Cleaning process center change
Our Knee System products are approved FDA and CE,
I would like to know one thing regarding the Cleaning process, we would like to change the cleaning process center then what are the Re validations needed to our approved system?
As per our knowledge ,change in the cleaning process will affect the biocompatibility so we are thinking to do the biocompatibility tests and cleaning re validation .
My question is Do we need to do revalidation on sterilization (Gamma and Eto)?
planB
srniu,
the response is: it depends.
If you come to the conclusion in your risk assessment, that the change in cleaning process
(1) might adversely affect the safety and efficacy of your cleaning process, then yes, you have to re-validate and also potentially re-evaluate the biological safety of your final product.
(2) has a potentially adverse impact on your subsequent validated sterilization process, then you would have to cover these aspects in re-validation activities. An example might be, that your intended change alters the bioburden / bacterial endotoxin levels of your product.
The extent of re-validation activities depends on the exact nature of your intended change (e.g. is it "just" a site change where you move your existing process to, or do you intend to use a different cleaning process?) and the associated risk assessment for this change.
You might also have a look into this thread and the change management sections in ISO 10993-1, ISO 11137-1 and ISO 11135, respectively.
Hope this helps,
Thanks @planB for your clarification..
We are changing the site with same conditions of cleaning process, only thing is in our risk management ,we are planning to verify like
1. we are thinking to do the Endotoxin test and check with our old report ,if the result are equivalent then the process is not effect
2.Another way to do the cleaning process validation and check the result with previous result, if the results are equivalent then the process in not effect
My question is which way is the best to prove the changed process is safe and avoid biological and sterilization revalidations?
If you put some light on this, it's more useful for me to conclude this risk assessment.
Thank you.
planB
Srinu,
you might want to gather objective evidence from your supplier that supports the conclusion that you are just transferring an unchanged process to a new site.
This could include (but not limited to):
- IQ, OQ, PQ data of the equipment at the new site, demonstrating equivalent functionality
- process at the new site: any OQ, PQ data available that the process conforms to the very same parameter specification? Any In-process monitoring data (such as conductivity, TOC, turbidity, ...) that would substantiate equivalency?
ad (1) In case (bacterial) endotoxins are one validation endpoint of your cleaning process, then yes, re-testing endotoxin levels at nominal (or at worst-case conditions) might help you building your case. What other validation acceptance criteria are in place?
ad (2) yes, this would be an option.
Hope this helps,
M Cleaning Process after laser marking on orthopaedic implants Other Medical Device and Orthopedic Related Topics 3
E Cleaning Validation Process of Non-Sterile Implants and Instruments Other Medical Device Related Standards 2
G Sampling Plan - Sample Size Justification for a Bottling Cleaning Process Validation Qualification and Validation (including 21 CFR Part 11) 2
F Airport Internal Audit Checklist (Cleaning, Project Process, Maintenance, etc) Internal Auditing 3
I Validation of Ultrasonic Cleaning Process for Basic Tools & Buffing Compound ISO 13485:2016 - Medical Device Quality Management Systems 5
B 5S Cleaning Standards Documentation and the 5S Process Quality Tools, Improvement and Analysis 7
C Nerve stimulation Implant cleaning services ISO 13485:2016 - Medical Device Quality Management Systems 3
D Preventive Maintenance vs. General Cleaning ISO 13485:2016 - Medical Device Quality Management Systems 2
C Cleaning of electronic transmitter Other Medical Device Related Standards 2
S California Cleaning Product Right to Know Act - Product Label advice needed Miscellaneous Environmental Standards and EMS Related Discussions 1
D Cleanroom Cleaning Products and Storage Other Medical Device and Orthopedic Related Topics 18
A Reprocessing, Cleaning, Disinfection and Sterilization Validation EU Medical Device Regulations 4
R 510(k) Biocompatibility vs Cleaning Validation Other Medical Device Related Standards 4
John Broomfield Cleaning plastic from our oceans - The Ocean Cleanup Venture Sustainability, Green Initiatives and Ecology 33
E Labware cleaning validation protocol example Qualification and Validation (including 21 CFR Part 11) 1
A Cleaning plastic parts for sterilized medical devices ISO 13485:2016 - Medical Device Quality Management Systems 15
C Documenting Cleaning Production Lines in Medical Device Pkg facility ISO 13485:2016 - Medical Device Quality Management Systems 1
C Documenting Cleaning Production Lines in Medical Device Pkg facility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
C Documenting Cleaning Production Lines in Medical Device Pkg facility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D ISO 13485 production cleaning requirements of reusable devices ISO 13485:2016 - Medical Device Quality Management Systems 2
P Production Line Pump Cleaning Validation Method Question Qualification and Validation (including 21 CFR Part 11) 0
S ISO 7 and 8 Clean Room Cleaning Other Medical Device Related Standards 2
I Material for PUW used to cleaning Other Medical Device and Orthopedic Related Topics 1
I Implantable medical device cleaning validation Other Medical Device Related Standards 3
S Looking for ISO Class 8 Cleanroom Cleaning Procedure example ISO 13485:2016 - Medical Device Quality Management Systems 1
P Cleaning and Disinfection Validation Sample Size of 3? Other Medical Device Related Standards 2
O Cleaning API Rotary Shoulder Connections Working Mates exposed to Outside Weather General Measurement Device and Calibration Topics 1
R Cleaning validation for Molding Tools? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
P Best Method of Cleaning of Thread Ring Gages Calibration Frequency (Interval) 5
E Cleaning Validations for Reusable Products - ANSI/AAMI TIR30:2011 Requirements Other Medical Device Related Standards 3
C Small Cast Iron Part Cleaning Manufacturing and Related Processes 7
S Cleaning Fluids and RoHS2 Declaration Requirements RoHS, REACH, ELV, IMDS and Restricted Substances 2
A Cleaning of Medical Device Storage Areas ISO 13485:2016 - Medical Device Quality Management Systems 1
N Using Ozone as a Cleaning Agent in Purified Water Manufacturing and Related Processes 2
R Best Practices in the Control of Cleaning Materials Document Control Systems, Procedures, Forms and Templates 1
W Production Equipment Cleaning Protocol Example Needed Other Medical Device and Orthopedic Related Topics 2
V Standard for Cleaning and Washing Metal Stampings Other ISO and International Standards and European Regulations 2
A Cleaning instructions for our non sterile, single use products EU Medical Device Regulations 6
W Cleanroom Cleaning Agents for cleaning a Class 7 Cleanroom Other Medical Device and Orthopedic Related Topics 9
G ISO 17664 Reusable Low Risk Medical Device Cleaning Other Medical Device Related Standards 2
B Isopropyl Alcohol cleaning of Peek Optima ISO 13485:2016 - Medical Device Quality Management Systems 5
N Seeking information on use of Byotrol Cleaning Chemicals in Pharma Industry Manufacturing and Related Processes 1
S Glass Bead Blast Cleaning Spec TT-C-490 Rev D Various Other Specifications, Standards, and related Requirements 1
L Can I produce Cosmetics and Household Cleaning Products in the same Facility? US Food and Drug Administration (FDA) 4
L Nutritional, Cosmetics and Cleaning Products in Same Facility Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
M Cleaning Resuable Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 11
I Cleaning Validation for Durapore Tape Cutting Machine EU Medical Device Regulations 2
R Cleaning Validation Protocols for a Pharmaceutical Company Qualification and Validation (including 21 CFR Part 11) 2
S Using 93 % alcohol for cleaning Printer Head Manufacturing and Related Processes 2
E Ultrasonic Cleaning of Orthopedic Implants - Detergents and Frequency - How? Other Medical Device and Orthopedic Related Topics 3