Confusion between Software Medical Device and Health Software to do a QM
thisby_
Hello everybody! :)
I am currently helping a company to write the Quality Manual (kind of a stage). This company develops software to use in health care environments, a kind of EHR.
There is a particular part in the Quality Manual, the 7th part , regarding Design and development process particularly risk management and software safety classification which, according to the 13485 can be achieved by the 62304 part 5th (sw development process) that we are doing now.
We know that our product is not only a medical device, it can also be regular software. We read some standards about health software (HS), but is not clear for us, which kind of software can be considered so. I was hoping you could help me. (standards studied for HS: ISO/TS 29321:2008, ISO/TS 29322:2008, ISO/TR 27809:2007 and ISO/TS 25238:2007)
We are using the classification given by the 62304 regarding safety. We want to add classes to it, to insert regular software and maybe health software. The risk analysis for health software according to 29321 is almost the same as 14971, so before dive in this issue I really wanted to give some inputs about HS.
As you know, an EHR is an enormous box with so many lithe boxes inside (patient scheduling, patient information, office dashboard, scheduling, ADT, medical report, medical records, electronic prescription, billing…) all of them can be inserted inside the MD software, regular Sw or health software, so different assessments need to be made.
I appreciate all inputs!
Thank you in advanced,
Emilia
Marcelo
Re: Confusion between sofware medical device and health software to do a QM!
You seem to imply that your software is a medical device software and a healthcare software at the same time. If this is the case, i find this strange, because definitions and standards for medical device software and health software generally exclude one another.
Just using an example, standards for health software often say that these standards are not applicable to software which is:
* necessary for the proper application of a medical device or
* an accessory to a medical device or
* a medical device in its own right.
These last ones are medical device software.
Even if your software is really "both" (and it really can be both, depending on your intented use of the software) what i think you need to do is just this: define the intented use of the softwatre, including inside the system or other software, and them see which regulations/standards would apply to them. Maybe you will need to apply all regulations if the software fits in more than one type.
thisby_
Thank you Marcelo, I think I didn’t explain myself, I am sorry!
I wanted to say that, the software made by this company can be divided in sub products, some medical devices other maybe health software, other regular software, however we want the quality manual as general as possible to adapt to all this kind of products. The big problem is that according to the definition of Health Software is difficult to decide whether a software is health software or general software.
According to the definition of Health Sw from 29322:2008 is “software product for use in the health sector for health related purposes”. Maybe I have this problem because I can’t understand the meaning of health related purposes. Can anyone help me?
Thank you!
Emilia
Marcelo
I understand you being confused because this definition seems too broad.
You´re trying to differentiate between medical evice software, healthcare software and "general" software". This general software would be software used in the healthcare sector but not for health related porposes.
What is a health related purpose? I don´t know, and i´m almost sure there´s no definition in the standards for those. There could be definitions in regulations, or at least you could exclude based on the definitions for other purposes (for example, the medical devic case cited before).
So, trying to build a hierarchy, lets say that:
1 - if the software is necessary for the proper application of a medical device or an accessory to a medical device or a medical device in its own right, it´s a medical device software.
2 - if it´s used in the healthcare sector and is not a medical device software, but it´s intented use is somehow related to the "process" of healthcare (another way of thinking is that it could impact the health of the patient), for example, a health record (note that this can be tricky sometimes, but i do not see other way of defining it).
3 - if it´s used in the healthcare sector, does not fit in the two above (not a medical device and not used in the "process" of healthcare), it´s a general software. Exempl,es in my point of view, would be a word processing software used to record information not related to healthcare process (anything from patient home address to financial reports of a hospital), software used for controlling the secutiry system of a hospital or clinic, etc. Again, anything which might not impact the health of the patient.
Please note that, in this line of thought, any software which is part of another software/system which, in turn, fits in one of the porposes above, should be treated in the same category.
Well, this is just a little brainstorm on the subject, not sure if it helped.
bobbt
Thank you!
Emilia
I want to point out that and identify that there are two harmonized standards for software, what I have found out is if you decide to be compliant to the 2nd edition of 60601-1 then the applicable sofware
stanadard is en 60601-1-4 IF you are complying to the 3rd edition of 60601-1 then you need to comply with EN 62304 for software. Cenelec
has an excellent website about when collateral standards can be used
in a simple nut shell one has till sometime between 2011 AND JUNE 1 2012
to comply to the 3rd ediion of 60601-1 standard, also the harmonized stanadrds directive dated 7/15/09 it will show that both standards for software are ok to use the same applies to the usablity standards as well
sfurlani
I know you're in Italy, but perhaps this will be of help.
If a failure of the software would result in direct or indirect (through presenting the wrong information to the doctor) harm to the patient, then I would use the strictest of the standards. (FDA Class II Medical device)
Other software used in the healthcare industry (such as billing or scheduling) would be subject to other rules here in the States (HIPAA).
General software would likely be stuff like word editors.
Hope that helps,
-Stephen
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
B EMPB (Erstmusterprufbericht) VDA form confusion + PSW VDA Standards - Germany's Automotive Standards 2
S Some confusion about the creepage and clearance distance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S CE Mark - Classification Confusion EU Medical Device Regulations 12
A CE mark - Measuring Instruments Directive confusion! CE Marking (Conformité Européene) / CB Scheme 0
B Confusion on the new FMEA guidebook - Are we supposed to replace our FMEAs? IATF 16949 - Automotive Quality Systems Standard 3
Q Buy American Act - COTS confusion Manufacturing and Related Processes 7
M IAS (USA) Accreditation with IAF for ISO 13485 Logo Confusion EU Medical Device Regulations 10
K Design and Development Exemption/NA confusion Design and Development of Products and Processes 6
M Accept/Reject (Ac/Re) Numbers (ISO 2859-1) Confusion Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
supadrai Confusion on 510(k) Transfer Guidance - Device Listing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
A Informational Confusion about Risks for Processes in ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 44
I Gage R&R confusion on a part that has little variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
M "Single Patient Use" Terminology Confusion Other Medical Device Related Standards 9
M Health Canada Private Label Guidance Confusion - Quality System Required? Canada Medical Device Regulations 5
S Ionograph Testing - MIL-PRF-55110 Confusion AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 2
N Confusion within CE Mark and European Regulations CE Marking (Conformité Européene) / CB Scheme 5
C Clause 6.3 Infrastructure - Clearing the Confusion ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
T Acronym COP - Acronym Confusion Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 21
O Hazards vs. Hazardous Situation Confusion ISO 14971 - Medical Device Risk Management 11
SteveK EN ISO 13485:2012 vs. ISO 13485:2003 - Confusion? ISO 13485:2016 - Medical Device Quality Management Systems 25
AmandaMusser Documentation Confusion - Agile BOM Structure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
M Defining what is a Medical Device Accessory and what isn't - Much Confusion EU Medical Device Regulations 8
T Product Realization Confusion - Manufacturer Specifications Aspects and Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
B 510k exempt Class I Medical Device - Confusion during the facility registration US Food and Drug Administration (FDA) 3
D Confusion in writing a SIPOC (Supplier, Input, Process, Output, Customer) Six Sigma 5
S Confusion DPU, PPM & DPMO vs. Sigma Level - ASQ Six Sigma Black Belt hand book Six Sigma 14
Hershal Participation...........(Which Clause(s) cause you the most confusion or question?) General Measurement Device and Calibration Topics 6
P Confusion on declared (part of the quality manual) and non-declared documents Document Control Systems, Procedures, Forms and Templates 18
Marc Confusion Reigns As Analog TV Begins Shutdown After Work and Weekend Discussion Topics 2
Q Confusion over REACH Chemical List RoHS, REACH, ELV, IMDS and Restricted Substances 5
V How to use Sampling Plans - Confusion on Lot Sampling Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
I Six Sigma - Confusion while creating SIPOC diagram Six Sigma 6
G Health & Safety Executive - ISO 14001 and OHSAS 18001 Confusion Occupational Health & Safety Management Standards 6
T ISO/TS 16949:2002 Audit Confusion Internal Auditing 21
Manix WEEE and the confusion surrounding whether we fall into the scope of the directive! Other ISO and International Standards and European Regulations 9
Domoreto IATF Confusion - Our Registrar bought by another Registrar IATF 16949 - Automotive Quality Systems Standard 6
D Confusion regarding design responsibility - What can arguably be excluded? Design and Development of Products and Processes 10
R Confusion - Which tools are used and why on a daily basis? Quality Tools, Improvement and Analysis 3
M PPAP vs. Annual layout confusion - Customer "emergency PPAP" request APQP and PPAP 14
Randy Confusion and Castor Oil Coffee Break and Water Cooler Discussions 6
Manix Staging Process - Some confusion about what this stage actually is Manufacturing and Related Processes 10
S Control chart limit calculation confusion concerning SPC control limit calculation Statistical Analysis Tools, Techniques and SPC 27
F ISO 9001:2000 Confusion - Process Aspects: Interactions, Inputs, Outputs, Etc. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M Clause 4.2.3 Confusion: Documents that predate the doc control procedure. Document Control Systems, Procedures, Forms and Templates 17
S Numbering Confusion - BS EN ISO 9001:2000 vs. Other national versions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Q ISO Confusion - Do we comply and other questions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
C Is my software an accessory? Telecommunication between HCP and patients EU Medical Device Regulations 10
K Verify Software Architecture - supporting interfaces between items IEC 62304 - Medical Device Software Life Cycle Processes 1
E ASTM F2118 - Fatigue testing of bone cement - Changes between the 2003 and the 2014? Other Medical Device Related Standards 1