Creepage and air clearance under vacuum conditions
Roland chung
Hi all,
I am wondering how to take into account the creepage and clearance under vacuum conditions (high voltage generator of X-ray, in my case). I can not find an answer from IEC 60664.
I will appreciate if you can provide some guidance to me.
Thanks,
Roland
Peter Selvey
Is it for functional or safety distances?
Normally the parts inside the vacuum tube would be functional, so you can base the distance on widely available research on breakdown in a vacuum (e.g. one paper I found gave a figure of ~45kV/mm, wikipedia gives 20-40kV/mm depending on shape/material, both of which far exceed the value for air which is around 3kV/mm).
Since there is no dust, pollution etc in vacuum, the distance would purely a function of clearance, not creepage.
Although some parts of the tube might be earthed, the circuit should be such that under failure conditions, the currents will only circulate inside the equipment, and cannot get outside.
Roland chung
Thank you.
Regarding the separation of high voltage, I find the answer from IEC 60601-2-44 CT scanners. No distances required but Hi-Pot test.
NOTE For CT SCANNERS installed with a fixed and PERMANENTLY INSTALLED PROTECTIVE EARTH CONDUCTOR it is assumed that there is no unacceptable RISK with regard to the reliability of the PROTECTIVE EARTH CONNECTION. The components with high reference voltages are: parts of X-RAY GENERATOR, X-RAY TUBE ASSEMBLY and in some cases parts of the digital acquisition system therefore, a reliable PROTECTIVE EARTH CONNECTION and sufficient dielectric strength of insulation are considered as equivalent to two MOPs. See also 201.8.8.3.
This also answers my another question, that is permanently installed protective earth alone can not be considered to provide 2 MOPs although the standard does not permit to interrupt such protective earth for SFC.
N PCB Creepage and Air Clearance on 8 Layer Boards IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Interpolation of creepage distance IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
H Secondary Circuits, Voltage Test and Clearance/Creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
S Some confusion about the creepage and clearance distance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Question on creepage/clearance requirements for HF Active Accessories for 2nd edition 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Creepage distance IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Electrosurgical generator with switchable sockets - creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Creepage - Does a surface mount isolator body negate a PCB slot? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Components in parallel to creepage\clearance IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H Defib Protected Applied Parts: 4mm clearance/creepage? IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
E Creepage distance accross insulation of step up transformer EU Medical Device Regulations 2
A IEC 60601 Creepage and Clearance Requirements on PCB IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
F IEC 61010 Creepage and Clearence Requirements Other ISO and International Standards and European Regulations 2
A Secondary Circuits, Voltage Test and Clearance/Creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F MOP (Means Of Protection) - Creepage and Clearance questions IEC 60601 - Medical Electrical Equipment Safety Standards Series 19
D Creepage and Clearance for high voltage low energy circuits IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G Creepage and Clearance Distance for MOOP on Secondary Circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C Creepage and Clearance for CR2032 powered fob for a Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
Steve McQuality Creepage Distance - Definition and Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T IEC 60601-1 Conformal Coating not an acceptable mitigation for creepage and clearance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
G We want to add Covid-2 /19 to our air filter 510 K Medical Device and FDA Regulations and Standards News 0
W Air Quality Measurement Hardware and Software General Measurement Device and Calibration Topics 11
M Informational US FDA – Ear, Nose, and Throat Devices; Classification of the Self-Fitting Air-Conduction Hearing Aid Medical Device and FDA Regulations and Standards News 0
E Employee complaining about air quality Occupational Health & Safety Management Standards 12
P HVAC System vs Air Handling Unit (AHU) - Differences Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
P Fan filter unit (un-ducted to air supply) vs fan filter unit (ducted to air supply) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
I Compressed air system qualification and 1 micron particle analysis Other Medical Device Regulations World-Wide 0
O Air Flow - Which is the operational difference between LAF (vertical and horizontal) and RLAF? Manufacturing and Related Processes 2
A Tubing Clog Tester - Convert from Air Bubble to Flow Rate Reliability Analysis - Predictions, Testing and Standards 1
triathlonx13 Leak test correlation - Bubble to Air Reliability Analysis - Predictions, Testing and Standards 8
I Process compressed air and contact with the product (ISO 8573-1:2010) Qualification and Validation (including 21 CFR Part 11) 0
S Performing PQ on a Small (tabletop) Air Heat Treating Oven ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
L Should Air Screwdrivers be Calibrated? General Measurement Device and Calibration Topics 15
N What does an effective design process look like? (shoots flare in the air) Design and Development of Products and Processes 2
R Measuring LiON Battery State of charge - Transporting batteries by Air RoHS, REACH, ELV, IMDS and Restricted Substances 1
R Air-powered dental handpiece with LED IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A 61010-1 Accessible Parts - Limit Values and Crepage and Air Clearance Distances Other Medical Device Related Standards 1
somashekar Is it a good practice to have Air Conditioner in the Gym Hospitals, Clinics & other Health Care Providers 2
S Is 0.1 micron monitoring required for ISO Class 5 Laminar Air Flow unit ? Other ISO and International Standards and European Regulations 4
M Regulations for oxygen/air-tubing and connectors inside of ME equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Setting air gauging outside of a part tolerance Capability, Accuracy and Stability - Processes, Machines, etc. 2
S Can Validation be done for fewer Air Changes than specified? Qualification and Validation (including 21 CFR Part 11) 6
Marc Air Pollution in the Early 21st Century Coffee Break and Water Cooler Discussions 5
C SOP for Door to Door Air Cargo Document Control Systems, Procedures, Forms and Templates 3
S What are air changes are required for ISO Class 8 Clean Room? US Food and Drug Administration (FDA) 2
V Validation Test and Acceptance Criteria for Compressed Air Qualification and Validation (including 21 CFR Part 11) 3
E Is Volumetric Air Sampling required for Class ISO 7 Cleanroom ? ISO 13485:2016 - Medical Device Quality Management Systems 1