Hello!
I'm trying to better understand the MDR requirements in regards to components for a finished medical device, that are shipped to both our Companies service team in the field, and also to authorized BioMeds at accounts. Occasionally these components, in this case an IV Pole, will break in transit. My concern is if an IV Pole were to break, and a service technician were to cut their hand, would I be obligated to file a MDR to capture the incident. Especially if the recipient of the component is a BioMed and not an employee of our company, in other words the component was no longer in our companies control. The IV Pole is not sold as a stand alone device. It can only be used with the finished good, and is only shipped to a tech/customer for repair. I'm unclear if this would fall under the realm of accessories.
Any thoughts would be greatly appreciated.
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Not an expert here but maybe it'll kickstart the discussion.

I don't think whether it's an accessory or not has any bearing.

I kept trying to convince myself that it wouldn't be reportable but if the event is considered a serious injury then it seems it would be since:
  • you might consider the injury a result of: ... (3) Improper or inadequate design,
  • the 'malfunction' may recur
I tend to take a more conservative approach to reporting. You'd definitely be wise to get regulatory expertise to weigh in.

You mention that it was out of your control; however you also say that it could be a result of damage in shipping. Sounds like your packaging (what I'm considering inadequate design above) may not adequately protect the component in shipping (and especially if this is recurring). Has there been any thought given to improving the packaging?
 
Thread starter Similar threads Forum Replies Date
C Warehousing two separate components to a finished device while meeting regulations Other US Medical Device Regulations 7
Z Proving REACH/RoHS status for electronics components from Digikey/Mouser REACH and RoHS Conversations 7
T Device & Accessory, Components in device, or System EU Medical Device Regulations 0
S Reporting REACH - Electronic components distributor REACH and RoHS Conversations 5
U Procedure pack components EU Medical Device Regulations 3
T Sterilized components expiration date ISO 13485:2016 - Medical Device Quality Management Systems 1
T Sterilized components shelf life Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
G Dealing with non conformity caused by Supplier Components detected in the production line IATF 16949 - Automotive Quality Systems Standard 14
C Shelf life electronic components 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
J Process FMEA Template with examples - Cold and Hot Forged components FMEA and Control Plans 4
F Biocompatibility evaluation for Hardware/Interface Components? Other Medical Device Related Standards 12
K 510K new manufacturer but same components Other US Medical Device Regulations 6
D Using non-conforming components even though the final assembly is conforming? Manufacturing and Related Processes 5
R DHR question: Traceability of components ISO 13485:2016 - Medical Device Quality Management Systems 2
R What is meant by Use of COMPONENTS WITH HIGH-INTEGRITY CHARACTERISTICS in ME EQUIPMENT (clause no 4.9) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J Medical Device Components US to EU commercialization EU Medical Device Regulations 7
B IATF 16949 - Is a Deviation required for sample components in a prototype build? IATF 16949 - Automotive Quality Systems Standard 13
A Components of a medical device - Hardware robotic hand EU Medical Device Regulations 3
D IMDS for Like Components (Interior Plastic, Leather/vinyl Armrest, etc.)other than Color APQP and PPAP 1
JoshuaFroud Serial numbers and labelling when sharing device components ISO 13485:2016 - Medical Device Quality Management Systems 1
A How to deal with changed shared components in 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Nicole Desouza Plastic Extrusion of Medical Device Components - No ISO or AS 9100 Certifications Other Medical Device and Orthopedic Related Topics 7
Q Handling Off-the-Shelf Components 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
I PPAP Assembly and Components Manufacturing and Related Processes 3
D CB Certificate - Distributor for a product with electrical components CE Marking (Conformité Européene) / CB Scheme 4
K Selling Medical Device Components to US hospitals and doctors US Food and Drug Administration (FDA) 4
S Need Assistance in Visual Inspection - Tiny electronic components AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
R FDA - How to discontinue/obsolete medical devices that share sub-components with others that will still be distributed Other US Medical Device Regulations 5
F Mig Welded Components - IMDS International Material Data System RoHS, REACH, ELV, IMDS and Restricted Substances 1
D Software Validation - Contract manufacturer of Components (PCBA's) Qualification and Validation (including 21 CFR Part 11) 7
G Audit finding - Components being transferred inter-plant Internal Auditing 3
J Referencing Medical Device in FDA records - Package contains several other components Other Medical Device Regulations World-Wide 1
K Article 23 (Parts and components) - Economic Operator EU Medical Device Regulations 1
S How are Medical Device Components and Subassemblies defined? ISO 13485:2016 - Medical Device Quality Management Systems 1
H Classification of a Class IIb Medical Device with 3 Components CE Marking (Conformité Européene) / CB Scheme 1
P REACH - A few of the hundreds components in my product no longer meet requirement REACH and RoHS Conversations 2
C Medical device file for components such as adhesives, gaskets, films, etc. ISO 13485:2016 - Medical Device Quality Management Systems 13
S Do suppliers of prototyping components need to undergo supplier qualification? ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Is it acceptable to mix components from two different lots into an assembly? ISO 13485:2016 - Medical Device Quality Management Systems 1
M CE / FCC Compliance for components + final product CE Marking (Conformité Européene) / CB Scheme 2
D Determining Critical Components for conformity with IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 21
R Components in parallel to creepage\clearance IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
J To Calibrate or not to calibrate electronics rack components Manufacturing and Related Processes 5
T Control of Nonconforming Product Output (Components Parts) IATF 16949 - Automotive Quality Systems Standard 3
H CE Marking on Multiple-Components Medical Products EU Medical Device Regulations 8
Q CE Mark Requirements for Components - Solenoid Actuator Type Devices CE Marking (Conformité Européene) / CB Scheme 3
P Total Variation in Data - Overall Importance of the Principal Components Reliability Analysis - Predictions, Testing and Standards 1
F FDA PMK 510(k) - IEC 62304 Software Components Segregation Other US Medical Device Regulations 3
J Obsolete or Delete - Accepted Components from Suppliers RoHS, REACH, ELV, IMDS and Restricted Substances 3
Highground Perfusion System - HLM Components with Vendor Neutrality Other US Medical Device Regulations 6

Similar threads

Top Bottom