Contradictions in Medical Device Verification vs. Validation - ISO 13485

M

MikeL

#21
A valid point

I like the waterfall as it gives a good sense of the relationships between customer needs and validation.

I always thought that verification relates to meeting a spec and that validation is checking fitness for purpose. A spec is only our trying to quantify customer needs so the fact that our product meets specs is no guarantee that the customer will be happy.

Validation is often customer approval.

I am currently working with a software provider.

In very basic terms

customer wants a program for providing decision making info from financial data

we write a spec for a database with reports

<program furiously>

we check that all the routines are working properly (verification)

we demo the program to the customer and they are delighted with all the buttons and charts and pretty colours (validation)
 
Elsmar Forum Sponsor
B

BZGeorge

#22
Think of it this way.

Verification answers the question: Did I make the product correctly? (Correctly=Conformance to the established product specification or Process parameters). Testing/Inspection Related

Validation answers the question: Did I make the correct product. A Fitness for Use focus. Correct=Meets User Needs. Satisfies some identified User Need.
 

Weiner Dog

Med Device Consultant
#25
Hi. I worked as an FDA investigator (Level II Certified International Medical Device Investigator) for over 21 years- primarily in the medical device arena. Currently, I am a contract international medical device consultant.

I want to comment about process validation v. verification philosophies in general.

As an example, I will first compare and contrast process validation pertaining to pharmaceutical and medical device operations. I will also show the differences between process validation and verification. I believe these examples will clarify validation v. verification (a concept many people are confused).

Basically, certain operations (such as change control, process validation, and acceptance testing) are "black or white" in the pharmaceutical area (no matter which manufacturing site or process is involved), where similar operations can be "gray" in the medical device area (even from product to product or process to process at one site).

For example, in the pharmaceutical area, one has to validate a quality or manufacturing process (traditionally via a IQ/OQ/PQ prospective validation scheme)- be it a new process or changed process. However, in the medical device area, one has to first determine whether the new or changed process can be fully verified by subsequent inspection and test. The key word is "fully"- because certain tests will cause product destruction or not be cost beneficial. If the process cannot be fully verified, then the process has to be validated (either by prospective validation-IQ/OQ/PQ/PPQ [PPQ is a step not required by the pharmaceutical industry], retrospective validation, or concurrent validation).

Sterilization is a prime example. If one is to fully verify the sterilization operation, every unit will have to undergo sterilization testing. This is great quality-wise because this acceptance testing will show that each unit has passed or failed applicable firm specifications, but the firm will have no products to sell because each unit in each batch (i.e. 100% testing) will have to be opened and tested (thus compromising the sterile barrier).

Here are some basics regarding validation and verification

CONFUSING TERMS

Specification means any requirement with which a product, process, service, or other activity must conform.

Validation means confirmation by examination and provision of objective evidence that the particular requirements for a specific intended use can be consistently fulfilled.

Process Validation means establishing by objective evidence that a process consistently produces a result or product meeting its predetermined specifications.

Design Validation means establishing by objective evidence that device specifications conform with user needs and intended use(s).

Verification means confirmation by examination and provision of objective evidence that specified requirements have been fulfilled.

Process validation protocol: a document stating how validation will be conducted, including test parameters, product characteristics, manufacturing equipment, and decision points on what constitutes acceptable test results.

Qualification: establishing by objective evidence that a process produces a result or product meeting its predetermined requirements.

Installation qualification (IQ): establishing by objective evidence that all key aspects of the process equipment and ancillary system installation adhere to the manufacturer’s approved specification and that the recommendations of the supplier of the equipment are suitably considered.

Operational qualification (OQ): establishing by objective evidence process control limits and action levels which result in product that meets all predetermined requirements.

Performance qualification (PQ): establishing by objective evidence that the process, under anticipated conditions, consistently produces a product which meets all predetermined requirements.

Product Performance qualification (PPQ): establishing by objective evidence that the process does not adversely affect the product to meet all predetermined requirements.

OPERATIONS REQUIRING PROCESS VALIDATION OR VERIFICATION

(1) Processes which should be validated

Sterilization processes Clean room ambient conditions
Aseptic filling processes Sterile packaging sealing processes
Lypholization process Heat treating processes
Plating processes Plastic injection molding processes

(2) Processes which may be satisfactorily covered by verification

Manual cutting processes
Testing for color, turbidity, total pH for solutions
Visual inspection of printed circuit boards
Manufacturing and testing of wiring harnesses

(3) Processes which may need to be validated

Cleaning processes Certain human assembly processes
Numerical control cutting processes Filling processes

I hope this helps. :)
 

Weiner Dog

Med Device Consultant
#26
In an earlier post, I commented about validation and verification in general. Now, I'd like to comment about design validation and verification.

In simple terms, design verification answers the question- "Did I make the device right?"; whereas design validation answers the question- "Did I make the right device?".

Basically, design verification is similar to product acceptance testing, in that you are testing a product to see if it meets predetermined specifications. Design verification takes it one step further, because you have to make sure you have the correct specifications defined before the tests can be conducted.

For example, if your written specifications (think design output) state to make sure that the product is 5" in length and is red and you transfer this specification to production, you will be making and testing a product that is 5" in length and is red in color. However, what if the customer actually wanted you to design a product that is 6" in length and is purple in color? You would be making the wrong product. This is why you have to confirm that the intended uses and user needs (i.e. final design inputs) are contained in the written procedures, labeling, packaging, etc (i.e design outputs) before any verification testing can be conducted.]

On the other hand, design validation shows if the product has met user needs and intended uses (i.e. final design inputs). In this case, if you are able to design an automated swing (due to technology, etc.), whereas the customer actually wanted a tire on a rope tied to a tree, you have not made the product per the customer needs and intended uses.

I stress- design validation is not process validation. Process validation may need to be performed during the design process, but is not design validation. For process validation, you are seeing whether your process is always under control. During this process, you will be conducting various verification tests (using sampled products), taking into consideration worst case scenarios. For design validation, you may need to conduct clinical studies (i.e. having products being used under actual use conditions by outsiders), simulated use conditions, review other product labels/specs/design files, etc., or a combination of these various activities. You will not be sampling product to test against predetermined specifications under worst case conditions, but be conducting actual or simulated use tests, risk analyses, et al. instead.
 
T

tata347

#28
Thanks Geochaz,

In your professional opinion, from an FDA stand point.. is there a "requirement" that the protocol for a validation be givin a number / controlled prior to the validation (process) being completed?
For example: We are a contract MFG for a class II non sterile device (accessory) and do not perform final testing or packaging. We have generated a protocol and validation report witht he results; but our customer is suggesting that all of the sites perform the same validation & the protocol should have been numbered. We do number these two documents but only after they have been completed and accepted. the protocls are process in general but are specific to each site due to the differences in equipment (models).. We are working on a general procedure (SOP) previously we only had a design procedure that was under the control on Eng (not quality) until design phases were completed.

Well forum??? :frust::confused:
 

Weiner Dog

Med Device Consultant
#29
Hi.

First of all, who is responsible for approving the process validations at your site? Design validations are usually the responsibility of the spec developer/ packager/ labeler (i.e. your customer). What is stated in the contract?

I am assuming we are talking about prospective validation. If so, what exactly are you validating- a process or product? (I know that it is process, but how could the customer require each validation protocol to be the same for each contract site?) If the customer supplies you with specs- are they product specs (i.e. product drawings, etc.) or production equipment specs (i.e. such as computer programs for CNC machines)? Unless the customer worked with you (and all other contract sites) during the design project (assuming the product underwent design controls), how could your customer possibly know about the actual workings at your site (i.e. variances in personnel, equipment/calibration tolerances/accuracies, shift, etc.)? These differences have to be taken into effect when conducting a validation study. It is not just "let's make a widget per drawing X". Worst case scenarios.

Validation protocols have to be approved before any validation work takes place. The QSR does not have any requirement that validation protocols are numbered.
 
T

tata347

#30
Yes, I guess I left out a bit of information. We were previously a spec designer, MFG.. Now striclty a design MFG w/2 other sites reporting into us. We have a strong tie with this customer who we assist in the design of some of their products. Their previous vendor had a much better yeild but also apparently had some processes that our customer was unaware of and unable to transfer to us. We must also supply yeild data with each shipment, that includes failure breakdowns.

Our customer also has previously recieved a 483, I believe for controlling the subcontractor (us) and the ability to have current documentation. QA, Eng & Ops must approve protocols.. (in reality this happens mostly after the fact when I insist on the documentation.) I need to get a simple system in place that the Engineering documentation phobic staff can work with. The processes.. UV appication & cure, epoxy application, and belt furnance. Our protocol was specific to our machinery but we will forward the same protocol to the other MFG sites to obtain their results.
 
Thread starter Similar threads Forum Replies Date
A Contradictions in the reporting timelines of IVD Directive and MEDDEV 2.12-1 Rev 5 ISO 13485:2016 - Medical Device Quality Management Systems 2
S IAOB Sanctioned Interpretations - Duplications/contradictions - TS update/interps? IATF 16949 - Automotive Quality Systems Standard 18
R "Medical devices" required in scope ISO 13485:2016 - Medical Device Quality Management Systems 2
R X-RAY Based Diagnostic Veterinary medical Devices Medical Device and FDA Regulations and Standards News 1
Z Over The Air (OTA) updates for medical device Other US Medical Device Regulations 1
H Tukery Medical Device Regulstion Other Medical Device Regulations World-Wide 0
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 1
A Brexit Mandate for EU Authorised Representative for non medical devices CE Marking (Conformité Européene) / CB Scheme 3
A Canadian Safety Labels for medical electrical equipment Canada Medical Device Regulations 0
A Applicability of Photobiological Safety Evaluation for LED used in medical devices Reliability Analysis - Predictions, Testing and Standards 2
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
T Google Fit Health datas and Medical devices Application Medical Information Technology, Medical Software and Health Informatics 1
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Procedure packs with non-medical devices EU Medical Device Regulations 1
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
A Customer Approval (Medical Devices) Document Control Systems, Procedures, Forms and Templates 4
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 1
D Electrical Medical Devices class I EU Medical Device Regulations 0
J Are complaints applicable to development of medical devices? Customer Complaints 2
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
M Do I need separation in my circuit with a medical charger? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
P ISO 8 classified medical manufacturing room Qualification and Validation (including 21 CFR Part 11) 1
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Brian Benzarti Medical Devices CLP Regulation Quality Assurance and Compliance Software Tools and Solutions 2
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
J Medical Device Regulations in Lebanon? Other Medical Device Regulations World-Wide 2
dgrainger Informational Medicines and Medical Devices Act 2021 UK Medical Device Regulations 0
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
S Medical Device MRI Compatibility EU Medical Device Regulations 3
M NICE Medical Technology Evaluation Programme - Recommendations Service Industry Specific Topics 0
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7

Similar threads

Top Bottom