SBS - The best value in QMS software

Clinical investigations of medical devices - not for CE marking

Michael Ayers

Starting to get Involved
#1
Article 62 of the regulations list the applicable articles to follow for clinical investigations for CE marking purposes
Clinical investigations shall be designed, authorised, conducted, recorded and reported in accordance with the provisions of this Article and of Articles 63 to 80, the acts adopted pursuant to Article 81, and Annex XV, where carried out as part of the clinical evaluation for conformity assessment purposes

For clinical investigation of non CE marked devices and not for the purposes of CE marking (e.g. pure research), are the requirements under the regulations covered by this part of Article 82?
Clinical investigations, not performed pursuant to any of the purposes listed in Article 62(1), shall comply with the provisions of Article 62 (2) and (3), points (b), (c), (d), (f), (h), and (l) of Article 62(4) and Article 62(6).
Are there any other requirements in the regulations for these devices and these studies?
 
Elsmar Forum Sponsor
Thread starter Similar threads Forum Replies Date
dgrainger Informational MHRA updated guidance: Clinical investigations of medical devices – guidance for manufacturers Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Draft Guidance – Patient Engagement in Design and Conduct of Medical Device Clinical Investigations Medical Device and FDA Regulations and Standards News 0
M Informational MHRA guidance updated to reflect the mDR – Clinical investigations of medical devices – Biological safety assessment Medical Device and FDA Regulations and Standards News 1
M Informational USFDA Final Rule – Human Subject Protection; Acceptance of Data From Clinical Investigations for Medical Devices Medical Device and FDA Regulations and Standards News 0
M Informational US FDA draft guidance. – Clinical Investigations for Prostate Tissue Ablation Devices Medical Device and FDA Regulations and Standards News 2
L Clinical Investigations under MDR EU Medical Device Regulations 4
M Informational USFDA draft guidance – A Risk-Based Approach to Monitoring of Clinical Investigations Questions and Answers Guidance for Industry Medical Device and FDA Regulations and Standards News 0
L EU MDR CRO (Contract Research Organization) requirements for Clinical Investigations EU Medical Device Regulations 1
V Use of e-records and e-Signatures in Clinical Investigations Under 21 CFR part 11 Q&A Qualification and Validation (including 21 CFR Part 11) 1
R Final guidance - FDA Decisions for IDE Clinical Investigations Other US Medical Device Regulations 1
N New MEDDEV 2.7/3 Guideline: Clinical Investigations: Serious Adverse Event Reporting EU Medical Device Regulations 4
bio_subbu New MEDDEV Guidance Documents - Clinical investigations and SAE Reporting EU Medical Device Regulations 3
R CRO (Contract Research Organization) Requirements for MDD Clinical Investigations EU Medical Device Regulations 5
T Application of ISO 14155 Clinical Trials Investigations EU Medical Device Regulations 5
P GSPRs / Clinical Evaluation EU Medical Device Regulations 2
P IDEs and Clinical Investigators - Advice About Records? Other US Medical Device Regulations 3
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 5
P Clinical Trial Requirement - No equivalent devices in the market CE Marking (Conformité Européene) / CB Scheme 3
M Clinical Decision Support Software Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Regular updates of clinical evaluation report EU Medical Device Regulations 6
M Clinical Benefit of device that only aids in a process for managing or treating disease EU Medical Device Regulations 2
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
C EU MDR - Annex II 6.1 Pre-clinical and clinical data EU Medical Device Regulations 4
M Clinical Evaluation Benchmark vs. Equivalent EU Medical Device Regulations 2
D Classification of product for clinical trials EU Medical Device Regulations 14
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5
S Australia TGA Clinical Evaluation Report (CER) Other Medical Device Regulations World-Wide 0
L Clinical research - Request from ANVISA ISO 14971 - Medical Device Risk Management 1
E Use of OQ cleared batch for stability and clinical study use ISO 13485:2016 - Medical Device Quality Management Systems 0
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
M Clinical Development Plan - Advice Requested EU Medical Device Regulations 11
T Clinical evaluation of a new medical device EU Medical Device Regulations 0
dgrainger Informational MDCG 2020:13 - Clinical evaluation assessment report template EU Medical Device Regulations 0
shimonv Minimum 60601 requirements for a clinical trial IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S A clinical performance evaluation study with an IVD product as Investagional Use product - Clinical Monitor requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
J Need for a cleanroom in the manufacture of a medical device for a clinical trial EU Medical Device Regulations 4
R Impact of changing manufacturer post clinical investigation CE Marking (Conformité Européene) / CB Scheme 3
C MDCG 2020-06 Clinical evidence legacy devices EU Medical Device Regulations 3
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 6
M Clinical Evaluation Plan vs. PMCF Plan EU Medical Device Regulations 21
S Selling Clinical Practice Advice Other US Medical Device Regulations 7
S FDA's E submitter - For clinical electronic thermometer Medical Device and FDA Regulations and Standards News 6
M Good Clinical Practice (GCP) - ISO 14155 vs. ICH Other ISO and International Standards and European Regulations 10
S Article 61 - MDR Clinical Investigation Requirement EU Medical Device Regulations 5
J Are DAM tools, Mobile application, clinical dashboard, etc. considered accessories to medical device or supportive functions? Manufacturing and Related Processes 4
M Can someone share a scrubbed version of Clinical Evaluation Plan (CEP) EU Medical Device Regulations 5
R Do Medical Device Complaints include Clinical Trials? Medical Device and FDA Regulations and Standards News 3
S EU MDR Annex XIV - Clinical Evaluation Plan - What do these methods mean? EU Medical Device Regulations 16

Similar threads

Top Bottom