CMM - Calibration Frequency and Variation
L
Leestrath
Happy New Year Everyone,...:),
I am currently working on CMM calibration frequency and variation.
Problem definition: We have morethan 25 CMM's in our manufacturing site, and it costs us lot of money to calibrate the CMM's. We are targeting 12 or 18-month calibration, current calibration frequency is less than 9months and in many cases it is less than 6 months....
There are recurring problems with the CMM repeatability, reproducibility and between machine variations in the manufacturing cells.
Example: I inspected the same component on different CMM's and the results varies between 3 to 6 microns within the machine and 14 to 24 microns between the machines,.... .
After these findings, now I am looking into
1. The compressed air quality supplied to CMM operation,
2. The influence of temperature (it varies between 23 to 27 degree Celsius),
3. The interface error between the OEM and Renishaw head,
4. Quality of Air filters used
5. Daily maintenance etc...
Has anyone come across this sort of situation before? Please share your views and experiences
Have A Great Year Ahead!
Cheers thanks,...
Ron Rompen
Depending on the tolerances you need to hold, 3 to 6 microns variation within 1 machine is not (in my opinion) excessive. In general, unless you are using a very high-end machine, this is approaching the limits of the ability of the equipment to measure.
Air pressure can have some effect on the equipment, but ambient temperature, vibration, part cleanliness and airflow will (probably) have a much bigger effect. Thermal compensation is available on some machines (I know Zeiss has it on theirs) and may be worth investing in if your manufacturing processes demand it.
Another thought, to reduce your costs, is to perform internal calibration of your equipment with a known standard, calling in an outside source only when the equipment approaches the limits of function. Most equipment manufacturers will provide the training necessary as well as the CMM standards you require to perform the calibration.
Richard Pike
I had a similar issue with a recent client. Assuming you can live with the error of 6 micron - then consider running a Stability Chart per station. This worked for us and reduced our calibration costs by almost 80%.. i.e. we only did anything to the gauges (other than routine care) when the Stability Chart indicated assignable variation.
L
Leestrath
Thanks Richard,...
It would be great if you could explain a little bit and share the approach/method followed by you to establish stability chart.
thank you
Z CMM (Coordinate Measuring Machine) Calibration Interval (Frequency) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
R CMM Calibration Frequency - Bendix manual Coordinate Measuring Machine Calibration Frequency (Interval) 5
D Calibration Frequency of CMM as per ISO/IEC 17025 ISO 17025 related Discussions 11
Ajit Basrur Reduction of calibration frequency for CMM (coordinate measuring machine) General Measurement Device and Calibration Topics 9
K Calibration Frequency of a CMM Machine (Coordinate Measuring Machine) Calibration Frequency (Interval) 1
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
J Manual Vision CMM Calibration General Measurement Device and Calibration Topics 7
K "Verification of Calibration" - CMM arm General Measurement Device and Calibration Topics 6
S Micrometer Calibration Using a CMM Calibration and Metrology Software and Hardware 12
J CMM calibration using a perfect sphere that is supplied Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
B Can I use a Laboratory CMM for Gauge Calibration General Measurement Device and Calibration Topics 11
B CMM Stylus Calibration - Required to record the As Found condition AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
R CMM Rotary Axis Calibration General Measurement Device and Calibration Topics 4
R Calibration of a portable FARO gage portable CMM # F04L2 gauge General Measurement Device and Calibration Topics 6
M CMM (coordinate measuring machine) calibration recommendations General Measurement Device and Calibration Topics 15
P Are we required to register to REACH? Small CMM service and calibration company RoHS, REACH, ELV, IMDS and Restricted Substances 3
M Calibration of ring gages with a CMM General Measurement Device and Calibration Topics 3
J Can I use the CMM after the Calibration has expired? Calibration Frequency (Interval) 18
V "Calibration" or "Validation with CMM machine" General Measurement Device and Calibration Topics 3
J Large CMM Calibration - Tetronom or Giant Ball Bar General Measurement Device and Calibration Topics 6
B Calibration of "Z" class ring gauges internally using a CMM General Measurement Device and Calibration Topics 5
K CMM Qualifying Sphere Calibration General Measurement Device and Calibration Topics 9
B Calibration Data Packages - CMM - Any way around "As found data?" General Measurement Device and Calibration Topics 17
G Gage Calibration - Using a CMM and Gage Blocks for In-House Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
L Understanding Calibration Results on CMM Reports Statistical Analysis Tools, Techniques and SPC 2
L Understanding calibration results on CMM reports General Measurement Device and Calibration Topics 4
K Dimensional Lab (CMM Room) vs. Physical Test/Calibration Lab - Same required temp/hum General Measurement Device and Calibration Topics 2
Q CMM Calibration Certification - Small CMM Calibration Service General Measurement Device and Calibration Topics 5
Nihls MSA Study Type 1 (CMM) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C CMM Granite too hard General Measurement Device and Calibration Topics 4
A Cg, Cgk and GRR of a CMM Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
Marco Bernardi MMC & LMC modifiers and CMM measuring techniques like diameter least squares and circularity or minimum/ maximum diameter Calibration and Metrology Software and Hardware 5
M Honeywell CMM's - Why Honeywell will no longer sell their manuals Federal Aviation Administration (FAA) Standards and Requirements 2
Marco Bernardi Plane measurement with a 3D touch machine like a CMM. Calibration and Metrology Software and Hardware 3
Marco Bernardi Diameters measurement with 3D touch machine like a CMM Calibration and Metrology Software and Hardware 4
A Gage type 1 study on CMM Capability, Accuracy and Stability - Processes, Machines, etc. 2
V Recording CMM dimensional results on CFG-1003 APQP and PPAP 3
G Measuring customer parts on a CMM - How many decimals to report to the customer? ISO 17025 related Discussions 28
J CMM Accuracy Formula General Measurement Device and Calibration Topics 3
D CMM Room cleanliness requirements ISO 17025 related Discussions 7
P IATF 16949 MSA Studies for CMM machines IATF 16949 - Automotive Quality Systems Standard 3
D Can a Mill be used as a CMM? General Measurement Device and Calibration Topics 2
Ron Rompen Best Practices in CMM Correlation Studies General Measurement Device and Calibration Topics 3
B Zeiss CMM Calypso output to Excel or CSV file Help Inspection, Prints (Drawings), Testing, Sampling and Related Topics 17
J CMM Probe Certification General Measurement Device and Calibration Topics 1
S Opinions on the "Best" CMM for High Accuracy Machining General Measurement Device and Calibration Topics 9
P Getting ready for Boeing's AC7130 requirement for suppliers using a CMM AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3