Condition of Item Calibrated - ISO 17025 Clause 5.10.2f - Calibration Certificate
howste
Thaumaturge
I'm looking at ISO 17025 clause 5.10.2f, which states that the calibration certificate must include "...the condition of... the item(s) tested or calibrated..."
My question is this: are the "as found" readings sufficient to meet this requirement, or is it requiring something more? If it's something more, then what should it be - what's the intent?
JAltmann
Re: ISO 17025 Calibration Certificate - Condition of Item Calibrated
I have always listed it as a seperate item, ie new, good, fair, damaged, etc.
Seems most of my cert.'s (external) and even my gage tracking software have a spot for this.
howste
Thaumaturge
Thanks for the response, James. Do you work for an accredited lab?
I'm looking for a response for a calibration lab that will be accredited in the future to ISO 17025, and they want to know the interpretation of the requirement for mandatory minimum content.
JAltmann
Formerly i was the quality manager of an accredited lab. Has been a few years so the standard is not as familiar to me off the top of my head anymore.
Currently among other duties i oversee the calibrations for 16949 company. I'll have to take a look at the 17025 standard to know for sure. But am thinking a few others on here will know for sure.
sokolasty
Hi
I work in ISO TS company in Poland and I see on some calibration certificates from Polish or German labs, that equipment condition is new or poor or fine etc. I did not see any certificate with more detailed description.
Stijloor
Hi
When your organization sends out devices to be calibrated, you need to specify in what format and level of detail you want the calibration reports to be produced and submitted. Your purchase order should specify all that. If you don't, you get the bare minimum.
Stijloor.
BradM
Stijloor.
That's correct. Too, this is in the 17025, so those requirements would be present for the lab that is accredited to that specification.
If the lab is not accredited, you can end up with a very wide degree of variation in how they prepare the calibration certificate.
Hershal
Metrologist-Auditor
As an assesor if I see as found/as left and do NOT see uncertatinty, that is a serious non-conformity.....
It tells me the laboratory has not done ANY homework.....
Hope this helps
Hershal
Metrologist-Auditor
There have been resolution of issues with respect to a range of issues.....let us rebound back to this baseline as an example.....
There is still a range of issues that include a range of perspectives.....
howste
Thaumaturge
Are you saying that there is no generally accepted answer? As an assessor, would you accept as-found readings on the certificate acceptable to address the requirement for the condition of the item calibrated? What would you consider to be best practice?
By the way, this company calibrates scales.
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
J Supplier Controlled Shipping (GM CS1) Condition Manufacturing and Related Processes 3
E Normal Condition Hazards in Risk Analysis ISO 14971 - Medical Device Risk Management 3
C Ambient temperature_how we claim storage condition in label? Other Medical Device Related Standards 2
M Which incubation condition should be selected to recover both bacteria and fungus effectively Miscellaneous Environmental Standards and EMS Related Discussions 3
K How to verify the Storage and Transport environmental condition for Medical Device Miscellaneous Environmental Standards and EMS Related Discussions 3
S VW Special status condition IATF 16949 - Automotive Quality Systems Standard 3
G IEC 61010 - Single Fault Condition - Protective Impedance Implementation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Calibration of a Width Gage Slot with an Attribute Gage - Maximum Width Condition General Measurement Device and Calibration Topics 16
B Maximum Material Condition with Bonus - Determine Cpk for this one kpc Measurement Uncertainty (MU) 2
E Single Fault Condition Simulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J ISO 13715 Edge Condition Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
C Calibration Certificate "As Found" condition - TS16949 Section 7.6.2 General Measurement Device and Calibration Topics 7
I Run a batch every year to revalidate the validated condition? Qualification and Validation (including 21 CFR Part 11) 4
C Q as the Datum Maximum Material Condition then MINUS 0.020 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
R Managing Tooling Condition at Suppliers Supplier Quality Assurance and other Supplier Issues 6
G What is the exact meaning of Single Fault condition? EU Medical Device Regulations 4
R Single Fault Condition for Internally Powered Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D As Found Condition - In or out of tolerance? General Measurement Device and Calibration Topics 4
W Supply Voltage Condition for Transformer O/L and S/C (open & short circuit) testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K Maximum Material Condition on Diameter Tolerance Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
M Zero Tolerancing on True Position using MMC (Maximum Material Condition) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
D "As Found" Condition and Calibration Interval Changes Calibration Frequency (Interval) 5
M Common Reasons for Out Of Tolerance condition of DC Calibrator or any reference std General Measurement Device and Calibration Topics 1
R Single Fault Condition - Motor Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
R What Standard or method to use to measure Bow condition in M6 Bolts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
R Is foreseeable misuse considered as single fault condition? IEC 60601 - Medical Electrical Equipment Safety Standards Series 33
G Single Fault Condition for 3-Phase Medical System - IEC 60601 Clause 8.1b IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C How Long it takes to Condition Pressure Gauge and Test Gauge General Measurement Device and Calibration Topics 4
B CMM Stylus Calibration - Required to record the As Found condition AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
D Process Capability Study which involves MMC (Maximum Material Condition) Capability, Accuracy and Stability - Processes, Machines, etc. 8
G ANSI & ASME regardless of Maximum Material Condition when none is specified Various Other Specifications, Standards, and related Requirements 3
D Herringbone Defect Condition in Rolled Steel Misc. Quality Assurance and Business Systems Related Topics 3
B Analytical Lab Environmental Condition (Temperature & Humidity Control Limit) General Measurement Device and Calibration Topics 5
L Root Cause Condition / Corrective Action Condition Problem Solving, Root Cause Fault and Failure Analysis 2
Chennaiite Process Capability for a specification with Maximum Material Condition (MMC) Statistical Analysis Tools, Techniques and SPC 4
S Condition of 100% inspection in organization. Capability, Accuracy and Stability - Processes, Machines, etc. 7
A Initial condition for establishing significant environmental aspects Miscellaneous Environmental Standards and EMS Related Discussions 1
J How is everyone else handling Machine Condition Standards? IATF 16949 - Automotive Quality Systems Standard 1
U Tolerancing Tool called 'Material Condition Modifier' created by Staneley Parker Misc. Quality Assurance and Business Systems Related Topics 0
S GD&T Rule 1: Gage size for form check at MMC (maximum material condition)? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
I Considering Mental Condition Related to Ill Health in Risk Assessment Occupational Health & Safety Management Standards 5
C Capability Study Required as a Condition to Adopt a Sampling Plan? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
T Automotive equipment sine wave vibration test condition - BMW requirement? Reliability Analysis - Predictions, Testing and Standards 4
W What condition is your company in, if you score 3.0? Lean in Manufacturing and Service Industries 12
U Nonnormal vs. Normal Operating Condition Aspect for each Environmental Aspect? Miscellaneous Environmental Standards and EMS Related Discussions 4
apestate Advice on maintaining surface plate condition - Starrett crystal pink surface plate Inspection, Prints (Drawings), Testing, Sampling and Related Topics 14
J Condition of Metrics prior to audit IATF 16949 - Automotive Quality Systems Standard 4
R Moving all of our SPC charts to an Electronic System - Software to flag OOC Condition Statistical Analysis Tools, Techniques and SPC 7