Hello everyone,
This is the first time I have ever done this, so please excuse my clumsiness with the process...
A company is improving its device and wants to prove equivalence to the older model for safety and effectiveness. The manufacture has hundreds of data files generated from patients who have used the predicate device and these same data files are able to be run through the new device to obtain the same results. Is this an acceptable way of proving the new device is substantially equivalent to the old device? If so, how many records should be used in the population for the test? Are there any other rules or issues to be aware of?
Thank you all for your assistance,
CPLLC
 
Elsmar Forum Sponsor
M

MIREGMGR

#3
I think your goal may be (1) to prove equivalence to the old model, and (2) to prove safety and effectiveness.

Maybe proving (1) also proves (2), but I'd think the FDA would want (2) proved from first principles unless it was self-evident.

The test process you propose may prove (1), but a better understanding of what the device does and what the data means is necessary to accept that supposition.

I don't see how it would prove (2).
 

Bev D

Heretical Statistician
Staff member
Super Moderator
#4
without knowing what the device does, I would have to say that recrunching the data would not generally be acceptable to the FDA if it were the sole test. Typically, one should run actual patient trials to ensure that the two devices are not substantially different (the statisticians at the FDA may croak if you say prove "equivalence")
 
Thread starter Similar threads Forum Replies Date
C EU MDR - Annex II 6.1 Pre-clinical and clinical data EU Medical Device Regulations 4
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 8
U Do we need clinical trial data for Class IIa medical device under MDR EU Medical Device Regulations 7
CPhelan Using clinical trial safety data for evidence for CE marking EU Medical Device Regulations 8
CPhelan Do you require MDSAP for CE Marking of a Medical Device or is ISO13485:2016 with clinical data sufficient? CE Marking (Conformité Européene) / CB Scheme 6
M Informational Health Canada begins release of clinical data Medical Device and FDA Regulations and Standards News 0
M Informational USFDA Final Rule – Human Subject Protection; Acceptance of Data From Clinical Investigations for Medical Devices Medical Device and FDA Regulations and Standards News 0
M IVD - Generating clinical data for IVD’s in the EU EU Medical Device Regulations 2
W ISO 14155 - Electronic clinical data systems - Definition Other Medical Device Related Standards 0
Y What Clinical Data is required for Drug Delivery System EU Medical Device Regulations 1
MDD_QNA Clinical data requirements - Changing a Class I Medical Device to Class II EU Medical Device Regulations 5
D Use of clinical data in a 510(k) submission Other US Medical Device Regulations 6
bio_subbu USFDA issues draft Guidance on the Acceptance of Medical Device Clinical Data Other US Medical Device Regulations 1
R FDA's guidance - Evaluation of Sex-Specific Data in Medical Device Clinical Studies Other US Medical Device Regulations 3
B Health Canada Acceptance Policy of Foreign Clinical Trial Data for Devices Canada Medical Device Regulations 2
M Category I Medical Device Clinical Data Requirements EU Medical Device Regulations 3
C Retention of Clinical Trial Data after a Clinical Trial has ended Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 5
C Key Links (Library) to search for updated Clinical Data (Is & IIa) EU Medical Device Regulations 1
E Need Corrective Action for Class III Device with no Clinical Data CE Marking (Conformité Européene) / CB Scheme 9
S Clinical or Non-Clinical Data in 510(k) Submission for a type II Medical Device US Food and Drug Administration (FDA) 1
D Registration of Class IV Device -Clinical Data and ISO Certificate Canada Medical Device Regulations 3
A Collecting Clinical Data using CE Marked Devices EU Medical Device Regulations 4
H 510(k) Filing and Predicate Device Clinical Data 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Technical File as per ANNEX VII and Sample Size for clinical data CE Marking (Conformité Européene) / CB Scheme 12
N Clinical Data for Class 1 Devices? CE Marking (Conformité Européene) / CB Scheme 6
I 510k Approval - Does the FDA look for Clinical Data 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
L Clinical Data Report - Literature Route - MEDDEV guidance (MEDDEV 2.7.1) EU Medical Device Regulations 28
L Clinical Trial and Finished Devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
P GSPRs / Clinical Evaluation EU Medical Device Regulations 3
P IDEs and Clinical Investigators - Advice About Records? Other US Medical Device Regulations 3
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 6
P Clinical Trial Requirement - No equivalent devices in the market CE Marking (Conformité Européene) / CB Scheme 3
M Clinical Decision Support Software Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Regular updates of clinical evaluation report EU Medical Device Regulations 6
M Clinical Benefit of device that only aids in a process for managing or treating disease EU Medical Device Regulations 2
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
M Clinical Evaluation Benchmark vs. Equivalent EU Medical Device Regulations 2
D Classification of product for clinical trials EU Medical Device Regulations 14
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5
S Australia TGA Clinical Evaluation Report (CER) Other Medical Device Regulations World-Wide 0
L Clinical research - Request from ANVISA ISO 14971 - Medical Device Risk Management 1
E Use of OQ cleared batch for stability and clinical study use ISO 13485:2016 - Medical Device Quality Management Systems 0
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
M Clinical Development Plan - Advice Requested EU Medical Device Regulations 11
T Clinical evaluation of a new medical device EU Medical Device Regulations 0
dgrainger Informational MDCG 2020:13 - Clinical evaluation assessment report template EU Medical Device Regulations 0

Similar threads

Top Bottom