Hello,
When combining medical equipment (IEC 60601-1) to other equipment (not IEC 60601-1, e.g. the hospital network) to form a medical system, usually there is a need to implement a separation device between that medical equipment and the rest of the equipment outside the medically used room.
However, according to IEC 60601-1, art. 16.5: "the separation device shall have the dielectric strength, creepage distances and air clearances required for one MOOP". There is no mention here of any MOPP requirement. The dielectric strength for one MOOP (230V) is 1.5 kV rms, which is exactly the same as equipment built to IEC 60950-1.
Does this mean that a common "modern" network switch (built to IEC 60950-1) in itself is suitable to be used as a separation device accoring to IEC 60601-1?
I cannot find any detailed information anywhere. Some say it is indeed suitable in itself (not taking into account that the cables from the switch to the network outlet could be damaged for example), others say that the requirements for 2 MOPPs also need to be considered. There are manufacturers who build network switches complying with IEC 60601-1.
I would be happy to hear your thoughts on this!
When combining medical equipment (IEC 60601-1) to other equipment (not IEC 60601-1, e.g. the hospital network) to form a medical system, usually there is a need to implement a separation device between that medical equipment and the rest of the equipment outside the medically used room.
However, according to IEC 60601-1, art. 16.5: "the separation device shall have the dielectric strength, creepage distances and air clearances required for one MOOP". There is no mention here of any MOPP requirement. The dielectric strength for one MOOP (230V) is 1.5 kV rms, which is exactly the same as equipment built to IEC 60950-1.
Does this mean that a common "modern" network switch (built to IEC 60950-1) in itself is suitable to be used as a separation device accoring to IEC 60601-1?
I cannot find any detailed information anywhere. Some say it is indeed suitable in itself (not taking into account that the cables from the switch to the network outlet could be damaged for example), others say that the requirements for 2 MOPPs also need to be considered. There are manufacturers who build network switches complying with IEC 60601-1.
I would be happy to hear your thoughts on this!