Cleaning and Sterilization - ISO 13485 sections 6.4 and 7.5.1.2.1
chasf
Section 7.5.1.2.1 states: If product is cleaned in accordance with a) or b) above, the requirements contained in 6. 4 a) and 6. 4 b) do not apply prior to the cleaning process.
How does this apply to our production setting?
Do 6.4 a and 6.4 b apply?
What would be the extent of documented controls needed?
Our company supplies components to the Medical Device manufacturer they are cleaned by us in an ultrasonic cleaner and in some cases passivated by our certified supplier then sent to the the customer. All sterilization takes place after shippment. We use controls for handling after cleaning takes place by use of protective gloves to keep device components in a clean condition.
Jimmy the Brit
Hi Chasf,
What I believe this clause is saying is that if you are cleaning your product using a validated cleaning process prior to sterilization or use, then you do not need to worry about control of the manufacturing environment or personnel hygiene prior to the washing stage.
In reality the burden of proof is on you - you must demonstrate that the U/S washing or passivation stages (I am assuming acid passivation??) are adequate to remove all contaminants from your product. I would recommend a worst case validation to test the limits of your process - I once coated injection moulded components in oil and releasing agents, just to see if the dry cleaning fluid in our U/S bath was robust enough to remove it all - it was, thankfully.
The important part is recovering the clean components and handling them in such a way to avoid recontamination prior to sterilization, because the ISO 13485 sections 6.4 a) and b) definitely apply post-washing. You use gloves and this may well be adequate for your process, but I suggest that you validate it, probably using a bioburden recovery as this is the easiest way to prove product protection. In my experience the critical thing is to ensure that the product is dry before packing as wet product will rapidly become contaminated (in 4 to 6 hours).
If there is a sterilization stage later in the process there should be pre-sterilization bioburden records available - this could give you the data you need to retrospectively validate the efficacy of your washing and recovery process (or it could show that bioburden is not well controlled, in which case your process has a hole in it).
I hope this helps, and I haven't just made your life a whole lot more complicated, ;)
Jim
chasf
Thanks for your concern but things here were already complicated as we make the transition from ISO 9001:2000 to ISO 13485:2003.
Our manufacturing process involves contact with synthetic water soluble coolants and or synthetic oils also stainless steel shot peen media as well as handling buy 4-6 different people.
In this type of setting what steps would we take to control contamination in relation to personal hygiene?
Would this control need to be validated?
Another question that comes to mind is if we did satisfy the requirements of 6.4 would we still need to validate our cleaning processes?
Jimmy the Brit
The easiest way to address the handling/personal hygiene issue is to separate the product from the personnel, post-washing, but of course that is expensive and technically demanding. If you recover washed components in a clean room, or laminar flow hood and immediately store into impermeable packaging then you only need to demonstrate control up until the point of sealing the primary pack - but you then have to valiidate and control the clean room etc...
You indicated in your earlier mail that you were using gloves to recover washed components, presumably to seal them into a polythene bag or similar. If your gloves are sterile, or at least non-linting and disinfected, regularly, and your bags are clean single use, then this may be adequate, as long as the bioburden downstream is within the specification required by your sterilization process.
As a personal comment, if I was auditing a facility and saw a cleaning process I would always ask for a justification and validation data for the process. You must be washing for a reason, and therefore I would want to see evidence that your process acheived its intended purpose.
It doesn't have to be too demanding to validate the cleaning process - as I said before, check the pre-sterilization bioburden, this may provide excellent data to support your current practices.
Jim
A Reprocessing, Cleaning, Disinfection and Sterilization Validation EU Medical Device Regulations 4
M Cleaning of Orthopaedic Implants before Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 15
A Guidance on Cleaning & Sterilization of an FDA Class II Medical Device US Food and Drug Administration (FDA) 4
C Nerve stimulation Implant cleaning services ISO 13485:2016 - Medical Device Quality Management Systems 3
D Preventive Maintenance vs. General Cleaning ISO 13485:2016 - Medical Device Quality Management Systems 2
S Cleaning process center change ISO 13485:2016 - Medical Device Quality Management Systems 4
C Cleaning of electronic transmitter Other Medical Device Related Standards 2
S California Cleaning Product Right to Know Act - Product Label advice needed Miscellaneous Environmental Standards and EMS Related Discussions 1
D Cleanroom Cleaning Products and Storage Other Medical Device and Orthopedic Related Topics 18
R 510(k) Biocompatibility vs Cleaning Validation Other Medical Device Related Standards 4
M Cleaning Process after laser marking on orthopaedic implants Other Medical Device and Orthopedic Related Topics 3
John Broomfield Cleaning plastic from our oceans - The Ocean Cleanup Venture Sustainability, Green Initiatives and Ecology 33
E Cleaning Validation Process of Non-Sterile Implants and Instruments Other Medical Device Related Standards 2
E Labware cleaning validation protocol example Qualification and Validation (including 21 CFR Part 11) 1
A Cleaning plastic parts for sterilized medical devices ISO 13485:2016 - Medical Device Quality Management Systems 15
C Documenting Cleaning Production Lines in Medical Device Pkg facility ISO 13485:2016 - Medical Device Quality Management Systems 1
C Documenting Cleaning Production Lines in Medical Device Pkg facility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
C Documenting Cleaning Production Lines in Medical Device Pkg facility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D ISO 13485 production cleaning requirements of reusable devices ISO 13485:2016 - Medical Device Quality Management Systems 2
P Production Line Pump Cleaning Validation Method Question Qualification and Validation (including 21 CFR Part 11) 0
S ISO 7 and 8 Clean Room Cleaning Other Medical Device Related Standards 2
I Material for PUW used to cleaning Other Medical Device and Orthopedic Related Topics 1
I Implantable medical device cleaning validation Other Medical Device Related Standards 3
S Looking for ISO Class 8 Cleanroom Cleaning Procedure example ISO 13485:2016 - Medical Device Quality Management Systems 1
P Cleaning and Disinfection Validation Sample Size of 3? Other Medical Device Related Standards 2
O Cleaning API Rotary Shoulder Connections Working Mates exposed to Outside Weather General Measurement Device and Calibration Topics 1
R Cleaning validation for Molding Tools? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
P Best Method of Cleaning of Thread Ring Gages Calibration Frequency (Interval) 5
E Cleaning Validations for Reusable Products - ANSI/AAMI TIR30:2011 Requirements Other Medical Device Related Standards 3
C Small Cast Iron Part Cleaning Manufacturing and Related Processes 7
S Cleaning Fluids and RoHS2 Declaration Requirements RoHS, REACH, ELV, IMDS and Restricted Substances 2
A Cleaning of Medical Device Storage Areas ISO 13485:2016 - Medical Device Quality Management Systems 1
N Using Ozone as a Cleaning Agent in Purified Water Manufacturing and Related Processes 2
R Best Practices in the Control of Cleaning Materials Document Control Systems, Procedures, Forms and Templates 1
W Production Equipment Cleaning Protocol Example Needed Other Medical Device and Orthopedic Related Topics 2
V Standard for Cleaning and Washing Metal Stampings Other ISO and International Standards and European Regulations 2
A Cleaning instructions for our non sterile, single use products EU Medical Device Regulations 6
W Cleanroom Cleaning Agents for cleaning a Class 7 Cleanroom Other Medical Device and Orthopedic Related Topics 9
G ISO 17664 Reusable Low Risk Medical Device Cleaning Other Medical Device Related Standards 2
B Isopropyl Alcohol cleaning of Peek Optima ISO 13485:2016 - Medical Device Quality Management Systems 5
N Seeking information on use of Byotrol Cleaning Chemicals in Pharma Industry Manufacturing and Related Processes 1
S Glass Bead Blast Cleaning Spec TT-C-490 Rev D Various Other Specifications, Standards, and related Requirements 1
L Can I produce Cosmetics and Household Cleaning Products in the same Facility? US Food and Drug Administration (FDA) 4
L Nutritional, Cosmetics and Cleaning Products in Same Facility Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
G Sampling Plan - Sample Size Justification for a Bottling Cleaning Process Validation Qualification and Validation (including 21 CFR Part 11) 2
M Cleaning Resuable Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 11
I Cleaning Validation for Durapore Tape Cutting Machine EU Medical Device Regulations 2
R Cleaning Validation Protocols for a Pharmaceutical Company Qualification and Validation (including 21 CFR Part 11) 2
S Using 93 % alcohol for cleaning Printer Head Manufacturing and Related Processes 2
E Ultrasonic Cleaning of Orthopedic Implants - Detergents and Frequency - How? Other Medical Device and Orthopedic Related Topics 3