You are too good for MDR
The MDR is almost intentionally ambiguous.
The following is the first paragraph from Annex XIV, PART B ,POST-MARKET CLINICAL FOLLOW-UP
"5. PMCF shall be understood to be a continuous process that updates the clinical evaluation referred to in Article 61 and Part A of this Annex and shall be addressed in the manufacturer's post-market surveillance plan. When conducting PMCF, the manufacturer shall proactively collect and evaluate
clinical data from the use in or on humans of a device which bears the CE marking and is placed on the market or put into service within its intended purpose as referred to in the relevant conformity assessment procedure, with the aim of confirming the safety and performance throughout the expected lifetime of the device, of ensuring the continued acceptability of identified risks and of detecting emerging risks on the basis of factual evidence."
The clinical evaluation is being updated with the results of clinical use of a device bearing the CE mark with the same intended use.
The phrase "clinical data" as defined in Article 2 (48) refers to clinical investigations and other published reports and relevant information.
To make intelligent distinction (i.e. making sense out of nonsense) we treat PMCF Plan as referring to clinical investigation plans whereas the clinical evaluation plan is the upper layer that includes both data from litrature review and the clinical trials.