Hello
I am with a company that is currently making a product that can be used in diagnostics as well as in the food industry.
We have a client that wishes our product to be comliant with CFR 21 part 11.
We are also certified in ISO 9001
My question is whether we have to be certified under ISO 13485 or ISO 17025 to claim compliance with CFR 21 part 11 or is it enough that we are certified under ISO 9001?
Thank you
sagai
Hello urishab,
Welcome to the Cove!
21CFR Part11 is the requirements set forth by FDA to determine such conditions when the Agency (aka FDA) considers electronic records equivalent with paper record including signatures on such electronic records whereas these signatures either electronic or digital signatures . Also, this set of requirements is for manufacturers subject to FDA scrutiny including food, drug, tobacco, medical device, etc. industries subject to the Food and Drug and Cosmetic Act in the US.
Further it is a requirement only for those records those are subject to the corresponding industry GMP requirement that also set forth by FDA.
So in case my impression is correct and you are a supplier for a manufacturere than unless you have some legacy kind of contractual agreement that would say so, my opinion picturized in one symbol would be ... let me see .... the following:
:deadhorse:
May be it was not that explicit so ... ISO13485 / ISO9001 has nothing to do with 21CFR11, it is a kind of oxymoron.
That's methink.
:bigwave:
Cheers!
v9991
Hello
Thank you
is your product used to generate any records/data which is required as per CFR 101 OR any CFR 820?
If yes, does your product involve software to maintain the data?
if yes, then you need to comply to the requirements set forth by 21 CFR part 11.
now the second part of question about being certified....
1st place, you need to have a pertinent development documentation maintained as per the respective guidelines. there is no specific recommended methodology/standard but depending upon the kind of software involved, a full scale GAMP or SEI CMM model OR a simple demonstration of qualification/verification could do.
sagai
NO!
ISO13485 is not a reason to comply with 21CFR11, sorry, there is no such connection between these two in any means.
and again if the company is not subject to GMP in US, than no reason for validation based on regulatory requirements.
Further, neither ISO13495 nor ISO9001 expect software validation other the software used for service provisioning or for manufacturing.
Another thing very silently is that I have never come across any warning letter that refers to 21CFR11, however I read more or less all of them for MDs in a last couple of years.
Regards
v9991
NO!
true, mentioning 13485 was by error; hence I corrected it immediately;
again geography related part is correct; although for Europe, one needs to comply with Annex-11; which is more or less similar to that of part-11;
and since the customer is requesting for part-21 compliance, I would guess that either customer or the target market would have something to do with US,
U
urishab
Wow
thanks guys for the very quick and detailed answers.
Our customer is European (Germany) and our product includes software that creates electronic records (images, analysis, etc...)
Those records are used in the manufacture of drugs or the quality assurance of food (livestock testing).
Therefore I believe our customers (such as veterinary labs) must comply with CFR 21 part 11.
Based on this information and given that part 11 does not mandate any specific ISO, do you believe that we need to be certified in anything beyond ISO 9001 to sell in Europer / US?
p.s. I had to think (and google) to understand the dead horse thing. Didn't know that phrase :applause:
yodon
Let me try to clarify the thinking just a bit.
Part 11 is for electronic records / electronic signatures under US regulations only. In and of itself, you wouldn't comply. When there are "predicate rules" by which you must abide and you choose to fulfill those requirements using electronic records and/or electronic signatures, then Part 11 applies. For example, 21 CFR 820 (Quality Systems Regulation) requires that design inputs be documented and approved (by signature). This is the predicate rule. If you choose to maintain this document electronically (not a paper copy) then you would need to comply with Part 11 for electronic records. If you execute approval via electronic means then you have to comply with Part 11 for electronic signatures.
As others said, there's no correlation whatsoever between Part 11 and ISO (any flavor).
So first understand what predicate rules apply and then determine if Part 11 is applicable.
Since you mention your customer is German, there might be another / a different aspect to consider. Annex 11 addresses computerized systems used in medicinal products for human / veterinary use. I believe this is just a guidance (compliance not compulsory). Make sure they really meant Part 11 and not Annex 11.
Hope that helps.
sagai
Mihzaga, I am lost regardless I had a read of your link.
Could you point me to the section that says part11 violation?
The other thing come into my mind for the original poster is that in this scenario the liability to comply with part11 lies on your customer, and actually, you are not in a position to issue any certificate what so ever that tells them that your product is part11 compliant, regardless you would willing to. Simple because the liability is on their shoulders but yours.
The things I have seen so far in these scenarios that the supplier (that is you actually) could provide a set of template to accelerate the manufacturer validation activity.
However you genuinely can not supply the full material simple because it is not possible to foresee the precise use and the extent of such use of your product.
that's come into my mind, could be foolish, up to you. :)
Cheers!
v9991
compliance to part 11 or annex 11, is not as much about the certification as it is about a structured approach towards implementation of software (supported by the design & development of software );
think from the perspective of your customer; He's the one who needs to demonstrate compliance to pertinent requirements. (e-records & e-signatures )
Now, as a part of it,
1) the software of the diagnostic device you are supplying has to support same. .., i.e., features pertaining to the managing the
users :- authorization profiles, access rights, signatures part wrt id&passwords, etc
records :- time stamps, versions, e-data vs human readable paper records, audit trails, back. etc
2) + demonstrate a "standard methodology" being adopted for software development & management.
this is point where I have earlier indicated some traceable methodology such as GAMP or SEI CMM level have some relevance. However it is not necessary/mandatory to be certified by any of above, it helps and simplifies the steps.
I believe at the minimum you should demonstrate, configuration management (representing the development, design document, ), test management(traceability matrix & integration& unit testing, etc.,), requirements management (URS , FRS )
L Wearables 21 CFR Part 11 compliance Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
Ed Panek Do Cloud services require 21 CFR Part 11 compliance? Qualification and Validation (including 21 CFR Part 11) 7
J Business Intelligence and 21 CFR Part 11 Compliance Qualification and Validation (including 21 CFR Part 11) 1
I GMP 21 CFR Part 11 Electronic Records Compliance Project Help Qualification and Validation (including 21 CFR Part 11) 9
N Change Control - Compliance with FDA 21 CFR Part 820 Document Control Systems, Procedures, Forms and Templates 3
I 510(k) Raw Data from Studies - 21 CFR Part 11 Compliance Qualification and Validation (including 21 CFR Part 11) 4
L Design software for a spectrophotometer for 21 CFR Part 11 compliance Qualification and Validation (including 21 CFR Part 11) 1
V Should template/formats be pre configured to claim 21 CFR part 11 compliance? Qualification and Validation (including 21 CFR Part 11) 5
D Compliance to 21 CFR Part 11 Require Registration to FDA? Qualification and Validation (including 21 CFR Part 11) 9
G Original Data Copying to Maintain Compliance with 21 CFR Part 820.180 US Food and Drug Administration (FDA) 2
L Compliance to 21 CFR Part 11 - Where to start?? Qualification and Validation (including 21 CFR Part 11) 20
T 21 CFR Part 11 Compliance - Generation of Complete Copies of Records Qualification and Validation (including 21 CFR Part 11) 3
T 21 CFR Part 11 Compliance - Using PDF files as our documentation masters Quality Assurance and Compliance Software Tools and Solutions 3
N Certificate of Compliance from a Sister Company for FDA 21 CFR 820 Exemption 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
T Quality Compliance Software to meet ISO 13485 QMS and 21 CFR 820 Requirements Quality Assurance and Compliance Software Tools and Solutions 3
N Upgrading ISO 13485:2003 QMS to FDA 21 CFR 820 Compliance ISO 13485:2016 - Medical Device Quality Management Systems 1
T Compliance Dept KPI Dashboard AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C Regulatory requirements of cloud platform for devices developed in compliance with 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
H Does this clause mean i must be in compliance with AS9100 or is 9001 enough? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
I Where to find MDR compliance device? EU Medical Device Regulations 6
E Compliance with GSPR 10.4 CMR/ED EU Medical Device Regulations 9
G Suggestions for environmental compliance with small machinists? REACH and RoHS Conversations 1
S Training for PRRC (person responsible for regulatory compliance) EU Medical Device Regulations 3
R Evidence of compliance with Directive 2013/59/Euratom EU Medical Device Regulations 3
T Electronic Systems and 8.5.2 compliance AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
Sidney Vianna ISO 37301 - Compliance management systems – Requirements with guidance for use Other ISO and International Standards and European Regulations 2
H Regarding compliance according WEEE Directive CE Marking (Conformité Européene) / CB Scheme 6
M Regulatory compliance ISO 14001 ISO 14001:2015 Specific Discussions 5
S Requirements for ANSI Z540.3 Compliance as End User of Calibrated Instruments AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
M How to build a winning strategy for EU MDR Compliance Book, Video, Blog and Web Site Reviews and Recommendations 0
A Compliance with iso 14155 Other Medical Device Related Standards 1
T Determining FIPS 140-2 compliance in a medical device Medical Information Technology, Medical Software and Health Informatics 2
C Deadline for compliance with added MDR harmonized standards CE Marking (Conformité Européene) / CB Scheme 2
I User ID verification for in-house e-signature compliance Qualification and Validation (including 21 CFR Part 11) 2
Ron Rompen 4.4.1.2 Product Safety Compliance IATF 16949 - Automotive Quality Systems Standard 29
W 6.1.3 Compliance obligations b) - problem with understanding ISO 14001:2015 Specific Discussions 0
S Question: Best alternatives to audits for ensuring compliance Quality Tools, Improvement and Analysis 5
K Compliance Obligations 6.1.3 of ISO 14001 ISO 14001:2015 Specific Discussions 3
Ed Panek Which choice of the following most closely matches the MDR compliance of your company? EU Medical Device Regulations 0
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Proof QMS compliance for class I ISO 13485:2016 - Medical Device Quality Management Systems 1
G Compliance with ISO 9001-2015 for ISO 17025 Accredited Labs? ISO 17025 related Discussions 8
L REACH compliance with MDR EU Medical Device Regulations 4
K ROHs compliance requirement REACH and RoHS Conversations 10
Z Choice of PTC heater voltage to maximize user safety and ease of CE compliance to LVD CE Marking (Conformité Européene) / CB Scheme 0
Z REACH compliance for low volume product REACH and RoHS Conversations 1
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
E Accredited vs. non-accredited labs for 60601 compliance in the US IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Accredited vs. non-accredited labs for 60601 compliance in the US Other Medical Device Related Standards 4
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14