Then what's the difference in between VALIDATION and VERIFICATION?

So....., here's my simple thoughts,

Is that clinical accuracy should be validated, to get value in a correct way?
There might not be acceptance criteria defined by ISO standards, we only need to meet customer's requirement.

And that laboratory accuracy must be verified, to get the correct value?
There is indeed acceptance criteria defined by ISO standards, we need to follow that.
I think you are getting on the right track. Verification answers the question, "Did we make the device right?" Validation answers the question, "Did we make the right device?" The validation will test whether the device performs as intended when used in the appropriate setting by the appropriate users.
I suggest you read the US FDA's guidance document on design control. I really think this will help you gain a better understanding.
https://www.fda.gov/media/116573/download
 
Elsmar Forum Sponsor

Cybel

Starting to get Involved
#12
When plan the clinical validation study of IR forehead thermometer, all procedures are based on ISO 80601-2-56 or ASTM-E1965-98.

These standards recommend explicit methods for the determination of laboratory accuracy, clinical accuracy and clinical repeatability.
However, only laboratory accuracy is assigned an maximum error criteria ( ±0,3 °C ) within the RATED OUTPUT RANGE.
For clinical accuracy and repeatability, no maximum error criteria is recommended.
So I don't know if the clinical result of IR forehead thermometer is acceptable or not.


Is there any explanations?
Hi.
From my understanding, in the clinical accuracy validation there is not a PASS/NON PASS threshold.
ISO 80601-2-56:2017 states that "The results of CLINICAL ACCURACY VALIDATION shall be disclosed in the ACCOMPANYING DOCUMENT. This disclosure shall include: the CLINICAL BIAS, Δcb, with its LIMITS OF AGREEMENT, LA, the CLINICAL REPEATABILITY, σr, the REFERENCE BODY SITE and MEASURING SITE for each OPERATING MODE."

It appears this means that it's up the user decision if the device is accurate enough or not.
This support the track of "customer requirements" as suggested by indubioush.
 
Thread starter Similar threads Forum Replies Date
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
P GSPRs / Clinical Evaluation EU Medical Device Regulations 3
P IDEs and Clinical Investigators - Advice About Records? Other US Medical Device Regulations 3
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 6
P Clinical Trial Requirement - No equivalent devices in the market CE Marking (Conformité Européene) / CB Scheme 3
M Clinical Decision Support Software Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Regular updates of clinical evaluation report EU Medical Device Regulations 6
M Clinical Benefit of device that only aids in a process for managing or treating disease EU Medical Device Regulations 2
C EU MDR - Annex II 6.1 Pre-clinical and clinical data EU Medical Device Regulations 4
M Clinical Evaluation Benchmark vs. Equivalent EU Medical Device Regulations 2
D Classification of product for clinical trials EU Medical Device Regulations 14
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5
S Australia TGA Clinical Evaluation Report (CER) Other Medical Device Regulations World-Wide 0
L Clinical research - Request from ANVISA ISO 14971 - Medical Device Risk Management 1
E Use of OQ cleared batch for stability and clinical study use ISO 13485:2016 - Medical Device Quality Management Systems 0
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
M Clinical Development Plan - Advice Requested EU Medical Device Regulations 11
T Clinical evaluation of a new medical device EU Medical Device Regulations 0
dgrainger Informational MDCG 2020:13 - Clinical evaluation assessment report template EU Medical Device Regulations 0
shimonv Minimum 60601 requirements for a clinical trial IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S A clinical performance evaluation study with an IVD product as Investagional Use product - Clinical Monitor requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
J Need for a cleanroom in the manufacture of a medical device for a clinical trial EU Medical Device Regulations 4
R Impact of changing manufacturer post clinical investigation CE Marking (Conformité Européene) / CB Scheme 3
C MDCG 2020-06 Clinical evidence legacy devices EU Medical Device Regulations 3
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 8
M Clinical Evaluation Plan vs. PMCF Plan EU Medical Device Regulations 24
S Selling Clinical Practice Advice Other US Medical Device Regulations 7
S FDA's E submitter - For clinical electronic thermometer Medical Device and FDA Regulations and Standards News 6
M Good Clinical Practice (GCP) - ISO 14155 vs. ICH Other ISO and International Standards and European Regulations 10
S Article 61 - MDR Clinical Investigation Requirement EU Medical Device Regulations 5
J Are DAM tools, Mobile application, clinical dashboard, etc. considered accessories to medical device or supportive functions? Manufacturing and Related Processes 4
M Can someone share a scrubbed version of Clinical Evaluation Plan (CEP) EU Medical Device Regulations 8
R Do Medical Device Complaints include Clinical Trials? Medical Device and FDA Regulations and Standards News 3
S EU MDR Annex XIV - Clinical Evaluation Plan - What do these methods mean? EU Medical Device Regulations 16
dgrainger Informational MHRA updated guidance: Clinical investigations of medical devices – guidance for manufacturers Medical Device and FDA Regulations and Standards News 0
vickyva Checklist for Clinical Investigation compliant to EU-MDR EU Medical Device Regulations 4
U Do we need clinical trial data for Class IIa medical device under MDR EU Medical Device Regulations 7
M Risk Classification For Supplier - Clinical Research Organisation (CRO) Supply Chain Security Management Systems 3
A Post-market clinical followup studies (PMCF) EU Medical Device Regulations 16
Ed Panek Adverse Event Clinical Trial using a 510K approved Device Other US Medical Device Regulations 6
S Clinical Evaluation - Is this an ISO 13485:2016 requirement? ISO 13485:2016 - Medical Device Quality Management Systems 4
M Informational How to perform a clinical evaluation of medical devices – Part 2 – Level of clinical evidence and what sufficient clinical evidence means Medical Device and FDA Regulations and Standards News 9
M EU – Chemical Safety Report – CSR – REACH Authorisation decisions – Triton X-100 – Ortho-clinical-Use1 REACH and RoHS Conversations 0
M One clinical design: Russia and Europe Other Medical Device and Orthopedic Related Topics 0
T Clinical trial - Medical device product not cleared (without FDA approval) in a drug trial Other US Medical Device Regulations 9
S Level of Clinical Evidence - MDR EU Medical Device Regulations 3
M Informational US FDA Draft Guidance – Clinical Decision Support Software Medical Device and FDA Regulations and Standards News 0

Similar threads

Top Bottom