Clinical Evaluation using the Literature Method - Example or information wanted
S
SK Vision
#1
Does anyone out there have a template or example of how a Clinical Evaluation using the Literature method might be formatted?
Thanks!
Sarah
Elsmar Forum Sponsor
E
E. Madsen - 2012
#3
SteveK,
How did your audit go? Do you have a more complete example. I have attached my literature review report example. Please provide feedback.
I have a question related to the literature review. We recognize the requirements for literature review. One of those requirements is to evaluate the authors’ background and expertise in relation to the device and/or procedures involved (NB-MED 2.7 section 4.2.1.iv.). It has been challenging for us to obtain author information on some of the articles we have reviewed, especially for those authors that are out of the country.
For those articles that we are unable to find sufficient author information, we were planning on using this statement in our report/technical file.
“The background of the author(s) of this article could not be established. The wait of this article in our overall review is less based on our lack of information."
Thank-you!!
Emily Madsen
S
SteveK
#4
Hi Emily,
Thanks for the interest, the audit went well – a few minor NCs (would have been surprised if none in a level 3), but all corrected to auditor’s satisfaction. With respect to the clinical evaluation, he made an observation (OFI) that some of the references I used did not show clearly enough the connection to the device in question. He was happy with the approach however. Your more detailed analysis (as you indicate) is based on NB-MED 2.7 – Clinical Investigations, Clinical Evaluation – i.e. in my eyes this means citing ‘clinical trials’ literature, totally different animal (see BS EN ISO 14155-1) – thus the authorship etc requirements – not a good idea. I have used MEDDEV 2.7.1 for my approach – I think you should have a look at this instead. So anyway, post audit, what I have done now is that for each reference cited, I have included either an abstract (all I could find in a number of cases) or a short precise of the article/book chapter – with pictures, diagrams etc as these often clearly show design or use equivalence. This document is therefore much more complete as you request, but has become a bit too specific and confidential to easily share a completed example.
All this said, I like your citation hits and applicability, because I do something very similar as I monitor alerts, recalls, advisory notices (FDA, MHRA, TGA, MDB, BfArM, Afssaps etc) as part of my post market surveillance.
For clinical evidence definitions, if not already done so, check out SG5/N1R8:2007 and SG5/N2R8:2007 at the GHTF website
Hope this helps.
Steve
E
E. Madsen - 2012
#5
Hi Emily,
Steve
Thanks for the information! Glad the audit went well. We have been using both NB-MED 2.7 and MEDDEV 2.7.1, but relied more heavily on MEDDEV. As for your comment regarding the evaluation of the authors, the first bullet-point under section 4.3.1.(i).e. states that "the relevance of the author's background and expertise in relation to the particular device and/or medical procedure involved" should be addressed. If we don't identify each author's backgrounds then how can we assess the articles properly.
What do you think? I appreciate your comments!!
Emily
S
SteveK
#6
Emily,
Where there are published papers referencing clinical trials information, these will have had a peer group and/or editorial review; otherwise they would not be published. Therefore, would not the author’s competency have been checked out by default? Having said this, think about falsified drug trial data reported over the years! I think it unlikely that you would be able to find unpublished primary data. Anyway, I think using this document (NB-MED 2.7) as a guide is OTT for class II devices. If your products are class III, then I suppose it would be a different matter. Would you not need to be a medic (which you may well be) to go into the required review details anyway?
Steve
S
schlupp
#7
Hi Emily and Steve,
I am Johannes from germany and we are just in this theme you are describing. We have still to do the literature route to our class I products, stehoscopes. Did your example work this way?
kind regards
Johannes
S
SteveK
#8
Hi Johannes,
I would expect there would be a large amount of references to stethoscopes out there on the web. When you do find something similar referenced, note the source and include a relevant extract or review from the journal, book, alert etc on your clinical evaluation document/form/report (I often use screen capture e.g. for pictures or diagrams of items similar to our own). You may only find an abstract due to internet sites requiring payment for a full article (try Google Books as one source of literature as well as my previous suggestions). Is this the sort of information you require – not totally clear from your post?
Steve
PS welcome to the community!
S
schlupp
#9
Hi Steve,
thanks for your quick answer. I will start searching for the correct google books. The PDF you posted some time ago as an example to the "literature route" was the search word in google, I found this forum. Did the build-up of your pdf work to pass the authorities?
Johannes
S
SteveK
#10
Hi Johannes,
As I indicated, the auditor was fine with the (pdf) form structure. It was just an observation (OFI) that originally there was just the reference quoted in my clinical evaluation form, but no actual detail e.g. an extract. Often the auditors just have a particular opinion. If you were writing a technical paper for publication you would tend to just cite references as supporting evidence - this is what I was used to.
Steve
Thread starter Similar threads Forum Replies Date
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 2
P GSPRs / Clinical Evaluation EU Medical Device Regulations 3
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 6
S Regular updates of clinical evaluation report EU Medical Device Regulations 6
M Clinical Evaluation Benchmark vs. Equivalent EU Medical Device Regulations 2
S Australia TGA Clinical Evaluation Report (CER) Other Medical Device Regulations World-Wide 0
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
T Clinical evaluation of a new medical device EU Medical Device Regulations 0
dgrainger Informational MDCG 2020:13 - Clinical evaluation assessment report template EU Medical Device Regulations 0
S A clinical performance evaluation study with an IVD product as Investagional Use product - Clinical Monitor requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 8
M Clinical Evaluation Plan vs. PMCF Plan EU Medical Device Regulations 24
M Can someone share a scrubbed version of Clinical Evaluation Plan (CEP) EU Medical Device Regulations 8
S EU MDR Annex XIV - Clinical Evaluation Plan - What do these methods mean? EU Medical Device Regulations 16
S Clinical Evaluation - Is this an ISO 13485:2016 requirement? ISO 13485:2016 - Medical Device Quality Management Systems 4
M Informational How to perform a clinical evaluation of medical devices – Part 2 – Level of clinical evidence and what sufficient clinical evidence means Medical Device and FDA Regulations and Standards News 9
M Informational How to perform a clinical evaluation of medical devices – Part 1 – Overview and sample of activities Medical Device and FDA Regulations and Standards News 0
T EU MDR Article 61- Clinical Evaluation EU Medical Device Regulations 2
pashah Looking for Clinical Evaluation SOP acc. MEDDEV and EU MDR Other Medical Device Related Standards 1
D Clinical Evaluation Report - Consultant Recommendations Consultants and Consulting 9
M CER (Clinical Evaluation Reports) updates - Product Codes EU Medical Device Regulations 2
A Medical device CER (clinical evaluation report) training/seminar services EU Medical Device Regulations 2
M Informational IMDRF – Proposed update to Clinical Evaluation documents Medical Device and FDA Regulations and Standards News 0
M Informational MDCG 2019-3 Interpretation of Article 54(2)b – Pre- market clinical evaluation consultation procedure with the involvement of expert panels Medical Device and FDA Regulations and Standards News 0
TheMightyWife Clinical Evaluation MEDDEV 2.7.1 Rev 4 demonstration of equivalence EU Medical Device Regulations 6
C How frequent to conduct Clinical / Performance Evaluation? CE Marking (Conformité Européene) / CB Scheme 4
H Addressing of Undesirable side-effects, harms, risks and side-effects in clinical evaluation report (CER) EU Medical Device Regulations 12
W Literature used in the clinical evaluation ISO 13485:2016 - Medical Device Quality Management Systems 4
S Medical device CER (clinical evaluation report) according to MDR EU Medical Device Regulations 24
M SOPs, Plans and Report templates for Clinical Evaluation of EU MDR Other Medical Device Related Standards 11
M Medical Device News CDRH Research Programs – VICTRE: Virtual Imaging Clinical Trials for Regulatory Evaluation Medical Device and FDA Regulations and Standards News 0
S Clinical evaluation for a class IIa medical device - EU MDR Requirements EU Medical Device Regulations 3
TheMightyWife Copyright for articles used in CER (Clinical Evaluation Report) EU Medical Device Regulations 16
Q Evidence of precautions (clinical evaluation report, risk management report) EU Medical Device Regulations 6
Rincewind Clinical Evaluation of very low risk medical devices EU Medical Device Regulations 3
SK13485 Can a literature alone be used for the purpose of Clinical Evaluation of a device? EU Medical Device Regulations 6
R Scope and plan of the clinical Evaluation CE Marking (Conformité Européene) / CB Scheme 3
M MDR Clinical Evaluation & investigation (SUMMARY of changes) EU Medical Device Regulations 2
R Clinical evaluation Meddev (2.7.1 revision 4) - Example or Template EU Medical Device Regulations 4
M Clinical Evaluation - MEDDEV 2.7/1 rev4 - Literature route Other Medical Device Related Standards 2
C Qualification of Evaluator - Clinical Evaluation Report MEDDEV 2.7.1 Rev 04 CE Marking (Conformité Européene) / CB Scheme 3
Q Sample of the MEDDEV 2.7.1 rev 4 Clinical Evaluation Report wanted Document Control Systems, Procedures, Forms and Templates 18
Sam Lazzara Clinical Evaluation MEDDEV Guidelines SmackDown - 2017 EU Medical Device Regulations 3
L Clinical Evaluation Report Updating Procedure Other Medical Device Related Standards 5
R Qualification of Evaluator - Clinical Evaluation Report MEDDEV 2.7.1 Rev 04 Other Medical Device and Orthopedic Related Topics 1
E CER (Clinical Evaluation Report) MEDDEV Version 4 Template wanted CE Marking (Conformité Européene) / CB Scheme 5
R Clinical Evaluation according to MEDDEV 2.7.1 (Rev 4) - seeking template ISO 13485:2016 - Medical Device Quality Management Systems 6
Z Labeling Requirements for Free Samples supporting Clinical Evaluation EU Medical Device Regulations 2
Gamula CFDA Clinical Evaluation Report (China) China Medical Device Regulations 4
Similar threads