Cleanroom Question
Milas
Hi
We are manufacturing some products in a class 8 cleanroom and regularly find contamination. The contamination, based on texture, looks like human hair and in some cases look like fibres that has come from the clothing which the operators are wearing. We have raised this issue many a time with our colleagues in Quality who continually monitor and assess the cleanrooms and claim there are no issues. Question is, should such contamination be detected during routine checks and assessments i.e. flow and pressure tests ?
Jen Kirley
Quality and Auditing Expert
Hello Milas,
Pardon me, it is late in the day: flow and pressure tests are designed to check what, besides flow and pressure? Seriously, particle size is a different type of check isn't it?
My understanding is that ISO 8 cleanrooms, also known as Class 100,000 cleanrooms, can be modular or soft-walled and have a maximum particle count of 100,000 particles (≥0.5 um) per cubic foot of interior air. They are required to have HEPA filtration coverage over 4-5% of the area and provide a minimum of 20 air exchanges per hour with air flow rates of 4-8 CFM per square foot.
What are your clean room controls?
ChrisM
Cleanroom air quality monitoring with the "usual" type of machine won't find large, settled contaminants such as human hairs. What controls on clothing etc are in place? I hope that you have a gowning area where cleanroom clothing including hair coverings etc must be donned before entering the clean room itself. Do you have a procedure that checks for settles contaminants such as debris from the building structure, human hairs, remnants of clothing etc?
AgnieszkaSz
Air cleanliness in cleanroom is just a part of the system targeting product cleanliness and safety. It detects airborne particles, nothing else. If hair and fibres are found, I would advise tightening gowning rules - coveralls, dedicated shoes, head caps - and checking whether the gowning rules are observed. Our operators are not allowed to wear jeans because cotton sheds fibers even when it is covered by coveralls. Bare legs are not acceptable, either, they shed enormous quantities of hair.
John Predmore
Hairs and other heavy fibers might not show up on filter tests and particle counters. You should have tacky mats at the entrance to your cleanroom which will trap loose hair/fibers sliding around on the floor from air currents, as well as fibers on the bottom of shoes. You can differentiate human hair from other fibers under a microscope. Instead of throwing mats away, maybe you can identify fibers (pull the big ones off with tweezers) and a Pareto chart will tell you the greatest source of contamination. Wrap a loop of clear shipping tape around your hand and pat work surfaces on your work stations for a spot-check sampling of contamination, close to where you manufacture product. As I recall, over 70% of clean room contamination comes from human sources.
ChrisM
I suspect that a vast majority of cleanroom contamination will indeed be human skin, since we are shedding and renewing our skin constantly, hence the need for hair nets, covered legs and arms etc (but not full face coverings and not gloves on hands in the lower levels of clean rooms)
Milas
Thank you for you feedback
We do have cleanroom controls in that, operators have a separate area for gowning prior to entering the cleanroom. However, we do not have a procedure which checks for settled contaminants nor, do our operators wear full cleanroom gowns, they wear hair nets and 3/4 lengths overalls
optomist1
A Sea of Statistics
Yep, great post, as I recall my first post graduation position...."shave the mustache, and a list of approved/prohibited clothing etc., this was for all clean room levels 100K to 100, a good procedure indeed...
Hope this helps
Optomist1
optomist1
A Sea of Statistics
Hi indubiosh, off to a meeting but will provide info later today, still pretty fresh in my mind though...Cheers Optomist 1
M Cleanroom Gloves Manufacturing and Related Processes 2
P Setting Cleanroom Alert and Action Limits Manufacturing and Related Processes 0
B Do I need a Cleanroom to manufacture Medical Devices? ISO 13485:2016 - Medical Device Quality Management Systems 2
D ESD cleanroom Manufacturing and Related Processes 4
B Sub-contractor cleanroom monitoring ISO 13485:2016 - Medical Device Quality Management Systems 0
D Cleanroom requirements for surgical apparel EU Medical Device Regulations 2
N Cleanroom Classification Change Other Medical Device Related Standards 7
P Using Zinc Plated Steel in a Cleanroom Other Medical Device Related Standards 3
J Cleanroom Differential Pressure Monitoring, frequent manual observation US Food and Drug Administration (FDA) 6
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
C Cleanroom - 209E to ISO14644 Other Medical Device Related Standards 2
P Surgical masks ( type IIR) - Cleanroom Requirements EU Medical Device Regulations 3
J Need for a cleanroom in the manufacture of a medical device for a clinical trial EU Medical Device Regulations 4
D Cleanroom Cleaning Products and Storage Other Medical Device and Orthopedic Related Topics 18
chris1price Cleanroom Monitoring Plan - ISO14644-2:2015 - Risk Assessment Other Medical Device Related Standards 3
C Looking for source of silicone contamination in ISO Class 8 Cleanroom Other Medical Device Related Standards 2
I Cleanroom Actions in case of Actions Limit Alarm Qualification and Validation (including 21 CFR Part 11) 3
W How to bring Cardstock into Cleanroom Environments Manufacturing and Related Processes 1
S Looking for ISO Class 8 Cleanroom Cleaning Procedure example ISO 13485:2016 - Medical Device Quality Management Systems 1
D Use of hand cream inside gloves in the cleanroom for medical devices Occupational Health & Safety Management Standards 7
T Converting an ISO Class 8 Cleanroom to ISO 7 Various Other Specifications, Standards, and related Requirements 4
jkuil Cleanroom Equipment and Furniture Engineering Standards Other Medical Device Related Standards 2
T Cleanroom Certification Duration Other ISO and International Standards and European Regulations 1
E ISO Class 8 Cleanroom Qualification Protocol Template Other Medical Device and Orthopedic Related Topics 8
B Cleanroom Control Limits: Cumulative or individual? Statistical Analysis Tools, Techniques and SPC 11
I FDA Cleanroom requirements for class II in vitro medical devices US Food and Drug Administration (FDA) 1
C Cleanroom Disaster Recovery for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 8
A Guidance for Cleanroom Qualification and Validation relevant for EU and USA ISO 13485:2016 - Medical Device Quality Management Systems 2
E Is Volumetric Air Sampling required for Class ISO 7 Cleanroom ? ISO 13485:2016 - Medical Device Quality Management Systems 1
B How to set Cleanroom Alert and Action Limits as part of Microbiological Evaluation Quality Assurance and Compliance Software Tools and Solutions 8
C What is an acceptable bioburden level or count when working in cleanroom ISO class7 ISO 13485:2016 - Medical Device Quality Management Systems 1
W ISO 14644 Cleanroom Validation - "Operational" Other Medical Device Related Standards 3
E A sink with water in an ISO Class 7 Cleanroom ISO 13485:2016 - Medical Device Quality Management Systems 4
W Thoughts on having Safety Shower in ISO Class 7 Cleanroom ISO 13485:2016 - Medical Device Quality Management Systems 5
C Cost Factors of Iso 8 vs. ISO 8/9 Cleanroom Other Medical Device and Orthopedic Related Topics 2
W Cleanroom Cleaning Agents for cleaning a Class 7 Cleanroom Other Medical Device and Orthopedic Related Topics 9
V Making Solar cells, Module, String - Which Cleanroom Level? Various Other Specifications, Standards, and related Requirements 2
H What is the ISO Class 7 cleanroom requirement on Temperature and Humidity ? Manufacturing and Related Processes 4
J Do I need a Cleanroom to manufacture Medical Devices? ISO 13485:2016 - Medical Device Quality Management Systems 12
C ISO 13485 Class 7 Cleanroom Requirements ISO 13485:2016 - Medical Device Quality Management Systems 7
J Yearly Cleanroom Revalidation Test Requirements - Questions for users clean room ISO7 Qualification and Validation (including 21 CFR Part 11) 5
J ISO 14644 - Monitoring Clean Room Particulates (Class 8 Cleanroom) Other ISO and International Standards and European Regulations 4
D Cleanroom Housekeeping and Personnel Hygiene Audit Checklist Other Medical Device Related Standards 25
C Cleanroom Class 8 (100,000) Particulate Control and Warning Limits Statistical Analysis Tools, Techniques and SPC 1
N Dehumidified (Dry) Cleanroom - Running a Cleanroom at very low RH Manufacturing and Related Processes 23
somashekar Use of Pencil in Class 1000 Cleanroom ISO 13485:2016 - Medical Device Quality Management Systems 12
S Cost Surrounding Cleanroom Construction Manufacturing and Related Processes 4
M Cleanroom Qualification and Standard Cleanroom Audit Checklist Process Audits and Layered Process Audits 6
C Pest Control - FDA approved list of Pesticides to use for Pest Pontrol in a Cleanroom Other Medical Device and Orthopedic Related Topics 10
B Checklist for Housekeeping and Cleaning of Class 10k Cleanroom? Other Medical Device and Orthopedic Related Topics 17