Comparing Laboratory Scope of Accreditation against internal equipment

RCW

Quite Involved in Discussions
#1
Sorry if this is a stupid question but I feel REAL clueless on this one. :confused:

Had an auditor in here this week and he said I really SHOULD be verifying that the calibration laboratory I use is accredited to calibrate the test equipment I send them. I already have a copy of their latest Scope of Accreditation but I am confused on what I should be comparing this against. I'm assuming I need to know what my equipment's capability is and then compare this against the cal lab's scope document? To me this looks difficult especially for items such as an oscilloscope.

Can someone point me in a direction or provide some guidance on how to address this?

Thanks!
 
Elsmar Forum Sponsor

Jim Wynne

Leader
Admin
#2
Have a look at the attachment, which is a page from the scope of one of our calibration providers. Because you're using oscilloscopes, I presume that there's someone in your company who's familiar with the technical aspects of their use, and the necessary calibration requirements. It's unlikely that anything you're doing wouldn't be addressed by your calibration provider's scope, but it wouldn't hurt to get authoritative advice.

It could be that your auditor only expected that you would verify that your provider is generally accredited for calibration of oscilloscopes and wasn't concerned about the details, but that's a question for the auditor.
 

Attachments

dwperron

Trusted Information Resource
#3
The reason why your auditor said that you should verify that your calibrations are within the laboratory's scope of accreditation is probably because of some less than honorable practices by some "accredited" calibration labs. Customers often have requirements that they use accredited labs for their calibrations. So a few labs went after a very limited scope of accreditation (cheaper and easier) just to be able to claim that they were accredited and get the work. Your auditor is making sure that you are doing your due diligence to verify that your calibration laboratory has demonstrated competence in the areas where you are using them, not just using an accredited lab. That is what ISO 17025 requires.
Incidentally, we have a local competitor who is accredited by A2LA. Their scope of accreditation consists of only micrometer and caliper calibration! No joke (A2LA Certificate # 2714.01). This is the kind of "accredited lab" your auditor is concerned about.
 

Hershal

Metrologist-Auditor
Trusted Information Resource
#4
<snip> Had an auditor in here this week and he said I really SHOULD be verifying that the calibration laboratory I use is accredited to calibrate the test equipment I send them. I already have a copy of their latest Scope of Accreditation but I am confused on what I should be comparing this against. I'm assuming I need to know what my equipment's capability is and then compare this against the cal lab's scope document? To me this looks difficult especially for items such as an oscilloscope. <snip>
The thing for electronic you need to know includes the discipline involved. Typically a scope include time/frequency, modulation, and similar influences.

Your lab should be accredited for these kinds of things. Sometimes the scope will be scope-specific. Generally it is not. Understanding equipment and the discipline(s) involved and what it requires may be the key you need.
 
Thread starter Similar threads Forum Replies Date
L Procedure for Comparing the Monitor Reading with Laboratory Result Statistical Analysis Tools, Techniques and SPC 4
D Comparing inline Conductivity instrument with offline test results General Measurement Device and Calibration Topics 0
B New to ISO 14971. Comparing to MIL-STD-882 ISO 14971 - Medical Device Risk Management 7
P Comparing Two Test Variables Using Attribute Data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
S Equivalency study - Comparing 2 similar equipments Reliability Analysis - Predictions, Testing and Standards 1
E Comparing process parameters on non-normal batches Capability, Accuracy and Stability - Processes, Machines, etc. 11
S Comparing measurement results for a part from two ring gage General Measurement Device and Calibration Topics 3
R Statistical Methods for comparing test and reference product equivalence for quality attributes US Food and Drug Administration (FDA) 3
O Software for comparing text of PDF files Contract Review Process 2
M Comparing data from destructive testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
A Demonstration of Equivalence - Need for comparing biological characteristics for an SamD EU Medical Device Regulations 1
G Comparing Gauge Blocks - Translating the calibration between the master set and the QA set General Measurement Device and Calibration Topics 2
V Statistical basis and justification while comparing / changing sampling plans Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
N Does FDA allow comparing with the competitors Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S Looking for a checklist comparing ISO 13485:2016 and 21 CFR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 14
M Comparing the MMOG (v3 or v4) with ISO / TS16949 Customer and Company Specific Requirements 3
A Comparing US/FDA Drug Regulations against US/FDA Device Regulations US Food and Drug Administration (FDA) 3
E Correct Statistical Test comparing 2 Groups Statistical Analysis Tools, Techniques and SPC 14
C Comparing Two Independent Group Observations - 95% Confidence Intervals Overlap Statistical Analysis Tools, Techniques and SPC 3
D Comparing the ISO 9001:2008 standard to ISO 17025 ISO 17025 related Discussions 2
S Comparing Quality System Software Features Quality Assurance and Compliance Software Tools and Solutions 18
C Comparing Two Test Methods - Acceptable Difference(s)? Statistical Analysis Tools, Techniques and SPC 3
M pH Calibration - Comparing Electrical and CRM Methods General Measurement Device and Calibration Topics 1
I Six Sigma comparing to other National and International Models Six Sigma 6
A Comparing Effect of Process Parameters on Equipment Performance during Start Ups Statistical Analysis Tools, Techniques and SPC 3
P Acceptable Adjusted R Squared Value Comparing 2 Measurement Devices in Minitab Using Minitab Software 3
C Comparing PPM Year over Year - Determining the PPM decrease from 2009 to 2010 Benchmarking 10
M z Test vs. F Test - The most appropriate for comparing the Variances Statistical Analysis Tools, Techniques and SPC 3
T Comparing Two Measurement Systems in terms of Accuracy and Capability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
R Comparing the Accuracy of Analog vs. Digital Multimeter General Measurement Device and Calibration Topics 6
T Comparing two different Ultrasonic Tester Calibration Techniques General Measurement Device and Calibration Topics 2
Q Internal Audit:Comparing Efficiency of Similar Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
D Comparing Variation - A CMM and an On-Line Inspection Station Statistical Analysis Tools, Techniques and SPC 6
N Statistically Comparing Technician Techniques Reliability Analysis - Predictions, Testing and Standards 16
C Comparing Histograms - Bimodal Distribution? Statistical Analysis Tools, Techniques and SPC 4
K Comparing CQE vs. MSc. Professional Certifications and Degrees 1
J To PPAP or Not? Comparing two machines (one old and one new) APQP and PPAP 6
C Comparing OHSAS 18001 to the new ANSI Z10 - Where I can get information? Occupational Health & Safety Management Standards 1
J Has anyone created a matrix comparing ISO 9001 VS BS 7799 and or ISO 27001? Document Control Systems, Procedures, Forms and Templates 1
A Comparing six sigma & EFQM (European Foundation for Quality Management) Six Sigma 2
C Comparing what I do against Canada request - Immigration Immigration and Expatriate Topics 4
J KAPPA Statistic - Comparing appraisers ratings to a known standard IATF 16949 - Automotive Quality Systems Standard 8
R Comparing MicroVu?s Vertex machine to OPG?s Smart Scope Flash General Measurement Device and Calibration Topics 3
D Comparing ISO 10012 and 17025 ISO 17025 related Discussions 2
S Comparing ISO with PMBOK (Project Management Body of Knowledge) Misc. Quality Assurance and Business Systems Related Topics 2
D Comparing Apples and Apples vs. Apples and Oranges QS-9000 - American Automotive Manufacturers Standard 8
C Gap Analysis comparing QS-9000 and TS 16949 IATF 16949 - Automotive Quality Systems Standard 33
D Is Good Laboratory Practice (GLP) Required in ISO 17025? ISO 17025 related Discussions 4
D IATF16949 7.1.5.3.2 External Laboratory IATF 16949 - Automotive Quality Systems Standard 17
A Laboratory scope on the IATF IATF 16949 - Automotive Quality Systems Standard 5

Similar threads