Correlating Maximum 'Range' of Values for a given "RSD" (Control Limits)

v9991

Trusted Information Resource
for setting a control limit of 'relative standard deviation',

we want to evaluate the maximum possible range for given rsd-1 and rsd-2; (viz., rsd 2% vs 3% vs 6%)

request help in calculating 'max possible range' for RSD of 2-3-6.
 

Bev D

Heretical Statistician
Leader
Super Moderator
By RSD do you really mean what is commonly referred to as CV?
CV is the coefficient of variation = SD/Average.

If not can you please clarify?

In any case wouldn't you just plot the average and the standard deviation on a traditional Xbar, S chart? In my experience the closer you are to the true data the better the insight. The more you manipulate (transform) the data the less insight you get...
 

v9991

Trusted Information Resource
Correlating Maximum 'Range' of Values for a given "RSD" relative standard devatio

Yes please, I meant RSD / Coefficient of variation.(i must have expanded the acronym...)

in our case, we are trying to set a limit for RSD for setting up a machine;

we want to understand the maximum range resulting from a given RSD value.


I also noted an usefull article on RSD
 

v9991

Trusted Information Resource
1010 ANALYTICAL DATA—INTERPRETATION AND TREATMENT
1010 ANALYTICAL DATA—INTERPRETATION AND TREATMENT
Determination of Sample Size
Sample size determination is based on the comparison of the accuracy and precision of the two methods4 and is similar to that for testing hypotheses about average differences in the former case and variance ratios in the latter case, but the meaning of some of the input is different. The first component to be specified is , the largest acceptable difference between the two methods that, if achieved, still leads to the conclusion of equivalence. That is, if the two methods differ by no more than , they are considered acceptably similar. The comparison can be two-sided as just expressed, considering a difference of in either direction, as would be used when comparing means. Alternatively, it can be one-sided as in the case of comparing variances where a decrease in variability is acceptable and equivalency is concluded if the ratio of the variances (new/current, as a proportion) is not more than 1.0 + . A researcher will need to state based on knowledge of the current method and/or its use, or it may be calculated. One consideration, when there are specifications to satisfy, is that the new method should not differ by so much from the current method as to risk generating out-of-specification results. One then chooses to have a low likelihood of this happening by, for example, comparing the distribution of data for the current method to the specification limits. This could be done graphically or by using a tolerance interval, an example of which is given in Appendix E. In general, the choice for must depend on the scientific requirements of the laboratory.

The next two components relate to the probability of error. The data could lead to a conclusion of similarity when the methods are unacceptably different (as defined by ). This is called a false positive or Type I error. The error could also be in the other direction; that is, the methods could be similar, but the data do not permit that conclusion. This is a false negative or Type II error. With statistical methods, it is not possible to completely eliminate the possibility of either error. However, by choosing the sample size appropriately, the probability of each of these errors can be made acceptably small. The acceptable maximum probability of a Type I error is commonly denoted as and is commonly taken as 5%, but may be chosen differently. The desired maximum probability of a Type II error is commonly denoted by . Often, is specified indirectly by choosing a desired level of 1 – , which is called the “power” of the test. In the context of equivalency testing, power is the probability of correctly concluding that two methods are equivalent. Power is commonly taken to be 80% or 90% (corresponding to a of 20% or 10%), though other values may be chosen. The protocol for the experiment should specify , , and power. The sample size will depend on all of these components. An example is given in Appendix E. Although Appendix E determines only a single value, it is often useful to determine a table of sample sizes corresponding to different choices of , , and power. Such a table often allows for a more informed choice of sample size to better balance the competing priorities of resources and risks (false negative and false positive conclusions).

Sample Size
Formulas are available that can be used for a specified , under the assumption that the population variances are known and equal, to calculate the number of samples required to be tested per method, n. The level of confidence and power must also be specified. [NOTE—Power refers to the probability of correctly concluding that two identical methods are equivalent.] For example, if = 4.7, and the two population variances are assumed to equal 4.0, then, for a 5% level test9 and 80% power (with associated z-values of 1.645 and 1.282, repectively), the sample size is approximated by the following formula:
Click to View Image
Click to View Image
Thus, assuming each method has a population variance, 2, of 4.0, the number of samples, n, required to conclude with 80% probability that the two methods are equivalent (90% confidence interval for the difference in the true means falls between –4.7 and +4.7) when in fact they are identical (the true mean difference is zero) is 4. Because the normal distribution was used in the above formula, 4 is actually a lower bound on the needed sample size. If feasible, one might want to use a larger sample size. Values for z for common confidence levels are presented in Table 8. The formula above makes three assumptions: 1) the variance used in the sample size calculation is based on a sufficiently large amount of prior data to be treated as known; 2) the prior known variance will be used in the analysis of the new experiment, or the sample size for the new experiment is sufficiently large so that the normal distribution is a good approximation to the t distribution; and 3) the laboratory is confident that there is no actual difference in the means, the most optimistic case. It is not common for all three of these assumptions to hold. The formula above should be treated most often as an initial approximation. Deviations from the three assumptions will lead to a larger required sample size. In general, we recommend seeking assistance from someone familiar with the necessary methods.
When a log transformation is required to achieve normality, the sample size formula needs to be slightly adjusted as shown below. Instead of formulating the problem in terms of the population variance and the largest acceptable difference, , between the two methods, it now is formulated in terms of the population RSD and the largest acceptable proportional difference between the two methods.
Click to View Image
where
Click to View Image
Click to View Image
and represents the largest acceptable proportional difference between the two methods ((alternative-current)/current) and the population RSDs are assumed known and equal.
 
Thread starter Similar threads Forum Replies Date
A Correlating Hazard Analysis and DFMEA Risk Acceptability Criteria FMEA and Control Plans 8
F XRF Capabilities - Correlating 2 different XRF machines - Ni/Au parts Capability, Accuracy and Stability - Processes, Machines, etc. 7
M IEC 60601-1-8 clause 6.4.1 (statistics of the distribution of the sum of the maximum ALARM CONDITION + SIGNAL GENERATION DELAY) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC Maximum Power Density for Muscle Stimulator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
P Maximum Mains Voltage (clause 8.5.3) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Maximum mains voltage used when applying 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
D Determining the the maximum number of reprocessing cycles of attachments CE Marking (Conformité Européene) / CB Scheme 2
Marco Bernardi MMC & LMC modifiers and CMM measuring techniques like diameter least squares and circularity or minimum/ maximum diameter Calibration and Metrology Software and Hardware 5
G Need to change KPI we called NC parts (maximum 3%.) to FTQ (first time quality) IATF 16949 - Automotive Quality Systems Standard 4
Judy Abbott Increase the maximum allowable dose for E-beam sterilization Reliability Analysis - Predictions, Testing and Standards 2
M Maximum Height for the stored material or stock materials Occupational Health & Safety Management Standards 1
S How to calculate the maximum allowable hysteresis of a pressure gauge General Measurement Device and Calibration Topics 1
F Irradiation Sterilization Validation Maximum Dose - VDmax 25 method (ISO 11137) EU Medical Device Regulations 3
V Acceptable maximum RSD (relative standard deviation) for an sample size Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Marc Definition MPEE - Maximum Permissible Error for Length Measurement Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 0
A Maximum bubble volume for subcutaneous injection IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D Calibration of a Width Gage Slot with an Attribute Gage - Maximum Width Condition General Measurement Device and Calibration Topics 16
B Maximum Material Condition with Bonus - Determine Cpk for this one kpc Measurement Uncertainty (MU) 2
M What is the maximum permissible period of delaying the validation of equipment? Other Medical Device Regulations World-Wide 4
C Q as the Datum Maximum Material Condition then MINUS 0.020 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
I Minimum and Maximum Validation Batch Size Requirements Qualification and Validation (including 21 CFR Part 11) 3
R Maximum Reverse Speed Limit for Powered Wheelchairs Other US Medical Device Regulations 2
C Maximum acceptable Tolerance - Uncertainty Ratio (TUR) General Measurement Device and Calibration Topics 2
R ISO 14644-1 maximum particulate concentration limits concern Other ISO and International Standards and European Regulations 2
K Maximum Material Condition on Diameter Tolerance Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
L Maximum voltage for IEC 60601? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
M Zero Tolerancing on True Position using MMC (Maximum Material Condition) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
K Step by Step Gauge R&R with 10% maximum allowable Gauge Error Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S Aseptic Processing - Maximum SAL using Aseptic Processing for Sterilization Other ISO and International Standards and European Regulations 4
J IEC60825 - Laser Safety - Calculating Maximum Permissible Exposure Correctly Other ISO and International Standards and European Regulations 1
optomist1 Maximum Shift Tolerance Restrained Part - Calculation of Datum Shifts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
K Is there a "maximum time limit" for nonconformity to fix? Internal Auditing 11
R Maximum allowed eccentricity for any shaft or hole diameter Misc. Quality Assurance and Business Systems Related Topics 7
R Can maximum ambient operating temperature be specified higher than 40C? IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
D Process Capability Study which involves MMC (Maximum Material Condition) Capability, Accuracy and Stability - Processes, Machines, etc. 8
S Maximum error of a z value from sample data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
G ANSI & ASME regardless of Maximum Material Condition when none is specified Various Other Specifications, Standards, and related Requirements 3
N ASME B89.1.13 question regarding Resolution and Maximum Permissible Error General Measurement Device and Calibration Topics 3
D ANSI IPC A-610 Maximum Jumper Wires Allowed on Boards by the IPC Standard Various Other Specifications, Standards, and related Requirements 1
D How to inspect maximum allowable contaminate Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
2 ISO 9001 should be implemented before CMMI for maximum effectiveness? Quality Manager and Management Related Issues 2
M Ergonomics - Human Hand - Maximum force - Wheel/Flywheel - Help ASAP :) Human Factors and Ergonomics in Engineering 8
F How to statistically determine the maximum performance life of a chemical Statistical Analysis Tools, Techniques and SPC 4
Chennaiite Process Capability for a specification with Maximum Material Condition (MMC) Statistical Analysis Tools, Techniques and SPC 4
F Maximum walking distance for assembly line worker Human Factors and Ergonomics in Engineering 28
M What is maximum limit for MU (Measurement Uncertainty) or acceptable criteria? Measurement Uncertainty (MU) 5
S GD&T Rule 1: Gage size for form check at MMC (maximum material condition)? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
D How to Measure a Diameter - Average or Maximum/Minimum? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 25
M Maximum Permissible Error for Monitoring and Measurement Devices General Measurement Device and Calibration Topics 19

Similar threads

Top Bottom