Clinical Evaluation Reports per the new MDD 93/42/EEC amendment

Elsmar Forum Sponsor

pkost

Trusted Information Resource
#32
That is what we use and it has been reviewed and accepted by our notified body...we claim compliance with the MEDDEV.

The meddev additionally requires that the results of the evaluation feed back into the risk assessment - I don't include that in the actual evaluation but separately as part of our risk management process - you mustn't forget to do that as well!
 
R

Robbie1

#33
That sounds encouraging
I have plenty of clinical study data ie case studies etc
Do you think putting this into a folder with the segments you suggested would be ok
Or do you have to actually write a new report based on the available data
 

pkost

Trusted Information Resource
#34
Although the studies would go a long way, you do have to critically evaluate the data and review it as a whole.

I'm guessing that these are your own case studies, you should also include other data from other sources, similar products, materials etc. You have to justify why you have excluded data sources. The watch words are fair and balanced, this is unlikely to be achieved from your own internal/marketing reports.
 
D

drewsky1

#35
We have been doing Clinical Evidence Reports for our medical devices and are about to update since they have not been done in 2+ years. We are going to discontinue some of these devices in the EU and Australia and the CE mark will be removed. Does anyone know if there are any other market requirements to do CER's besides required by the MDD? I don't believe this is required by FDA or any other international markets. We will continue to do regular PMS's. Thanks.
 

lilybef

Involved In Discussions
#36
For our class 1 sterile device, we have had to provide a Clinical Evaluation for Singapore, UAE, Mexico, Brazil, and Argentina so far apart from the Technical File for the EU. We have not had any requests from other countries, but we are still registering the product in the major markets. FDA does not require one at the moment.

Hope this helps!:cfingers:
 
D

drewsky1

#37
Thanks. Do they require an update of the CER at different time periods? Or were these just for the initial registration/dossier?
 

lilybef

Involved In Discussions
#38
It was just for the initial submission. We have not had the product long enough to do any reregistrations, so I am not sure if a new report would be required. If I find out anything, I will let you know!:bigwave:
 
D

drewsky1

#39
Thanks, I would appreciate that. I am also going to investigate our international colleagues and will let you know if they can offer any insight as well
 
D

drewsky1

#40
In an NB audit for one of our Technical Files, the reviewer noted:

Please submit clinical evaluation that complies with Annex X. It is recommended referring to MEDDEV 2.7.1 Rev 3 (2009)

The Clinical Evidence Report did not note Annex X as a guidline for the report, just MEDDEV 2.7.1 Rev 3 (2009). The later was considered to be the current guideline. Can anyone shed any light on why the reviewer may have noted this for Annex X?
 
Thread starter Similar threads Forum Replies Date
M CER (Clinical Evaluation Reports) updates - Product Codes EU Medical Device Regulations 2
L Clarification on Clinical Evaluations and Clinical Evaluation Reports Other Medical Device and Orthopedic Related Topics 5
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 1
P GSPRs / Clinical Evaluation EU Medical Device Regulations 3
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 6
S Regular updates of clinical evaluation report EU Medical Device Regulations 6
M Clinical Evaluation Benchmark vs. Equivalent EU Medical Device Regulations 2
S Australia TGA Clinical Evaluation Report (CER) Other Medical Device Regulations World-Wide 0
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
T Clinical evaluation of a new medical device EU Medical Device Regulations 0
dgrainger Informational MDCG 2020:13 - Clinical evaluation assessment report template EU Medical Device Regulations 0
S A clinical performance evaluation study with an IVD product as Investagional Use product - Clinical Monitor requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 8
M Clinical Evaluation Plan vs. PMCF Plan EU Medical Device Regulations 24
M Can someone share a scrubbed version of Clinical Evaluation Plan (CEP) EU Medical Device Regulations 8
S EU MDR Annex XIV - Clinical Evaluation Plan - What do these methods mean? EU Medical Device Regulations 16
S Clinical Evaluation - Is this an ISO 13485:2016 requirement? ISO 13485:2016 - Medical Device Quality Management Systems 4
M Informational How to perform a clinical evaluation of medical devices – Part 2 – Level of clinical evidence and what sufficient clinical evidence means Medical Device and FDA Regulations and Standards News 9
M Informational How to perform a clinical evaluation of medical devices – Part 1 – Overview and sample of activities Medical Device and FDA Regulations and Standards News 0
T EU MDR Article 61- Clinical Evaluation EU Medical Device Regulations 2
pashah Looking for Clinical Evaluation SOP acc. MEDDEV and EU MDR Other Medical Device Related Standards 1
D Clinical Evaluation Report - Consultant Recommendations Consultants and Consulting 9
A Medical device CER (clinical evaluation report) training/seminar services EU Medical Device Regulations 2
M Informational IMDRF – Proposed update to Clinical Evaluation documents Medical Device and FDA Regulations and Standards News 0
M Informational MDCG 2019-3 Interpretation of Article 54(2)b – Pre- market clinical evaluation consultation procedure with the involvement of expert panels Medical Device and FDA Regulations and Standards News 0
TheMightyWife Clinical Evaluation MEDDEV 2.7.1 Rev 4 demonstration of equivalence EU Medical Device Regulations 6
C How frequent to conduct Clinical / Performance Evaluation? CE Marking (Conformité Européene) / CB Scheme 4
H Addressing of Undesirable side-effects, harms, risks and side-effects in clinical evaluation report (CER) EU Medical Device Regulations 12
W Literature used in the clinical evaluation ISO 13485:2016 - Medical Device Quality Management Systems 4
S Medical device CER (clinical evaluation report) according to MDR EU Medical Device Regulations 24
M SOPs, Plans and Report templates for Clinical Evaluation of EU MDR Other Medical Device Related Standards 11
M Medical Device News CDRH Research Programs – VICTRE: Virtual Imaging Clinical Trials for Regulatory Evaluation Medical Device and FDA Regulations and Standards News 0
S Clinical evaluation for a class IIa medical device - EU MDR Requirements EU Medical Device Regulations 3
TheMightyWife Copyright for articles used in CER (Clinical Evaluation Report) EU Medical Device Regulations 16
Q Evidence of precautions (clinical evaluation report, risk management report) EU Medical Device Regulations 6
Rincewind Clinical Evaluation of very low risk medical devices EU Medical Device Regulations 3
SK13485 Can a literature alone be used for the purpose of Clinical Evaluation of a device? EU Medical Device Regulations 6
R Scope and plan of the clinical Evaluation CE Marking (Conformité Européene) / CB Scheme 3
M MDR Clinical Evaluation & investigation (SUMMARY of changes) EU Medical Device Regulations 2
R Clinical evaluation Meddev (2.7.1 revision 4) - Example or Template EU Medical Device Regulations 4
M Clinical Evaluation - MEDDEV 2.7/1 rev4 - Literature route Other Medical Device Related Standards 2
C Qualification of Evaluator - Clinical Evaluation Report MEDDEV 2.7.1 Rev 04 CE Marking (Conformité Européene) / CB Scheme 3
Q Sample of the MEDDEV 2.7.1 rev 4 Clinical Evaluation Report wanted Document Control Systems, Procedures, Forms and Templates 18
Sam Lazzara Clinical Evaluation MEDDEV Guidelines SmackDown - 2017 EU Medical Device Regulations 3
L Clinical Evaluation Report Updating Procedure Other Medical Device Related Standards 5
R Qualification of Evaluator - Clinical Evaluation Report MEDDEV 2.7.1 Rev 04 Other Medical Device and Orthopedic Related Topics 1
E CER (Clinical Evaluation Report) MEDDEV Version 4 Template wanted CE Marking (Conformité Européene) / CB Scheme 5
R Clinical Evaluation according to MEDDEV 2.7.1 (Rev 4) - seeking template ISO 13485:2016 - Medical Device Quality Management Systems 6
Z Labeling Requirements for Free Samples supporting Clinical Evaluation EU Medical Device Regulations 2