So I had a "Shower Thought" the other day that I'm trying to check if it's insane or if this is going to be a fundamental shift in Complaint Handling application under the proposed QMSR. This may come across as rambling, but there are a few moving pieces here.
We know that under the QSR the following is a complaint:
They state in the definition section:
So in summary: If the FDA is choosing to incorporate the 13485 definition of a complaint and no longer retain theirs, does that infer that the FDA will expect Complaint Handling to be performed on devices no longer in the organization's control too? Could you realistically get a complaint from a vendor and/or a customer? Examples off the top of my head, mislabeled devices sent to a supplier for service? CMs getting SCARs from a Customer. I can see the cases in some scenarios where complaint handling may already be occurring per the 13485 definitions, but I can see plenty where organizations may not have any complaints (QSR only contract manufacturers) and will in the future. Any other interpretations of the above?
We know that under the QSR the following is a complaint:
It's always been my understanding, that distribution, coupled with the following definition from 21CFR807, refers to a device that is commercially offered for sale on the market.(b) Complaint means any written, electronic, or oral communication that alleges deficiencies related to the identity, quality, durability, reliability, safety, effectiveness, or performance of a device after it is released for distribution. - 21CFR820.3(b)
That said, if you were manufacturing a device that had not yet been released on the market, Complaint Handling would not apply. Further, Complaint Handling under the QSR has been the responsibility of the Manufacturer, unless written agreements state otherwise, per the language in 820.198.Commercial distribution means any distribution of a device intended for human use which is held or offered for sale - 21CFR807.3(b).
Now let's get into the application of the QMSR. We know the FDA is proposing to amend the CGMP to incorporate 13485 and add additional requirements on top of it.(a) Each manufacturer shall maintain complaint files. Each manufacturer shall establish and maintain procedures for receiving, reviewing, and evaluating complaints by a formally designated unit. - 21CFR820.198(a).
They state in the definition section:
We have reviewed the definitions in ISO 13485 to determine their suitability for FDA's purposes. We find that most of the definitions in Clause 3 are acceptable; thus, unless identified in this section, we are not proposing any modifications to the terms and definitions in Clause 3 and are proposing to remove the correlating terms and definitions from the current part 820.
We are retaining the majority of the definition of “rework”; however, we are proposing to remove the term “device master record (DMR)” (§ 820.3(j)) from the regulation.
It is noted that Complaint is NOT one of the terms the FDA has identified to retain from the QSR. I was curious to see if "Finished Device" was going to be the out here, but that does not appear to be called into 21CFR820.198 as a distinction between a device and a finished device. So that then points us to the 13485 definition of a Complaint, which states:We are retaining without change the terms and definitions for “component” (§ 820.3(c)); “finished device” (§ 820.3(l)); “human cell, tissue, or cellular or tissue-based product (HCT/P) regulated as a device” (820.3(bb)); “design validation” (§ 820.3(z)(2)); “remanufacturer” (§ 820.3(w)); “nonconformity” (§ 820.3(q)); and “verification” (820.3(aa)) because these terms are necessary for implementing part 820.
It's been my understanding if you were both ISO 13485 certified and compliant with the QSR, based on the language in section 0.1, you could adopt the FDA definition of a complaint and go with it. I've not been challenged on that in my entire career, as both a manufacturer and a contract manufacturer, however, my frame of reference is usually FDA-based.“written, electronic or oral communication that alleges deficiencies related to the identity, quality, durability, reliability, usability, safety or performance of a medical device that has been released from the organization’s control or related to a service that affects the performance of such medical devices.” - ISO 13485:2016 3.4.
So in summary: If the FDA is choosing to incorporate the 13485 definition of a complaint and no longer retain theirs, does that infer that the FDA will expect Complaint Handling to be performed on devices no longer in the organization's control too? Could you realistically get a complaint from a vendor and/or a customer? Examples off the top of my head, mislabeled devices sent to a supplier for service? CMs getting SCARs from a Customer. I can see the cases in some scenarios where complaint handling may already be occurring per the 13485 definitions, but I can see plenty where organizations may not have any complaints (QSR only contract manufacturers) and will in the future. Any other interpretations of the above?