Greetings,

I have a few questions pertaining to the control over R&D parts/equipment.

Currently, we state in our Design Control procedure that units built for V&V must use the intended manufacturing processes and should have DHRs. We'd like to not require part inventory and lot histories before V&V. Is this acceptable and why?

For suppliers who provide us with R&D parts do they have to be on the approved supplier list if those parts won't be part of design V&V?

Basically, how much control over equipment is required if not used for design V&V or assessing quality of finished devices?

Any assistance/feedback would be most appreciated.

Tonia
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Might be helpful if we split up R and D. On the research side - which it sounds like maybe you're talking about - the only controls are up to you. If you're really into design, for what purpose are you building the items (and purchasing parts, etc.)?
 

tebusse

Involved In Discussions
#3
The purpose of building items is for R&D to determine feasibility and compatibility. Is it a regulatory requirement that we control R&D equipment? As previously stated, we'd continue to control any equipment used for design verification and validation.

Thank you.
 

JeantheBigone

Quite Involved in Discussions
#4
I agree with yodon. Especially if you are talking about medical devices to be approved / cleared by FDA, "R" and "D" are two completely different worlds.

FDA does not regulate research, but it certainly does have something to say about design.

I think it is up to your company to have a procedure in place to decide when a research project has resulted in the recognition that a device is to be developed. Once that decision is made, design controls apply, but not before.
 
Thread starter Similar threads Forum Replies Date
D Minimum Size of Control Sample in Medical Research Lab Work Statistical Analysis Tools, Techniques and SPC 2
D GAMP CAT 5 Change Control Qualification and Validation (including 21 CFR Part 11) 9
D Use Error Risk Controls and Control Verification ISO 14971 - Medical Device Risk Management 5
S Maintenence related action in Control plan FMEA and Control Plans 3
M ISO 13485 and document control for programs ISO 13485:2016 - Medical Device Quality Management Systems 6
R Control Methods in Control Plan FMEA and Control Plans 4
M Design Control - Management of Electronic CAD files Design and Development of Products and Processes 11
S TSCA (Toxic Substances Control Act) PIP 3:1 RoHS, REACH, ELV, IMDS and Restricted Substances 0
S Use of Attributes Control Charts Statistical Analysis Tools, Techniques and SPC 1
U Appraiser Out of Control Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
L Looking for Control Plan and FMEA Stamping process FMEA and Control Plans 2
Z AS9100 Control of Nonconforming Outputs - Rework Dispositions AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
K ASME, ANSI, ASTM and similar specifications and their requirements for document control. Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
Z Software for design control ISO 13485:2016 - Medical Device Quality Management Systems 5
K Quality Control/Assurance in the Metal Industry Quality Tools, Improvement and Analysis 10
K Defining risk control measures IEC 62304 - Medical Device Software Life Cycle Processes 14
J Warehouse temperature and humidity control EU Medical Device Regulations 7
Jimmy123 IATF16949 Control Plan FMEA and Control Plans 1
D Control Number for Class III and IV devices Canada Medical Device Regulations 1
J Document Control Metrics Document Control Systems, Procedures, Forms and Templates 3
I Document Control Software Document Control Systems, Procedures, Forms and Templates 2
T Controlling Expandable Forms in Paper-Based Document Control System Document Control Systems, Procedures, Forms and Templates 10
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
C UDI Questions on Control Units Other US Medical Device Regulations 6
A Quality Control Datasheets Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Control-self assessment Internal Auditing 5
GreatNate Master Control QMS software Quality Tools, Improvement and Analysis 1
A Prototype control plan FMEA and Control Plans 2
J Records Control - Does each individual record need to be numbered? Records and Data - Quality, Legal and Other Evidence 2
shimonv Document Control Procedure Header Content Document Control Systems, Procedures, Forms and Templates 4
W FMEA - Current control and occurrence rating FMEA and Control Plans 3
S Which department should prepare the control plan? could you show me a standard regarding to this matter. FMEA and Control Plans 17
G Control Plan & PFMEA Review Procedure? FMEA and Control Plans 8
N What are the software audit and control steps Reliability Analysis - Predictions, Testing and Standards 2
A Fabric roll inspection - What type of Control Chart to use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
C Monitoring and Control Instruments RoHS, REACH, ELV, IMDS and Restricted Substances 3
M ISO14971:2019 - Verification of implementation and effectiveness of risk control ISO 14971 - Medical Device Risk Management 3
T Risks arising from control measures vs. ineffective control measures ISO 14971 - Medical Device Risk Management 11
J Need Change Control Yes/No Decision Tree Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
P IQ, OQ, PQ protocol, and report templates for Distributed Control Systems in Pharmaceutical Manufacturing Systems. Qualification and Validation (including 21 CFR Part 11) 0
R Risk control measures as per ISO 14971 ISO 14971 - Medical Device Risk Management 6
T Linking Control Plans and PFMEA's FMEA and Control Plans 3
L How to add exemption or statement to control of document procedure? ISO 13485:2016 - Medical Device Quality Management Systems 5
T PFMEA and Control Plans on legacy product FMEA and Control Plans 5
P IATF 16949 requirement - error-proofing in control plan IATF 16949 - Automotive Quality Systems Standard 3
C 8.5.1.1 Control of Equipment, Tools, and Software Programs - Questions about the extent of control of NC programs AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
Q Version/Revision Control of CAD files Document Control Systems, Procedures, Forms and Templates 2
earl62 What is the best way to control special characteristics in Control plan? Is it Mandatory to have SPC for IATF 16949? IATF 16949 - Automotive Quality Systems Standard 12
I Control Plan (Product/Process specification/ Tolerance) acceptance FMEA and Control Plans 27
Sravan Manchikanti How to interpret '8.3 Control of nonconforming product' for SaMD device while implementing ISO 13485 & MDSAP ISO 13485:2016 - Medical Device Quality Management Systems 7

Similar threads

Top Bottom