Complaint Handling Procedure for Medical Devices

S

sokku

#1
Dear Professionals

We are the medical devices company and implementing FDA QSR, As part of implementation, we need to implement complaint handling procedures.

Is there any different between customer feedback and complaint handling? If so

A significant compliant was received from the field regarding the performance of our device, this particular complaint was reported from a large number of customers.

What procedure do we follow to handle these complaints?

How would we decide whether to remove the affected device from the field?

I appreciate your valuable comments

-Sokku
 
Elsmar Forum Sponsor
A

arios

#2
Re: Complaint Handling

Hi Sokku

Based on your comments regarding the amount and type of complaints you may be facing a possible need to perform a Field Safety Corrective Action. The decision will have have to be based on a risk assessment taking in consideration the actual or potential impact on the patient, user or other persons.

About the other question about customer feedback and complaint handling, they are not necessarily the same. A feedback could be positive or a suggestion, thus not necessarily a customer complaint.

You will need to determine if the complaints represent reportable events (MDR's), document an investigation, and initiate your CAPA process.

Alberto
 

sreenu927

Quite Involved in Discussions
#3
Hi Sukku,

To be in compliance with QSR, you must have documented procedures for Complaint Handling and Medical Device Reporting (MDR), apart from other SOPs.

As told by Alberto, feedback and complaint handling may or may not be iincluded in the single procedure.

For ur case, the complaint handing procedure shud capture the complaints, which you need to evaluate the complaint, whether or not it is reportable to FDA. If so, that triggers your MDR. During your investigation, if you decide to do any field action (relabel, modification, repair and/or recall), then a Field Safety Corrective Action procedure will be triggered. If no field action,then investigate the complaint and close CAPA.

Hope this helps.

Regards
Sreenu
 
D

Dudes

#4
Hi Sokku,

let's try to answer your different questions...


Dear Professionals

We are the medical devices company and implementing FDA QSR, As part of implementation, we need to implement complaint handling procedures.

Is there any different between customer feedback and complaint handling? If so
As other stated, feedback can be either positive or negative. Complaint would only be negative.

Although you have to document both, complaint handling will give you much more work. Once you have received a praise, there isn't much more you can do than document it, file it, and add to your monitoring for the management review.

When you receive a complaint, you have to investigate it. Is it relevant? Where does the problem comes from? Can you correct it? Can you find the ROOT cause and take preventive actions to be sure it doesn't happen again (CAPA)? ...

A good monitoring of your complaints shoud allow you to discover and address problem early on...



A significant compliant was received from the field regarding the performance of our device, this particular complaint was reported from a large number of customers.

What procedure do we follow to handle these complaints?
The procedure would be the same for one complaint as for many. Depending on the risk (both to the patient/user and commercial), you may want to put more ressources in the resolution.



How would we decide whether to remove the affected device from the field?

I appreciate your valuable comments

-Sokku
You have to perform a risk analysis. What are the risks of your device? Does it present any danger to the patient? In that case, you may have to notify it to your regulatory agency.

A recall, withdrawal or any other FSCA (field safety corrective action) has to happen based on this risk analysis.


I hope it helps :2cents:
 

Ronen E

Problem Solver
Staff member
Moderator
#5
Dear Professionals

We are the medical devices company and implementing FDA QSR, As part of implementation, we need to implement complaint handling procedures.

Is there any different between customer feedback and complaint handling? If so

A significant compliant was received from the field regarding the performance of our device, this particular complaint was reported from a large number of customers.

What procedure do we follow to handle these complaints?

How would we decide whether to remove the affected device from the field?

I appreciate your valuable comments

-Sokku
Dear Sokku,

After reading your post, I’m not 100% sure of your situation. Are you just preparing your QSR compliant procedures, or do you already have them in place?

If you already have written procedures in place, you should follow them. It sounds as if this is not a good time to start making changes. If you are unsure whether or not your current procedures meet the QSR, it is best to have someone experienced carefully examine them. It would be very difficult to tell, without actually reading your procedures, and you would probably not want to expose them in a public forum.

If, on the other hand, you still don’t have such written procedures, or you are in the process of preparing them (or upgrading existing ones to meet the QSR), then you shouldn’t be selling into the US market. If you are still not selling into the US market then you probably don’t have to report the FDA.

As to your questions –

Apart from customer feedback being possibly positive, another difference is that customer complaints are received passively, while you are expected to be active in gathering general customer feedback (sending out questionnaires, conducting annual survey meetings etc.). Sometimes a customer complaint is not declared as such by the source, and can be received as part of the general customer feedback collection system. It is recommended, however, to record any specific negative feedback received from the field, regardless of the way it was received, as a formal complaint, and handle it as such.

In handling customer complaints you should follow your documented customer complaints procedure. This, in turn, should be based on relevant FDA guidance. I think going to the source and studying it thoroughly is much better than relying on abbreviated second-hand advice. Having said that, here is some abbreviated second-hand advice :D

1. Formal MDR is required in the following situations:
· Your device may have caused or contributed to a death or serious injury, or
· Your device has malfunctioned, while recurrence of the malfunction in this type of device or a similar one, marketed by you, would be likely to cause or contribute to a death or serious injury.
“Significant” is a term that may have different meanings in different settings, and therefore can't be used for decision-making. The above, on the other hand, is FDA terminology and further elaborated definitions for the terms used are available from the FDA.

2. For a decision whether or not to remove the affected device from the field, it is best to consult the FDA guidance on RECALL (or your recall procedure, if based on FDA guidance).

3. To itself, the large number of complaints of a similar nature does not necessarily mean that the problem is severe (i.e. in risk terms). However, several incidents of a minor severity may call for a response similar to that given to a single major incident (similarly, several major may be equivalent to one critical).

Good luck,
Ronen.
 
S

sueliu - 2012

#6
I have a question regarding FDA's requirement on complaint handling as well.

We just went through an ISO 13485/FDA inspection (combined) audit, the auditor mentioned we should have a question list to assess whether the patient health is at risk for each single complaint we get. I am wondering if anybody could show me the links or locations where I can find this question list. Or, if anybody can share me an example.
 
V

Vincnet

#7
Sueliu

afssaps in France issued a very usefull document that we are supposed to use when declaring a vigilance event.
the last page of the document is a flowchart designed to help people decinding what it the impact of the compliant and what should be done.

I attached a copy if this doc here.
So far i never used it with non french clients so i do not have a translation but wording is very simple and google should be abble to translate it.

Hope this will help you

Vincent
 

Attachments

Q

QA compliance

#9
I have found all my answers to this question plus much, much more right here. Click on the Post Attachments List on the top of this page. Then for a keyword, type in "customer". Give this a try and see if it helps!
 
Thread starter Similar threads Forum Replies Date
B Software Validation Procedure Question - Complaint Handling Process software Software Quality Assurance 9
B Procedures for Complaint Handling and Post Market Surveillance EU Medical Device Regulations 7
J KPIs or Metrics to Measure a New Complaint Handling Process 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
M Complaint Handling Responsibilities for a Design Partner - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 9
S Merging Post Market Surveillance and Complaint Handling ISO 13485:2016 - Medical Device Quality Management Systems 3
A Production and Post-Production and Complaint Handling ISO 14971 - Medical Device Risk Management 2
A Risk Management, complaint handling and CAPA system ISO 14971 - Medical Device Risk Management 5
D Complaint Handling Call Centers ISO 13485:2016 - Medical Device Quality Management Systems 7
M Manufacturer's handling of Distributor's Historical Complaint Records US Food and Drug Administration (FDA) 5
Michael Malis What is the best Complaint Handling system? Customer Complaints 16
T US Clinical Study conducted in EU - What are EU Complaint handling reqts? EU Medical Device Regulations 4
M Reduce occurrence rating based on the PMS data and customer complaint data ISO 14971 - Medical Device Risk Management 2
B Complaint Records - Accessing records on Easy Track Software Records and Data - Quality, Legal and Other Evidence 3
J Customer Complaint & SCAR, false data Nonconformance and Corrective Action 14
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
J Customer Complaint Response 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S How to treat a customer complaint ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Requirements for Complaint Files - Distributed Goods Customer Complaints 3
S Is any dissatisfaction over a Medical Device considered as a complaint? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
A Complaint alert levels - Industry standards pertaining to alert levels for complaints Customer Complaints 5
D Class II medical device - When should a complaint be closed? Customer Complaints 6
PhilM Nonconformance report, Customer complaint investigations and RMAs ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
G Problem Resolution Report Monitoring - Customer complaint or PRR as general motors use Customer Complaints 12
R Does an RMA = a Customer Complaint Customer Complaints 4
G Customer complaint over a Reference Dimension Manufacturing and Related Processes 9
Z Security for Approvals - Cloud based Complaint, NC, and CAPA systems Qualification and Validation (including 21 CFR Part 11) 8
S How to collect complaint data from customers Customer Complaints 24
E Is there an Automotive Customer Complaint Index? IATF 16949 - Automotive Quality Systems Standard 3
R In a software development company: Is every bug reported by the customer a complaint? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 27
E Corrective Action or Customer Complaint ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
T Decontamination responsibility on customer complaint investigation Customer Complaints 4
N Possibly worlds first customer service complaint Supply Chain Security Management Systems 0
A What is a Customer Complaint? (ISO 9001:2015) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
S What is the difference between a service request and a complaint? ISO 13485:2016 - Medical Device Quality Management Systems 2
M Should Potential Customer Complaint Outcome Define Registrar NC Rating? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
J FDA definition of a complaint - Customer orders spare parts 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R 510(k) - Finished Test Reports - RTA - Bit of a Complaint 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
P Customer Complaint Definition - Food Distributor Customer Complaints 5
J Using Complaint Rate to Update FMEA P.O.C ISO 14971 - Medical Device Risk Management 4
C Complaint Training - FDA - How to recognize complaints and what actions to take 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
R Customer Complaint - Samples vs. Removal - FDA Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
J Retention Requirements of Complaint Medical Devices Customer Complaints 2
C Web based Complaint Software which interfaces with our Enovia system Document Control Systems, Procedures, Forms and Templates 0
Moncia Complaint / Non-Conforming Product / CAPA Procedures - Please share your experience ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Moncia Customer Complaint Procedure example needed Document Control Systems, Procedures, Forms and Templates 1
C Medical Device Malfunction - Customer Complaint - Response vs. Correction Other US Medical Device Regulations 3
J ISO 13485 Complaints (what is a complaint) Customer Complaints 5
Q Internal Audit of Product Quality Complaint System (21 CFR Part 820) Customer Complaints 9
T PPAP - Complaint from a Tier One customer concerning Product Dimension IATF 16949 - Automotive Quality Systems Standard 26
M Using a Complaint Management Software (Medical Device companies) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2