I work in a medical device company with a clean room that classified as a Class 8 ISO. We conduct the revalidation annually, when all the production activities are stopped, which is right on the religious holiday. My boss want us to have the full holiday at least once a year, and started to considering to "reduce" the frequency of the revalidation, since it will be hard if we ask the management to totally stop the production activities beside that religious holiday. We have some plan that we only do the at-rest monitoring (not checking all the sampling points as we usually do in validation, only some points that used in monthly monitoring), then the validation will be held once in 2-3 years, or do the revalidation every 2-3 years by third party.
I was ordered to find any references that could support this plan, it will be good if it based on ISO standard, especially ISO 14644.
In ISO 14644-2:2015 point 5, it stated "Periodic classification test in shall be undertaken annually in accordance with ISO 14644-1. The frequency can be extended based on risk assessment, the extent of monitoring system, and data that are consistently in compliance with acceptance limits or levels defined in the monitoring plan."
Could anyone please explain what that statement means? Does that mean we can do the reclassification/revalidation not in annual basis?
*additional info, our products is a terminally-sterilized class II medical device.
I work in a medical device company with a clean room that classified as a Class 8 ISO. We conduct the revalidation annually, when all the production activities are stopped, which is right on the religious holiday. My boss want us to have the full holiday at least once a year, and started to considering to "reduce" the frequency of the revalidation, since it will be hard if we ask the management to totally stop the production activities beside that religious holiday. We have some plan that we only do the at-rest monitoring (not checking all the sampling points as we usually do in validation, only some points that used in monthly monitoring), then the validation will be held once in 2-3 years, or do the revalidation every 2-3 years by third party.
I was ordered to find any references that could support this plan, it will be good if it based on ISO standard, especially ISO 14644.
In ISO 14644-2:2015 point 5, it stated "Periodic classification test in shall be undertaken annually in accordance with ISO 14644-1. The frequency can be extended based on risk assessment, the extent of monitoring system, and data that are consistently in compliance with acceptance limits or levels defined in the monitoring plan."
Could anyone please explain what that statement means? Does that mean we can do the reclassification/revalidation not in annual basis?
*additional info, our products is a terminally-sterilized class II medical device.