Comparing Gauge Blocks - Translating the calibration between the master set and the QA set

G

grnfish

Hey,

I am trying to determine the best course of action for my calibration plans.

I am anticipating getting Grade 0 or 1 for my 'Master Block'. The quantity would be two or three blocks (maybe more). Then use some sort of 'gauge comparator' to calibrate to a QA set Grade 1 or 2 blocks which will be used for calibration. I was also hoping to use this device to translate the 'Master Block' to my sets of XX or XXX gauge pins also.

Does anyone know of an instrument with a good way of translating the calibration between the master set and the QA set? Or is there a better approach all together?

Pratt and Whitney make some instruments such as the Laser Ruler.

Opinions welcomed, cheers
 

dwperron

Trusted Information Resource
When you start dealing with gauge blocks and class XXX pins you are moving into areas of metrology where allowable errors are very small. Also I am assuming here that you are dealing with ASME B89 tolerances, better referred to as Grade AS-1 and AS-2. At 1 inch the maximum deviation for your blocks would be ±6 µin for Grade 0, ± 12 for AS-1 and ±24 for AS-2 which, for gauge blocks, are fairly wide tolerances. For the XXX and XX pins the tolerances are 15 and 30 µin, similar values. You could use a machine like a Pratt and Whitney Labmaster to compare these "master" blocks to your working sets, adding an uncertainty of about 4 µin for every inch of difference between your master block and the block being measured.

Another very large consideration would be temperature stability. Tool steel has an expansion coefficient of about 12 µin/in/°F, so if the 1 inch master block and the block being measured are 2°F apart in temperature there could be a 24 µin difference in the measurement. That's why strong environmental controls are needed for hard gauging, why you need to lets tools "soak" to stabilize at lab temperature before calibration, and why you need gloves and shields to prevent body heat from affecting measurements.

In the metrology world you typically would use a "grand master" gauge block set to compare to sets being calibrated. The one to one size comparison can reduce you measurement uncertainty. There is a lot involved in calibrating gauge block sets, so unless you have a number of sets you should probably send them out to the experts who specialize in this field.
 
G

grnfish

Thank you for the reply

Currently creating a plan (that will span the next few years) to establish a high end metrology laboratory within our company's facility. I do understand the importance of a controlled environment and the small room for error when preforming these procedures, after consulting the NIST gauge block text available on the internet (along with other sources). I have continued researching the appropriate ways of going about this task. I suppose since this is new territory for the company (and me), I have come to the forum with questions I will eventually (hopefully) answer myself -- but always looking for quick opinions and inputs since Google does not always provide resources with a black and white answer (welcome to the real world)

Enjoy
 
Thread starter Similar threads Forum Replies Date
Y Gap assessment comparing RDC ANVISA 665/2022 and RDC ANVISA 16/2013? Other Medical Device Related Standards 0
D Comparing inline Conductivity instrument with offline test results General Measurement Device and Calibration Topics 0
B New to ISO 14971. Comparing to MIL-STD-882 ISO 14971 - Medical Device Risk Management 7
P Comparing Two Test Variables Using Attribute Data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
adztesla Equivalency study - Comparing 2 similar equipments Reliability Analysis - Predictions, Testing and Standards 1
E Comparing process parameters on non-normal batches Capability, Accuracy and Stability - Processes, Machines, etc. 11
S Comparing measurement results for a part from two ring gage General Measurement Device and Calibration Topics 3
R Statistical Methods for comparing test and reference product equivalence for quality attributes US Food and Drug Administration (FDA) 3
O Software for comparing text of PDF files Contract Review Process 2
M Comparing data from destructive testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
A Demonstration of Equivalence - Need for comparing biological characteristics for an SamD EU Medical Device Regulations 1
V Statistical basis and justification while comparing / changing sampling plans Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
N Does FDA allow comparing with the competitors Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
L Procedure for Comparing the Monitor Reading with Laboratory Result Statistical Analysis Tools, Techniques and SPC 4
S Looking for a checklist comparing ISO 13485:2016 and 21 CFR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 14
M Comparing the MMOG (v3 or v4) with ISO / TS16949 Customer and Company Specific Requirements 3
A Comparing US/FDA Drug Regulations against US/FDA Device Regulations US Food and Drug Administration (FDA) 3
E Correct Statistical Test comparing 2 Groups Statistical Analysis Tools, Techniques and SPC 14
R Comparing Laboratory Scope of Accreditation against internal equipment General Measurement Device and Calibration Topics 3
C Comparing Two Independent Group Observations - 95% Confidence Intervals Overlap Statistical Analysis Tools, Techniques and SPC 3
D Comparing the ISO 9001:2008 standard to ISO 17025 ISO 17025 related Discussions 2
S Comparing Quality System Software Features Quality Assurance and Compliance Software Tools and Solutions 18
C Comparing Two Test Methods - Acceptable Difference(s)? Statistical Analysis Tools, Techniques and SPC 3
M pH Calibration - Comparing Electrical and CRM Methods General Measurement Device and Calibration Topics 1
I Six Sigma comparing to other National and International Models Six Sigma 6
A Comparing Effect of Process Parameters on Equipment Performance during Start Ups Statistical Analysis Tools, Techniques and SPC 3
P Acceptable Adjusted R Squared Value Comparing 2 Measurement Devices in Minitab Using Minitab Software 3
C Comparing PPM Year over Year - Determining the PPM decrease from 2009 to 2010 Benchmarking 10
M z Test vs. F Test - The most appropriate for comparing the Variances Statistical Analysis Tools, Techniques and SPC 3
T Comparing Two Measurement Systems in terms of Accuracy and Capability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
R Comparing the Accuracy of Analog vs. Digital Multimeter General Measurement Device and Calibration Topics 6
T Comparing two different Ultrasonic Tester Calibration Techniques General Measurement Device and Calibration Topics 2
Q Internal Audit:Comparing Efficiency of Similar Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
D Comparing Variation - A CMM and an On-Line Inspection Station Statistical Analysis Tools, Techniques and SPC 6
N Statistically Comparing Technician Techniques Reliability Analysis - Predictions, Testing and Standards 16
C Comparing Histograms - Bimodal Distribution? Statistical Analysis Tools, Techniques and SPC 4
K Comparing CQE vs. MSc. Professional Certifications and Degrees 1
J To PPAP or Not? Comparing two machines (one old and one new) APQP and PPAP 6
C Comparing OHSAS 18001 to the new ANSI Z10 - Where I can get information? Occupational Health & Safety Management Standards 1
J Has anyone created a matrix comparing ISO 9001 VS BS 7799 and or ISO 27001? Document Control Systems, Procedures, Forms and Templates 1
A Comparing six sigma & EFQM (European Foundation for Quality Management) Six Sigma 2
C Comparing what I do against Canada request - Immigration Immigration and Expatriate Topics 4
J KAPPA Statistic - Comparing appraisers ratings to a known standard IATF 16949 - Automotive Quality Systems Standard 8
R Comparing MicroVu?s Vertex machine to OPG?s Smart Scope Flash General Measurement Device and Calibration Topics 3
D Comparing ISO 10012 and 17025 ISO 17025 related Discussions 2
S Comparing ISO with PMBOK (Project Management Body of Knowledge) Misc. Quality Assurance and Business Systems Related Topics 2
D Comparing Apples and Apples vs. Apples and Oranges QS-9000 - American Automotive Manufacturers Standard 8
C Gap Analysis comparing QS-9000 and TS 16949 IATF 16949 - Automotive Quality Systems Standard 33
S Gauge R&R for Profilometer Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J Using current gauge to verify masters General Measurement Device and Calibration Topics 1

Similar threads

Top Bottom