Control Charts
Emine
Hello everyone,
We are ISO 17025 accredited testing laboratory. We create control charts and monitor them regularly. I have just realized that the size of the points are very variable in different test methods (some labs calculating 60 points, other 25 points etc). Are there an optimum number of points in creating control charts that will give a correct statistical analysis?
Thank you very much, any thoughts are very much appreciated,
Emine
When you say the "size of the points", do you mean the subgroup size or the number of subgroups used to calculate control limits?
Emine
Hi Miner,
Thank you for your response, the points representing averages of measurements of QC s (CRM) that are run together with the samples. And as you have mentioned we calculate the control limits from standard deviation. We are a bit confused on the optimum points should be used to calculate warning/action limits, I realized that sometimes when there are a lot of QC measurements are included the control limits are wider...
Many Thanks,
Emine
Emine
And what are you plotting?
Hi BD,
Thank you for your reply,
We are plotting our QC measurements from day to day analysis in a time sequence. And we calculate average and action and warning limits from the standard deviation.
Emine
Hi Miner,
Many Thanks,
Emine
I am interpreting this to mean that it is your subgroup size that varies from 25-60. This is probably too large and is making your chart overly sensitive to small process changes. The subgroup size balances the risks of type 1 and type 2 errors. Too small a subgroup size and you risk missing a process change, and too large a subgroup size and you see changes that are too small to be practical. Most subgroup sizes should fall between 1 and 9.
The linked article contains a formula for calculating an appropriate size, but you can also determine an appropriate size experimentally. Start with a midrange size such as 5. Is it responding to the size of process shift that you need? If not, increase it to 6 and try again. Likewise, if it is too sensitive, reduce it to 4 and try again.
Your control limits should get tighter with larger subgroup sizes. Your experience makes me suspect that you do not have rational subgroups (i.e., you may have mixtures, or or other issues).
If you are plotting QC results then your subgroup size should be 1. Two articles that cover SPC for QC results are: Link to article 1 and link to article 2
if your situation is not what I think it is please elaborate.
Steve Prevette
Deming Disciple
I usually worked with keeping a minimum of 25 points displayed on any control chart, which was based upon Dr. Shewhart's original declaration of do not declare a process stable without 25 stable points. The rule of thumb always worked for me. Visually, I usually ended up keeping the past two years plus the current year on the chart of monthly data.
PQ Systems Understanding Capability and Control Charts Series Using SQCpack Software 0
PQ Systems Control Charts & Capability Analysis 101 Using SQCpack Software 0
D How to calculate the A4 constants in Median/Range control charts? Statistical Analysis Tools, Techniques and SPC 2
S Use of Attributes Control Charts Statistical Analysis Tools, Techniques and SPC 1
U Electronic templates for making paper control charts Reliability Analysis - Predictions, Testing and Standards 2
Rameshwar25 Other types of Control Charts described in Chapter II of SPC Manual Statistical Analysis Tools, Techniques and SPC 6
S Standard Deviation Selection on Control Charts - Minitab "pooled deviation" Statistical Analysis Tools, Techniques and SPC 3
N Injection Molding - Median individual control charts Statistical Analysis Tools, Techniques and SPC 6
Chennaiite Do Attribute Control Charts really help? Statistical Analysis Tools, Techniques and SPC 9
B Setting Lower Control Limit on Attributes Charts Statistical Analysis Tools, Techniques and SPC 9
M Interpreting X bar and R Control Charts Statistical Analysis Tools, Techniques and SPC 2
M Large Sample Size - up to 32 - for Control Charts Capability, Accuracy and Stability - Processes, Machines, etc. 2
T Is the use of Control Charts Mandatory or Optional? ISO 17025 Requirements ISO 17025 related Discussions 6
E Minitab for hourly patient census - Control Charts Using Minitab Software 6
J Using P-Charts for Glass Manufacturing to Monitor and Control Attributes Statistical Analysis Tools, Techniques and SPC 2
M On Subgroups in Control Charts and Process Capability Statistical Analysis Tools, Techniques and SPC 3
S Help with Table of Constants and Formulas for Control Charts Statistical Analysis Tools, Techniques and SPC 5
Q Monthly Data Review through Control Charts - Monitoring Complaints ISO 13485:2016 - Medical Device Quality Management Systems 3
R "Control" as used in 4.1 c) - Process Flow Charts (Process Sequence Chart) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Suitability of Attribute Control Charts for ppm Level Control Statistical Analysis Tools, Techniques and SPC 6
R Metrology Control Charts for Reference and Working Standards General Measurement Device and Calibration Topics 6
D One-Two-Three Sigma Control Limits - Calculating Control Limits for X-bar Charts Statistical Analysis Tools, Techniques and SPC 3
bobdoering Myth or Mythunderstanding: Implications of the Economic Design of Control Charts Statistical Analysis Tools, Techniques and SPC 1
E Performance Charts - Purchasing Department and Material Control Department Preventive Action and Continuous Improvement 6
J Control Charts for 29 Devices that I tested Reliability Analysis - Predictions, Testing and Standards 10
J Control Charts with different Sample Sizes Reliability Analysis - Predictions, Testing and Standards 2
T why CPK <1.0 regarding two pieces of control charts of torque wrench? Statistical Analysis Tools, Techniques and SPC 2
R Electronic Templates for making Paper Control Charts Document Control Systems, Procedures, Forms and Templates 9
S Multivariate Control Charts in Chemical Batch Processing Statistical Analysis Tools, Techniques and SPC 7
G Control Charts - Plotting Unwanted Data Excel .xls Spreadsheet Templates and Tools 5
Y Control Charts + Ishikawa Manufacturing and Related Processes 26
N SPC Light Program - Easy way of Cpk extraction from multiple control charts Statistical Analysis Tools, Techniques and SPC 4
K Calculating Control Limits in Average and Range Charts Statistical Analysis Tools, Techniques and SPC 3
G Adjusted Control Limits - Usage of Control Charts Statistical Analysis Tools, Techniques and SPC 26
U P Charts - Subgroup Sizes? Process Control Indicator Statistical Analysis Tools, Techniques and SPC 3
E Which control charts to monitor waste produced on the shop floor Statistical Analysis Tools, Techniques and SPC 5
J Control limits for one sided control charts using WECO rules Statistical Analysis Tools, Techniques and SPC 5
T Rules for interpreting control charts Statistical Analysis Tools, Techniques and SPC 2
G Control charts of finished product characteristics..NonSense? Statistical Analysis Tools, Techniques and SPC 8
D Control limit calculation (Xbar R charts) using Minitab Statistical Analysis Tools, Techniques and SPC 14
H How to create Stoplight Control charts with excel? Excel .xls Spreadsheet Templates and Tools 16
S Calculating 1 sigma and 2 Sigma limits in Control Charts Statistical Analysis Tools, Techniques and SPC 1
M Control Charts - Points TOUCHING Limits = Stable? Statistical Analysis Tools, Techniques and SPC 12
T What is Average Run Length & Average Production Length when used in Control Charts? Statistical Analysis Tools, Techniques and SPC 4
T Using SPC Control Charts to Reduce AQL Sample Size AQL - Acceptable Quality Level 16
B Recommendation for a book to Explain Control Charts Book, Video, Blog and Web Site Reviews and Recommendations 6
C Designing Control Charts - Variable with Small Variation and Tight Control Limit Statistical Analysis Tools, Techniques and SPC 20
K Understanding Control Charts Methodology for Attributes like Plug Guages. Statistical Analysis Tools, Techniques and SPC 1
J Selection Procedure for the Use of Control Charts (Apendix C) Statistical Analysis Tools, Techniques and SPC 2
B Reaction to Out of Control SPC Charts Statistical Analysis Tools, Techniques and SPC 14