Complaint alert levels - Industry standards pertaining to alert levels for complaints

ASteel

Starting to get Involved
#1
The company I work for has established complaint alert levels and I have been tasked with the inquiry of industry standards. All I have been able to find are levels regarding bioburden. Question to the fellow Quality members of the forum is...are you familiar with any industry standards pertaining to alert levels for complaints. Thanks for your help!

Are current levels are as follows:
Triggers by Absolute Numbers:
  • Other Complaint Types > 1/1,000 (0.1%)
  • Potential Adverse Events > 1/10,000 (0.01%)
    • Children’s Potential Adverse Events > 1/10,000 (0.01%)
Triggers by Rate:
  • Other Complaint Types > 2 Standard Deviation above the Average
  • Potential Adverse Events > 2 Standard Deviation above the Average
    • Children’s Potential Adverse Events > 2 Standard Deviation above the Average
 
Elsmar Forum Sponsor
#2
Do you work for a medical device company? Are the alert levels assigned only for low risk complaints? What happens when the alert limit is reached? All complaints need to be assessed according to a previously performed risk analysis. Are you doing that? If you answer the questions, maybe we can help you further.
 

ASteel

Starting to get Involved
#3
Do you work for a medical device company? Are the alert levels assigned only for low risk complaints? What happens when the alert limit is reached? All complaints need to be assessed according to a previously performed risk analysis. Are you doing that? If you answer the questions, maybe we can help you further.
Yes, it is a medical device company. The alert levels are separated into Other, which are low risk complaints, and Potential Adverse and Children's
Potential Adverse,
which are higher risk (e.g., efficacy, safety, etc.). We perform an investigation for all complaints (e.g., batch record review); however, when an alert level is reached, we perform a more in-depth investigation (e.g., HHE). Please clarify what you mean when you say "assessed according to a previously performed risk analysis". During the investigation of every complaint we currently ensure that the complaint has been established within our risk management and document if the risk hazard index is in need of adjustment.
 
#4
When I say assessed according to a previously performed risk analysis, I mean you go back and ensure the hazardous situation that occurred for the complaint is recorded in your hazard analysis and that the risk level is correct. It seems as though you are doing this.

As for having complaint alert levels, I am not aware of any standards that pertain to alert levels. The only thing I can think of that may be related is trend reporting per MEDDEV 2.12-1. However, I don't believe this is what you are taking about. Also, I do not fully understand the what this does for your organization. You should be performing a review on a regular basis to meet the requirements of ISO 14971 production and post production information. To do this review, you need to look at way more info than occurrence levels.

I hope someone else can chime in and provide additional insight into the idea of alert levels for complaints.
 

ASteel

Starting to get Involved
#5
Our alert levels are a way of noticing an upward trend in certain types of complaints. This analysis, among others, is included in a quarterly feedback report. I wouldn't think that we are revolutionary in our setting alert levels for complaints, so hopefully someone else can give some more insight.

Thank you for your input.
 

Bev D

Heretical Statistician
Leader
Super Moderator
#6
I would strongly caution against using 2 SD. by the rule of the formula for the standard deviation, there will be ~5% false alarms. Real control charts are much better for alert levels if you are simply trying to see real upward trends.

For high severity failures I recommend using a severity based approach. Higher severity failures will have lower limits for the number or proportion of complaints. For example, we have a limit of one complaint for any power supply that smokes or catches on fire.
 
Thread starter Similar threads Forum Replies Date
A Complaint VS CAPA report ISO 13485:2016 - Medical Device Quality Management Systems 5
P "Missing Label" Complaint Manufacturing and Related Processes 5
P Out-of-Box Failures during installation, complaint? Customer Complaints 2
G Customer complaint AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 22
M Is there a mandatory requirement to notify customers of the complaint resolution in ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
B IATF 16949 clause 10.2 - requirements following a customer complaint. IATF 16949 - Automotive Quality Systems Standard 9
B Spreadsheet - Used for complaint investigation - Validation required or not ISO 13485:2016 - Medical Device Quality Management Systems 9
Sidney Vianna IATF 16949 News IATF Performance Complaint Management System (IATF CMS) within the IATF Database IATF 16949 - Automotive Quality Systems Standard 4
A Are defective returned spare parts to be considered as a complaint? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
L Customer Complaint Internal Auditing 3
shimonv Device deficiencies (~complaint) discovered during a clinical trial ISO 13485:2016 - Medical Device Quality Management Systems 3
M Non conformance/complaint and CAPA ISO 13485:2016 - Medical Device Quality Management Systems 6
C Complaint Return Sample Size Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Melissa Complaint about Adhesive ISO 13485:2016 - Medical Device Quality Management Systems 9
K Do Contract Manufacturers need a complaint procedure? Medical Device and FDA Regulations and Standards News 10
A Complaint review as part of the complaint handling process? ISO 13485:2016 - Medical Device Quality Management Systems 3
M Reduce occurrence rating based on the PMS data and customer complaint data ISO 14971 - Medical Device Risk Management 2
B Complaint Records - Accessing records on Easy Track Software Records and Data - Quality, Legal and Other Evidence 3
J Customer Complaint & SCAR, false data Nonconformance and Corrective Action 14
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
J Customer Complaint Response 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S How to treat a customer complaint ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Requirements for Complaint Files - Distributed Goods Customer Complaints 3
S Is any dissatisfaction over a Medical Device considered as a complaint? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
B Procedures for Complaint Handling and Post Market Surveillance EU Medical Device Regulations 10
J KPIs or Metrics to Measure a New Complaint Handling Process 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
D Class II medical device - When should a complaint be closed? Customer Complaints 6
PhilM Nonconformance report, Customer complaint investigations and RMAs ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
G Problem Resolution Report Monitoring - Customer complaint or PRR as general motors use Customer Complaints 12
R Does an RMA = a Customer Complaint Customer Complaints 4
G Customer complaint over a Reference Dimension Manufacturing and Related Processes 9
Z Security for Approvals - Cloud based Complaint, NC, and CAPA systems Qualification and Validation (including 21 CFR Part 11) 8
S How to collect complaint data from customers Customer Complaints 24
E Is there an Automotive Customer Complaint Index? IATF 16949 - Automotive Quality Systems Standard 3
R In a software development company: Is every bug reported by the customer a complaint? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 27
E Corrective Action or Customer Complaint ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
T Decontamination responsibility on customer complaint investigation Customer Complaints 6
N Possibly worlds first customer service complaint Supply Chain Security Management Systems 0
A What is a Customer Complaint? (ISO 9001:2015) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
S What is the difference between a service request and a complaint? ISO 13485:2016 - Medical Device Quality Management Systems 2
M Should Potential Customer Complaint Outcome Define Registrar NC Rating? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
J FDA definition of a complaint - Customer orders spare parts 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R 510(k) - Finished Test Reports - RTA - Bit of a Complaint 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
P Customer Complaint Definition - Food Distributor Customer Complaints 5
J Using Complaint Rate to Update FMEA P.O.C ISO 14971 - Medical Device Risk Management 4
C Complaint Training - FDA - How to recognize complaints and what actions to take 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
M Complaint Handling Responsibilities for a Design Partner - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 9
S Merging Post Market Surveillance and Complaint Handling ISO 13485:2016 - Medical Device Quality Management Systems 3
R Customer Complaint - Samples vs. Removal - FDA Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
J Retention Requirements of Complaint Medical Devices Customer Complaints 2

Similar threads

Top Bottom