Conducted disturbances induced by RF fields test for battery-powered devices

HamidRNZ

Starting to get Involved
Hi everyone,

We have a battery-powered ECG logger with two electrodes to log the ECGs.
The electrodes are connected through wires to the central unit, which is battery-powered.
Is "Conducted disturbance induced by RF field" applicable to the cables?

Thanks in advance, and I look forward to hearing from you.

Ref:
  • IEC 60601-1-2, Table 7 "PATIENT coupling PORT"
  • IEC 60601-4-2, Table 5 "PATIENT COUPLING PORT"
 

Attachments

  • Conducted disturbances induced by RF fields test for battery-powered devices
    thumb_255_0_480_0_0_crop.jpg
    50.2 KB · Views: 17

Jaydub

Involved In Discussions
It seems like the ECG electrodes and leads would be a patient coupling port and therefore the test would be applicable.
 