Color measurements R&R: what's wrong with white?

L

lexaclev

Dear members,

This is my first post so first of all thank you all for valuable and helpful threads and advices in MSA topic. Most of my questions are already answered here, but I still need you collective wisdom in the field of instrumental color measurements.

The problem we ran into is GR&R for instrumental (spectrophotometer) measurements of white colored painted plastic auto parts. GRR for pure whites is almost twice as bad as for other colors and unacceptable for practical purposes. Measured values are dL, da, db and resulting deltaE. The troublesome component is dL with most between-appraisers variation. Same parts, same appraisers, same experimental plan as for other colors is used. GRR for all other colors is acceptable.

What might be wrong with white colors? Is instrumental control unreliable for them and visual inspection should be used instead? (BTW, customer requires instrumental control)

Thank you in advance for your help!
 
L

lexaclev

Are you certain that the wavelength and geometry are identical for each instrument?

Yes, the instrument was also the same without settings changes, ambient light as well. The whole experiment was well controlled and was repeated second time for another set of parts.

GR&R was calculated as %Tolerance and we use same tolerances for all colors: deltaE < 1,7
 

Ron Rompen

Trusted Information Resource
You are measuring the reflected color of the 'white light' which is in fact, made up of all other colors. This will skew your observations, and unless you measure the exact same spot, at exactly the same orientation, with the same temp and humidity (etc etc ad naseum) then you will get different results each time. The degree of difference may vary, but it will always be there.
If this is a gloss white (vs flat or eggshell) then the effect is even worse.
 

julsbear

Involved In Discussions
For whites almost all your variation is going to be in L and it will also vary quite a bit more than in the other achromatic colors (grays and black) because the absolute value is so high.

Placement of part on the port may also be key. Are the parts opaque? Even some light due to lack of opacity could have a big affect especially if the operator keeps their hand on the plunger (if a desktop unit) or where they stand (shadow).
 
Last edited:

Hershal

Metrologist-Auditor
Trusted Information Resource
If you want to do GR&R, ignore what may be called light, and go solely by wavelength. No other approach is valid.

The reason is because if with sphere, is the blue coming on stronger? Chips inside? If on a Goneo, then distance?

Wavelength is the only constant as both instruments run through a Spectrophotometer.

Hope this helps.
 
Thread starter Similar threads Forum Replies Date
S Labelling Questions (use of color and symbols) US Medical Device Regulations 1
Z Change color or shape of individual data point in control chart Using Minitab Software 6
D Checking fixture color coding General Measurement Device and Calibration Topics 2
J SSQR-01 Vision Requirements- Color Blindness AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
Marc Visual Inspection - Color Vision Coffee Break and Water Cooler Discussions 15
D IMDS for Like Components (Interior Plastic, Leather/vinyl Armrest, etc.)other than Color APQP and PPAP 1
Dobby1979 Requirements for the dimensions / color of medical device labels - Software as a Medical Device IEC 62304 - Medical Device Software Life Cycle Processes 2
M Informational Work in progress at the FDA for biological evaluation – Color Hazard and RISk calculator (CHRIS) Medical Device and FDA Regulations and Standards News 0
S Device name change and label color change EU requirement? Other Medical Device Regulations World-Wide 6
erozas2003 Color test on molded plastic parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
S Testing (Evaluation of) and Verification of Employee Color Vision Human Factors and Ergonomics in Engineering 19
C ISO 7010 m002 mandatory action - Color requirements on Packaging IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Paint Color Standards Manufacturing and Related Processes 5
V Query on PA66+GF15% - Color variation (Natural to Yellowish) problem Manufacturing and Related Processes 20
K Color to be used for the Packing box Symbols Other Medical Device Related Standards 2
R Paint Color Control Manufacturing and Related Processes 6
S Organisms postive to Wellcolex? Color Salmonella Test Kit US Food and Drug Administration (FDA) 1
S Labeling Symbols for EU - E.g. EN980 - Dimension, color, font, etc. Other ISO and International Standards and European Regulations 6
optomist1 Error Mistake Proofing Color Parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
B ISO 3864-4 Symbol Color Regulations Other ISO and International Standards and European Regulations 4
M Molded Component Color Tolerance Specification Manufacturing and Related Processes 6
A Reporting of Design Changes (Color Change Only) EU Medical Device Regulations 5
K Black Anodize Color Specification per MIL Spec Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
R AC 7120 Vision and Color Blindness Testing AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
somashekar How do you specify color tolerance ... Manufacturing and Related Processes 6
QMMike Color Measurement (Monitoring) during Processes Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
K How do I autofill cell color based on other cells' color in Excel? Excel .xls Spreadsheet Templates and Tools 11
Steve Prevette Color Coded SPC Dashboards The Reading Room 0
D MSA (Measurement Systems Analysis) on Visual Color Check IATF 16949 - Automotive Quality Systems Standard 13
W Approximation of a Gardner color value to the APHA Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
S Color Coded Shirt System on our Shop Floor Lean in Manufacturing and Service Industries 18
T Enclosure Flammability Rating of Molded-In Custom Color Liquid Concentrates IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R 5S Red Tag and ISO Procedure Red Tag - Tag Color Conflict Document Control Systems, Procedures, Forms and Templates 7
1 What device to use to measure LCD display devices color/brightness General Measurement Device and Calibration Topics 3
B Calibration Interval of Munsell Color Charts General Measurement Device and Calibration Topics 3
J Master Color Plaque recommended sources Quality Tools, Improvement and Analysis 16
B Color Individual Points on a Control Chart? Using Minitab Software 7
G Organoleptic (Color, Odor, Taste, Texture, etc) lexicon? General Measurement Device and Calibration Topics 8
V Color Assessment Cabinet - Calibration IATF 16949 - Automotive Quality Systems Standard 1
P Color Cross Reference - PMS and NCS colors samples Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
D Need suggestions for a mobile Spectrophotometer (a.k.a. color meter) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
Proud Liberal Color Evaluation of Tinted Transparent Plastic Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
CalRich How to display color chips - Injection molding color verification Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
Ajit Basrur Color Codes for different gases Manufacturing and Related Processes 2
S Phenolic Resin - Colour Variation - Will oxidation change its color? Manufacturing and Related Processes 4
P Color Coding of Tools for 5S - The Visual Workplace Lean in Manufacturing and Service Industries 24
Wes Bucey Big Brother! - Even in your color laser printer After Work and Weekend Discussion Topics 15
M Color Codes and visual floor marking - Are there any Standards? Manufacturing and Related Processes 11
E How to specify Paint or Plastic Color - Pantone or Manufacturer Specification? Manufacturing and Related Processes 15
B TS16949 Clause 7.1 in plastic pellet 'color mixing' and color compounding' process IATF 16949 - Automotive Quality Systems Standard 1

Similar threads

Top Bottom