Converting ETO Residual Limits from PPM to mg
Nash27
Involved In Discussions
#1
Hi Could someone help me converting ETO residual limits from PPM to mg please. Our overseas contract manufacturer (CM) utilises IS10150 (Indian Standard) that provides EO, ECH, EG limits in PPM. Where as ISO 10993-7: 2008 list the quantities in mg. Our CM states that Indian standard is more stringent and listed below rationale;
EO ECH EG
AS PER IS 10150 PPM 25 25 250
AS PER ISO 10993-7 MG 60 60 NA
EXTRACT 5 ML 12 12
0.1 % MG 1.2 1.2
MCG 1200 1200
PPM 1200 1200
i.e. the Indian standard far more stringent than ISO10993-7: 2008. However, I am unable to understand the logic. The questioned device is class IIa simple IV set that come in contact with patient for Prolonged Exposure (>24 hours to <30 days) hence the ISO 10993-7: 2008 limit taken as 60mg.
Any help would be much appreciated. Thanks
Elsmar Forum Sponsor
planB
Super Moderator
#2
Re: ETO Residual PPM vs mg
Nash27,
the limits listed in ISO 10993-7:2008, section 4.3 (and specifically section 4.3.3 for your classification) are absolute masses per device and applicable time period.
Values in ppm, i.e. parts per million, are ratios, and are 1-dimensional,such as mass per mass or volume per volume or sometimes more losely mass/volume.
Moreover, googling for "IS 10250" seems to return a standard applicable for the automotive industry, not for medical devices.
I would strongly recommend sticking to ISO 10993-7, which is applicable for medical devices. At least on an international level, you will have a hard time defending not complying to this standard.
And note that _all_ of these limits apply to a prolonged exposure device:
4.3.3 Prolonged exposure devices
The average daily dose of EO to patient shall not exceed 2 mg/d. In addition, the maximum EO dose shall
not exceed:
⎯ 4 mg in the first 24 h;
⎯ 60 mg in the first 30 d.
The average daily dose of ECH to patient shall not exceed 2 mg/d. In addition, the maximum ECH dose
shall not exceed:
⎯ 9 mg in the first 24 h;
4.3.5.2 Tolerable contact limit for EO
Either the EO TCL for surface contacting devices and implants shall not exceed 10 μg/cm2 or it shall
exhibit negligible irritation as specified in ISO 10993-10.
4.3.5.3 Tolerable contact limit for ECH for surface contacting devices
Either the ECH TCL for surface contacting devices and implants shall not exceed 5 mg/cm2 or it shall
HTH,
Nash27
#3
Dear HHT,
My apologies the standard I have quoted is IS10150: 1981 (reaffirmed 1996) see attached. I am referring to the table 4 of the section Appendix B-5. I have attached the same standard here
(Deleted)
I hope this might help to review my query again. Thanks
planB
Super Moderator
#5
Nash27,
IS10150:1981, App. B-5, table 4 only list one single requirement for EO and ECH, respectively. Thus, this standard cannot be more stringent than ISO10993-7:2008, which lists multiple requirements for limits (see my first posting).
Moreover, I think IS10150:1981 is not state-of-the-art anymore. Similar sterilization-related ISO standards have undergone multiple revisions in the past 35 years (i.e. the age of this Indian standard).
ISO 11135:2014 and ISO 10993-7:2008 (the latter being currently amended) are the standards that are internationally regarded as state of the art.
HTH,
Gerhard
Nash27
#6
Dear Gerhard,
Pity on me for not extrapolating HHT ..is happy to help..I remember a joke, during the university studies we were trying to help one of our friends to buy a second hand car..going through the car advertisement he found a car within his budget... he said he is going to call, so we said ok go on.. he called and asked 'Hi, can I talk with Datsun'...there was a pause from other end ..and came reply.. mate do you want to talk with me or my car... I am selling my car which is a Datsun brand...we laughed as he was unaware, that the Datsun is a car....
I will remember HHT is not a name.
We have worked around the requirements, as our CM complies with requirements of ISO 11135 and ISO 10993-7 with an exception of the results mentioned in format that complies to IS standard. The EO extraction method, chromatogarphic analysis meets (performed by Lab) the ISO10993-7 requirements. I just needed to show the result relevance which I have got now.
Thanks for your help.
planB
Super Moderator
#7
Nash27,
no problem. Make sure that the lab report states compliance to ISO 10993-7 for the used analysis method. And best of all, the lab would be accredited for this type of analysis. Then you will have an easier time defending the validity of your test results.
HTH (= hope this helps ;) ),
Gerhard
Thread starter Similar threads Forum Replies Date
S Converting SOP MDD to MDR Noob ISO 13485:2016 - Medical Device Quality Management Systems 5
Q Converting Manufacturing Drawings from a Contract Supplier ISO 13485:2016 - Medical Device Quality Management Systems 7
P Converting DMF III into eCTD (Electronic Common Technical Document) Medical Information Technology, Medical Software and Health Informatics 4
T Converting an ISO Class 8 Cleanroom to ISO 7 Various Other Specifications, Standards, and related Requirements 4
N Converting Paper Records to Electronic Format Records Document Control Systems, Procedures, Forms and Templates 6
V Converting Newtons to Newton Meters APQP and PPAP 4
optomist1 Converting Defect Rate To Capability index Capability, Accuracy and Stability - Processes, Machines, etc. 11
B Conversion and Rounding Standard for converting metric to english and vice versa Other ISO and International Standards and European Regulations 4
A Converting 1-Sided Design Tolerance to a Bilateral Tolerance Statistical Analysis Tools, Techniques and SPC 8
D Converting to Electronic Document Control - Procedure Content ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
Stijloor Converting a Bear... Funny Stuff - Jokes and Humour 0
optomist1 ASME GD&T Certification Test - Converting Coordinate Tolerances to Profile Tolerances Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
J Converting SOUP to "SOKP" ... IEC 62304 - Medical Device Software Life Cycle Processes 1
M Converting PowerPoint Presentations into MS Word Documents After Work and Weekend Discussion Topics 6
N Thermocouple Calibration - Digital Display vs. Converting the Millivolt into ?C General Measurement Device and Calibration Topics 8
R Converting Best Measurement (BMC) to Calibration & Measurement Capability (CMC) Nonconformance and Corrective Action 2
A Basic Math training - I.e. fractions converting from inches to mm, etc. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
I Converting mG(milli g's of acceleration) to dB(deciBel) Misc. Quality Assurance and Business Systems Related Topics 20
J Converting solid model formats for a Calypso CMM Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
Q Curve Surface Roughness Measurement & Converting Ra to Rz General Measurement Device and Calibration Topics 8
Sidney Vianna Converting PowerPoint presentations into video - software recommendations? After Work and Weekend Discussion Topics 12
A CMM Software Conversion - Converting from Zeiss/Calypso to Brown & sharpe PCDMIS Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
ScottK Learning about converting an AutoCAD .dwg to a .STEP file Manufacturing and Related Processes 5
B Converting from QS9000 to TS 16949 ... HOW? WHO? HOW MUCH? IATF 16949 - Automotive Quality Systems Standard 2
J Converting torque to pressure - One for the Math Folks Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
M Converting ISO9001:1994 to new ISO9001:2000 - Transition ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
C Converting to Process Audits - Seeking: Process Audit Form layout Process Audits and Layered Process Audits 4
J Table for converting Cp values into corresponding Capability % Statistical Analysis Tools, Techniques and SPC 1
J Converting DPMO to SIGMA - Seeking Forumula Statistical Analysis Tools, Techniques and SPC 16
B Re-use of ETO Sterilized products Medical Device and FDA Regulations and Standards News 6
B ETO Annual review ISO 13485:2016 - Medical Device Quality Management Systems 1
R ETO Sterilization EU Medical Device Regulations 7
shimonv Frequency and sample size for lot release of EtO sterilized products EU Medical Device Regulations 2
chris1price EtO Sterilisation Consultant Wanted Job Openings, Consulting and Employment Opportunities 0
C ETO Sterilised Class II Medical Device - Required Temperature Storage ISO 13485:2016 - Medical Device Quality Management Systems 1
A Is Aeration Process in ETO Sterilization Validation Required? Manufacturing and Related Processes 7
N ETO Sterilisation Validation - EO Residual Minimum Sample Requirement ISO 13485:2016 - Medical Device Quality Management Systems 2
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
M Is it essential to do the Fractional Test for ETO Revalidation? ISO 13485:2016 - Medical Device Quality Management Systems 1
K Re-validation of EtO sterilization after change of tape Other Medical Device and Orthopedic Related Topics 1
N Decision to file a new 510(k) - ETO (Ethylene Oxide) Sterilization Other US Medical Device Regulations 4
C ETO (Ethylene Oxide) Sterilization Audit Question Other Medical Device Related Standards 9
A How to overcome brown stains after ETO sterilization process ? Other Medical Device Related Standards 2
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
S Fractional cycle fail in Eto sterilization validation ISO 13485:2016 - Medical Device Quality Management Systems 3
R Eto Sterilization Product Adoption SOP ISO 13485:2016 - Medical Device Quality Management Systems 4
M EtO compatible or recommended materials Other Medical Device Regulations World-Wide 1
X EtO Steriliser Software (PACS) Update Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
S IV Catheter Biological Evaluation - EtO Residuals - Permanent Contact Definition Qualification and Validation (including 21 CFR Part 11) 2
S ETO ISO 11135-1 & ISO 17665-2 - Purity and Quality of Steam (Humidity) Qualification and Validation (including 21 CFR Part 11) 3
Similar threads
Top Bottom