Clean Room as a Controlled Environment - Requirements
Ronen E
Problem Solver
Moderator
#12
Unfortunately MDD annex I s. 8 sheds no additinal light WRT the questions above:
8. Infection and microbial contamination
8.1. The devices and manufacturing processes must be designed in such a way as to eliminate or reduce as far as possible the risk of infection to the patient, user and third parties. The design must allow easy handling and, where necessary, minimize contamination of the device by the patient or vice versa during use.
8.2. Tissues of animal origin must originate from animals that have been subjected to veterinary controls and surveillance adapted to the intended use of the tissues.
Notified bodies shall retain information on the geographical origin of the animals.
Processing, preservation, testing and handling of tissues, cells and substances of animal origin must be carried out so as to provide optimal security. In particular safety with regard to viruses and other transmissible agents must be addressed by implementation of validated methods of elimination or viral inactivation in the course of the manufacturing process.
8.3. Devices delivered in a sterile state must be designed, manufactured and packed in a non-reusable pack and/or according to appropriate procedures to ensure that they are sterile when placed on the market and remain sterile, under the storage and transport conditions laid down, until the protective packaging is damaged or opened.
8.4. Devices delivered in a sterile state must have been manufactured and sterilized by an appropriate, validated method.
8.5. Devices intended to be sterilized must be manufactured in appropriately controlled (e. g. environmental) conditions.
8.6. Packaging systems for non-sterile devices must keep the product without deterioration at the level of cleanliness stipulated and, if the devices are to be sterilized prior to use, minimize the risk of microbial contamination; the packaging system must be suitable taking account of the method of sterilization indicated by the manufacturer.
8.7. The packaging and/or label of the device must distinguish between identical or similar products sold in both sterile and non-sterile condition.
TWA - not the airline
#13
You have to keep in mind that the bioburden is related to the sterilization validation, i.e. your sterilization validation only stays valid as long as the amount and type of germs on your devices (after manufacturing/before sterilization) are the same as during the validation runs (the validation should state detailed acceptance criteria regarding that issue). Now, that does not directly mean you need to control the bioburden in your clean room, but then typically you need to do more monitoring on your finished devices. If you monitor your clean room more closely and have a good link between the bioburden of the clean room and your product then you can reduce the bioburden testing on finished devices, which will save you some money and also alert you much more ealier if something gets out of control...
A Clean Room Gowning Work Instruction - Controlled Document or 'Guideline'? Lean in Manufacturing and Service Industries 18
V Manufacturing requirements for respiratory ventilators - clean room required? Medical Device and FDA Regulations and Standards News 6
C Clean room injection molding, raw material mixing requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
C Raw material Mixing / Preparation for clean room class C injection molding Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
A Construction of an ISO 8 Clean Room - ISO 14644-1 Manufacturing and Related Processes 2
P ISO 5 Pharmaceutical clean room design Process Maps, Process Mapping and Turtle Diagrams 2
O Clean room performance qualification (pq) - How many times can I do each test? Manufacturing and Related Processes 8
O Localized clean room dust extraction system coupled in HVAC - Manufacturing solid dosage form Manufacturing and Related Processes 9
I Clean room recovery program Other Medical Device and Orthopedic Related Topics 1
M ISO 7 Clean Room Requirements? ISO 13485:2016 - Medical Device Quality Management Systems 4
M Scarf in Clean Room instead of Hair cover? Manufacturing and Related Processes 8
M What are ISO 7 Clean Room Requirements Manufacturing and Related Processes 0
S Should the primary drug containers be molded in ISO Class 7 clean room? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 6
S ISO 7 and 8 Clean Room Cleaning Other Medical Device Related Standards 2
S New clean room change submission - FDA Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
A Sterile Garbage Bag for Clean Room Manufacturing and Related Processes 6
D What should an ISO 14644 Clean Room PQ (Performance Qualification) look like? Manufacturing and Related Processes 3
N ISO 14644-1 Clean Room Occupancy State Validation Requirements Manufacturing and Related Processes 3
S What is the minimum pressure differential required for ISO Class 8 Clean Room? ISO 13485:2016 - Medical Device Quality Management Systems 3
V Not using gloves in Clean Room Class 8 (ISO 14644) Manufacturing and Related Processes 21
M Bioburden testing for Class 10000 (Grade C) clean room ISO 13485:2016 - Medical Device Quality Management Systems 1
L How do I set a limits to cfu in an ISO class 7 Clean Room? ISO 13485:2016 - Medical Device Quality Management Systems 2
L Clean Room Monitoring Plan Other Medical Device and Orthopedic Related Topics 6
L AS9100 7.5.2 Special Processes - Clean Room Apply? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
G Clean Room Classification for Contact Lens Packaging ISO 13485:2016 - Medical Device Quality Management Systems 4
D Do I need to test "as built" for an expanded ISO Class 8 Clean Room ? Other Medical Device Related Standards 2
S What are air changes are required for ISO Class 8 Clean Room? US Food and Drug Administration (FDA) 2
A Clean Room Qualification Regulatory Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
L FDA Inspection of Clean Room Processes 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
somashekar Determination of Clean Room Class for Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 1
E Do I need to audit the supplier supplying mop heads for our Clean Room ? ISO 13485:2016 - Medical Device Quality Management Systems 8
M What should be in an SOP for a Machine Operation - LAF Bench in our Clean Room ISO 13485:2016 - Medical Device Quality Management Systems 3
P Clean Room Calibration (Validation) Other ISO and International Standards and European Regulations 4
C Recommended Sealing for Granite Top in a Clean Room US Food and Drug Administration (FDA) 1
L ISO 14644 Class 7 Clean Room Requirements Training Presentation Training - Internal, External, Online and Distance Learning 2
K SPC to apply for Particle in Clean Room Class 100 Statistical Analysis Tools, Techniques and SPC 7
V Updated Clean Room Classification Table of ISO 14644 Other ISO and International Standards and European Regulations 2
N Clean Room Validation and Certification Methods and Definitions ISO 13485:2016 - Medical Device Quality Management Systems 2
S Clean Room Training for Manufacturing Operators (Personal Hygiene, Gowning, etc.) Other Medical Device and Orthopedic Related Topics 4
J Yearly Cleanroom Revalidation Test Requirements - Questions for users clean room ISO7 Qualification and Validation (including 21 CFR Part 11) 5
V AHU (Air Handler Units) and Clean Room References Qualification and Validation (including 21 CFR Part 11) 3
B Hospital Operating Room Regulations - Separate Entrances for Clean and Dirty Supplies Hospitals, Clinics & other Health Care Providers 4
J ISO 14644 - Monitoring Clean Room Particulates (Class 8 Cleanroom) Other ISO and International Standards and European Regulations 4
B Building a Class 10,000 Clean Room Manufacturing and Related Processes 2
A Identifying the Clean Room Requirements for a Medical Device in US and EU ISO 13485:2016 - Medical Device Quality Management Systems 3
G Is Microbiological Monitoring required for Clean Room? Other Medical Device Regulations World-Wide 8
R Clean Room Validation Requirements ISO 13485:2016 - Medical Device Quality Management Systems 7
T Disinfectant usage in a Clean Room Setting procedure ISO 13485:2016 - Medical Device Quality Management Systems 6
jkuil Clean Room Class for Aseptic Preparation Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
G How I am suppose to chose the clean room class for my product? ISO 13485:2016 - Medical Device Quality Management Systems 9