotive Quality Systems Standard 5
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J Scope of ISO 9001 clause 10.2 in the product life cycle ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S MDR GSPR Clause 17 - Software Requirements EU Medical Device Regulations 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
R Clause 7.7 Replicate, Recalibration and Intermediate checks using Artifact ISO 17025 related Discussions 1
A ISO 13485 - Clause 8.3.3 appropriate actions in response to nonconforming product ISO 13485:2016 - Medical Device Quality Management Systems 2
M Description of the requirements of clause 9.2.2.3 manufacturing process audit- needs your feedback IATF 16949 - Automotive Quality Systems Standard 0
A ISO 41001:2018 - Clause No.8 Operations Part Quality Management System (QMS) Manuals 2
L AS5553 Clause 3.1.7 e "Control packaging material ..................reused" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
R What is meant by Use of COMPONENTS WITH HIGH-INTEGRITY CHARACTERISTICS in ME EQUIPMENT (clause no 4.9) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P Definition of production tooling in IATF16949 clause 8.5.1.6 IATF 16949 - Automotive Quality Systems Standard 4
K ISO 13485:2016, Clause 4.2.3 Medical Device File ISO 13485:2016 - Medical Device Quality Management Systems 4
I Master Document Access - ISO 9001:2015 clause 7.5.3 Document Control Systems, Procedures, Forms and Templates 5
qualprod Why new clause (7.1.6) in ISO 9001:2015? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
F ISO 13485 8.2.3 Reporting to regulatory authorities: Question regarding a procedure for this clause. ISO 13485:2016 - Medical Device Quality Management Systems 4
R Process control documents - Context of API Spec Q1, Clause 5.7.1.3 Other ISO and International Standards and European Regulations 0
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
O How can I justify excluding the R&D group and the design and development clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1

Similar threads

Top Bottom