Hi. I worked as an FDA investigator (Level II Certified International Medical Device Investigator) for over 21 years- primarily in the medical device arena. Currently, I am a contract international medical device consultant.
I want to comment about process validation v. verification philosophies in general.
As an example, I will first compare and contrast process validation pertaining to pharmaceutical and medical device operations. I will also show the differences between process validation and verification. I believe these examples will clarify validation v. verification (a concept many people are confused).
Basically, certain operations (such as change control, process validation, and acceptance testing) are "black or white" in the pharmaceutical area (no matter which manufacturing site or process is involved), where similar operations can be "gray" in the medical device area (even from product to product or process to process at one site).
For example, in the pharmaceutical area, one has to validate a quality or manufacturing process (traditionally via a IQ/OQ/PQ prospective validation scheme)- be it a new process or changed process. However, in the medical device area, one has to first determine whether the new or changed process can be fully verified by subsequent inspection and test. The key word is "fully"- because certain tests will cause product destruction or not be cost beneficial. If the process cannot be fully verified, then the process has to be validated (either by prospective validation-IQ/OQ/PQ/PPQ [PPQ is a step not required by the pharmaceutical industry], retrospective validation, or concurrent validation).
Sterilization is a prime example. If one is to fully verify the sterilization operation, every unit will have to undergo sterilization testing. This is great quality-wise because this acceptance testing will show that each unit has passed or failed applicable firm specifications, but the firm will have no products to sell because each unit in each batch (i.e. 100% testing) will have to be opened and tested (thus compromising the sterile barrier).
Here are some basics regarding validation and verification
CONFUSING TERMS
Specification means any requirement with which a product, process, service, or other activity must conform.
Validation means confirmation by examination and provision of objective evidence that the particular requirements for a specific intended use can be consistently fulfilled.
Process Validation means establishing by objective evidence that a process consistently produces a result or product meeting its predetermined specifications.
Design Validation means establishing by objective evidence that device specifications conform with user needs and intended use(s).
Verification means confirmation by examination and provision of objective evidence that specified requirements have been fulfilled.
Process validation protocol: a document stating how validation will be conducted, including test parameters, product characteristics, manufacturing equipment, and decision points on what constitutes acceptable test results.
Qualification: establishing by objective evidence that a process produces a result or product meeting its predetermined requirements.
Installation qualification (IQ): establishing by objective evidence that all key aspects of the process equipment and ancillary system installation adhere to the manufacturer’s approved specification and that the recommendations of the supplier of the equipment are suitably considered.
Operational qualification (OQ): establishing by objective evidence process control limits and action levels which result in product that meets all predetermined requirements.
Performance qualification (PQ): establishing by objective evidence that the process, under anticipated conditions, consistently produces a product which meets all predetermined requirements.
Product Performance qualification (PPQ): establishing by objective evidence that the process does not adversely affect the product to meet all predetermined requirements.
OPERATIONS REQUIRING PROCESS VALIDATION OR VERIFICATION
(1) Processes which should be validated
Sterilization processes Clean room ambient conditions
Aseptic filling processes Sterile packaging sealing processes
Lypholization process Heat treating processes
Plating processes Plastic injection molding processes
(2) Processes which may be satisfactorily covered by verification
Manual cutting processes
Testing for color, turbidity, total pH for solutions
Visual inspection of printed circuit boards
Manufacturing and testing of wiring harnesses
(3) Processes which may need to be validated
Cleaning processes Certain human assembly processes
Numerical control cutting processes Filling processes
I hope this helps.