We are working with a company who produces a regular device (lamp) which now has class I indication. We want to use this lamp as a medical device and make it into a class II (a or b) device. The claim is that the lamp which is an indoor lamp replicates daylight and operates in the same spectrum. My question is:
When we expand the medical claims, what is necessary from a clinical perspective? Can we rely on published literature solely?
Or do we need to do our own clinical trials?
Is it enough to show equivalence between the two spectrums (daylight and the spectrum the lamp operates on) technically?
Do you have any examples of devices who has used the support of other published reports?

Thanks!
 
Elsmar Forum Sponsor
#2
Your classification is dependent upon the medical purpose of your lamp.

Is it intended for therapeutic use? (Treatment of SAD; seasonal affective disorder). This would be Class IIa (Rule 9).

If it is for illumination purposes in the visible spectrum it is Class I (Rule 10).
 

Ronen E

Problem Solver
Staff member
Moderator
#3
Can we rely on published literature solely?
Or do we need to do our own clinical trials?
There is no overwhelming requirement to conduct clinical trials. In some cases the clinical evidence / clinical evaluation may be purely literary.

Recently the clinical evaluation bar has been raised, when rev. 4 of the official MEDDEV guidance was released. The answer to your question is in there, and depends on the device's details and claims.
 

Kees Fremerij

Starting to get Involved
#4
What I read is that it is intended for illumination purposes only. Following the rules would say Class I according to rule 12:

Active diagnostic devices intended to illuminate the patient's body in the visible spectrum such as examination lights

The question is not: what classification you would prefer but where is the device classified in.
 

MDD_QNA

Involved In Discussions
#5
What I read is that it is intended for illumination purposes only. Following the rules would say Class I according to rule 12:

Active diagnostic devices intended to illuminate the patient's body in the visible spectrum such as examination lights

The question is not: what classification you would prefer but where is the device classified in.
It is a class I device now and because of the expanded medical claim, it will probably be a class IIa medical device.
 

dgrainger

Trusted Information Resource
#6
It will depend on your new claims but the requirement is given in Annex X:

1.1.1. Either a critical evaluation of the relevant scientific literature currently available relating to the safety, performance, design characteristics and intended purpose of the device, where:
— there is demonstration of equivalence of the device to the device to which the data relates, and
— the data adequately demonstrate compliance with the relevant essential requirements.
1.1.2. Or a critical evaluation of the results of all clinical investigations made.
1.1.3. Or a critical evaluation of the combined clinical data provided in 1.1.1
and 1.1.2.

Unless you can show equivalence, you will need clinical data.
 
Thread starter Similar threads Forum Replies Date
M Category I Medical Device Clinical Data Requirements EU Medical Device Regulations 3
C EU MDR - Annex II 6.1 Pre-clinical and clinical data EU Medical Device Regulations 4
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 8
U Do we need clinical trial data for Class IIa medical device under MDR EU Medical Device Regulations 7
CPhelan Using clinical trial safety data for evidence for CE marking EU Medical Device Regulations 8
CPhelan Do you require MDSAP for CE Marking of a Medical Device or is ISO13485:2016 with clinical data sufficient? CE Marking (Conformité Européene) / CB Scheme 6
M Informational Health Canada begins release of clinical data Medical Device and FDA Regulations and Standards News 0
M Informational USFDA Final Rule – Human Subject Protection; Acceptance of Data From Clinical Investigations for Medical Devices Medical Device and FDA Regulations and Standards News 0
M IVD - Generating clinical data for IVD’s in the EU EU Medical Device Regulations 2
W ISO 14155 - Electronic clinical data systems - Definition Other Medical Device Related Standards 0
Y What Clinical Data is required for Drug Delivery System EU Medical Device Regulations 1
D Use of clinical data in a 510(k) submission Other US Medical Device Regulations 6
bio_subbu USFDA issues draft Guidance on the Acceptance of Medical Device Clinical Data Other US Medical Device Regulations 1
R FDA's guidance - Evaluation of Sex-Specific Data in Medical Device Clinical Studies Other US Medical Device Regulations 3
B Health Canada Acceptance Policy of Foreign Clinical Trial Data for Devices Canada Medical Device Regulations 2
C Retention of Clinical Trial Data after a Clinical Trial has ended Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 5
C Key Links (Library) to search for updated Clinical Data (Is & IIa) EU Medical Device Regulations 1
E Need Corrective Action for Class III Device with no Clinical Data CE Marking (Conformité Européene) / CB Scheme 9
S Clinical or Non-Clinical Data in 510(k) Submission for a type II Medical Device US Food and Drug Administration (FDA) 1
D Registration of Class IV Device -Clinical Data and ISO Certificate Canada Medical Device Regulations 3
A Collecting Clinical Data using CE Marked Devices EU Medical Device Regulations 4
H 510(k) Filing and Predicate Device Clinical Data 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Technical File as per ANNEX VII and Sample Size for clinical data CE Marking (Conformité Européene) / CB Scheme 12
C Clinical Data to Prove Equivalence Other US Medical Device Regulations 4
N Clinical Data for Class 1 Devices? CE Marking (Conformité Européene) / CB Scheme 6
I 510k Approval - Does the FDA look for Clinical Data 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
L Clinical Data Report - Literature Route - MEDDEV guidance (MEDDEV 2.7.1) EU Medical Device Regulations 28
E ISO 13485 in Clinical Trial conduct: Applicable or No ISO 13485:2016 - Medical Device Quality Management Systems 2
L Clinical Trial and Finished Devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
P GSPRs / Clinical Evaluation EU Medical Device Regulations 3
P IDEs and Clinical Investigators - Advice About Records? Other US Medical Device Regulations 3
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 6
P Clinical Trial Requirement - No equivalent devices in the market CE Marking (Conformité Européene) / CB Scheme 3
M Clinical Decision Support Software Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Regular updates of clinical evaluation report EU Medical Device Regulations 6
M Clinical Benefit of device that only aids in a process for managing or treating disease EU Medical Device Regulations 2
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
M Clinical Evaluation Benchmark vs. Equivalent EU Medical Device Regulations 2
D Classification of product for clinical trials EU Medical Device Regulations 14
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5
S Australia TGA Clinical Evaluation Report (CER) Other Medical Device Regulations World-Wide 0
L Clinical research - Request from ANVISA ISO 14971 - Medical Device Risk Management 1
E Use of OQ cleared batch for stability and clinical study use ISO 13485:2016 - Medical Device Quality Management Systems 0
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
M Clinical Development Plan - Advice Requested EU Medical Device Regulations 11
T Clinical evaluation of a new medical device EU Medical Device Regulations 0

Similar threads

Top Bottom