Cotton Swab Method Validation for clean uniforms, bins and poly bags

C

Chithra86

#1
Hi, my company has been practicing cotton swab method for clean uniforms, bins and poly bags. The same method has been in use for more than 10 years, and i plan to re validate the method. Since, i am new to Microbiology, i am not very sure how do i perform the validation for cotton swab method. Any guidance will be very much appreciated. Thank you
 
Elsmar Forum Sponsor

Jen Kirley

Quality and Auditing Expert
Staff member
Admin
#2
Good day Chithra86,

I can not guarantee this is the current approved method, but I found The Microbiology Guidelines document: OFFICIAL METHODS OF ANALYSIS (Appendix J, 2012) which talks about swabbing, with specific reference in page 6:
Sponges and swabs are premoistened with a neutralizing broth, such as Dey-Engley (2), prior to sampling. The entire sampling area is sponged or swabbed in both a horizontal and vertical motion. Use the sponges to sample a 100 cm 2 (4′′ ? 4′′) area and swabs to sample a 5 cm 2 (1′′ ? 1′′) area. Sponges/swabs containing samples are placed back into their individual respective bag or tube and held at room temperature for 2 hours prior to initiation of testing.
I hope this helps!
 
C

Chithra86

#4
Hi, from our historical data, we have never encountered any growth from the swab test method, and now i am not able to perform a recovery if i am not getting and, unless i perform the artificial innoculum. But as far as bin/uniform is concerned, i do not think its feasible to perform it. Any help would be much appreciated.
 
Thread starter Similar threads Forum Replies Date
Q US FDA Standard Limit for Swab Test Microbial and ATP Other US Medical Device Regulations 1
I Microbiological Monitoring for Surfaces - Switching from Swab to Contact Plates Other Medical Device Related Standards 2
R Swab, Rinse and Visual Cleaning Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
earl62 IATF 16949 Clause 9.1.1.1 - What is the batch conformance to specification method? IATF 16949 - Automotive Quality Systems Standard 3
A Beginners help with ISO3951-2 Combined control s-method n>5 what is Phi ?? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
S What is the best method to bond urethane foam to urethane foam? Manufacturing and Related Processes 4
R BS EN 12060:1998 - Zinc and zinc alloys Method of sampling Specifications Manufacturing and Related Processes 4
M Test method validation - Is MSA (MSA1, MSA2, MSA3 and linearity) a good solution? Medical Device and FDA Regulations and Standards News 1
Pau Calvo How to motivate colleagues to use 8D method or similar Quality Manager and Management Related Issues 37
M Proper method to test flow of a regulator Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
V Minitab GRR crossed Xbar and r method macro needed Using Minitab Software 7
R Please share your method used for PEMS Validation IEC 62304 - Medical Device Software Life Cycle Processes 5
S AS9100D PEAR - Examples for organization's method for determining process results? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
Y Post Market Surveillance Indicators and Analysis method EU Medical Device Regulations 1
S Seeking efficient method to manage install base data Manufacturing and Related Processes 0
D 4M (Man, machine, method, materials) Management Manufacturing and Related Processes 3
Ron Rompen Measurement Method Challenge - Measure feature #91 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
C FDA on changing acceptance criteria re: analytical method validation US Food and Drug Administration (FDA) 1
M Issuing NCR for improper method statement submissions Internal Auditing 10
F Irradiation Sterilization Validation Maximum Dose - VDmax 25 method (ISO 11137) EU Medical Device Regulations 2
I "Method" vs "Process" - Differences Document Control Systems, Procedures, Forms and Templates 33
P Test Method Validation for Tensile Strength Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
C ICP versus AA: How to determine appropriate specification for assay method verification Qualification and Validation (including 21 CFR Part 11) 1
I Flatness - Rectangular part using the three jack method Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
B AS9102 Supplier Flow Down Method AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
M Control Method correspondence between PFMEA and CP FMEA and Control Plans 8
qualprod Qualitative or Quantitative? Risk method evaluation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Ron Rompen Dual level holes - Measurement method suggestions wanted General Measurement Device and Calibration Topics 9
D Sampling method that is representative of the devices supplied by that distributor EU Medical Device Regulations 5
P Production Line Pump Cleaning Validation Method Question Qualification and Validation (including 21 CFR Part 11) 0
A Risks related to Method Validation and Stability Studies Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
qualprod ISO 9001 Risk control method - What could be the better way to control risks? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Ron Rompen Ford Method - Position Capability with MMC Modifier Capability, Accuracy and Stability - Processes, Machines, etc. 14
J Inspection method for T10 Hex driver General Measurement Device and Calibration Topics 13
qualprod ISO 9001:2015 - Right method to address clause 6.1? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
S Stakeholder analysis Method Six Sigma 1
T ER Checklist - Method(s) of conformity EU Medical Device Regulations 1
T ISO 17025 Method Verification - One sided “F” test or two sided “F” test ISO 17025 related Discussions 1
B Dosimeter Calibration Method General Measurement Device and Calibration Topics 5
A Test method for presence or absence of hexavalent chromium on a metal coating RoHS, REACH, ELV, IMDS and Restricted Substances 4
S Response Surface Method to create design matrix in Minitab Using Minitab Software 5
T Method to assess Measurement System of a Continuous Process Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
G HPLC Method Equivalence between USP and EP Various Other Specifications, Standards, and related Requirements 1
S Test Method Validation or MSA for Batch Manufacturing Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A Test Method Validation on Elongation Machine General Measurement Device and Calibration Topics 0
B Gage R & R on Visual Inspection using the Cross Tabulation Method IATF 16949 - Automotive Quality Systems Standard 4
C Method of Monitoring our Internally Produced Non-Conforming Products Quality Tools, Improvement and Analysis 7
C O-Ring Groove Measurement Method in CNC Machined Parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
G What's your favourite method of Electronic Folder Filing? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6

Similar threads

Top Bottom