Creepage distance accross insulation of step up transformer
EMengineer
I am a little confused about determination of creepage distances. There is switchin PSU in which mains voltage (230) are transformed to 400V and transformed again to 1500 V (output) and then rectified to DC. 60601-1 standard requires to apply creepage distance according working voltage accross that insulation. Working voltage is defined as the maximum voltage accross that insulation. In this case the maximum voltage in transformer (400 V to 1500 V) is 1500 V, so I should apply values for 1500 V?
Somewhere I have found that voltages are summed, but in this case (400 V + 1500 V = 1900 V) the creepage distance would be enourmously large.
Also if my output "-" is earthed. Therefore, in case we intend to make protection from dangerous currents/voltages on the enclosure we should provide protection from positive output to negative output. However, positive and negative outputs are connected by a diode bridge rectifier and some other components (like capacitors etc.). So, these components (diodes, capacitors) are bridging a MEANS OF PROTECTION just the same way as resistors are mentioned in Subclause 4.8. In fact, these compotes are mounted in parallel to feedback resistors. I do doubt that we will be able to find diodes, which can "have a construction that cannot fail short-circuit".
Thnak you for your answers in advance!
CharlieUK
What is this circuit used for / connected to?
Creepage distances are required between Hazardous Voltages and Accessible Parts or Accessible Circuits - they are not required between Harzardous Voltages and Hazardous Voltages
EMengineer
I attached insulation diagram bellow:
D Interpolation of creepage distance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
S Some confusion about the creepage and clearance distance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Creepage distance IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Creepage and Clearance Distance for MOOP on Secondary Circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Steve McQuality Creepage Distance - Definition and Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C Creepage requirement Wearable ECG IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Creepage and Clearance 8.9.1.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Secondary Circuits, Voltage Test and Clearance/Creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Question on creepage/clearance requirements for HF Active Accessories for 2nd edition 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N Electrosurgical generator with switchable sockets - creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Creepage - Does a surface mount isolator body negate a PCB slot? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
rothlis Components in parallel to creepage\clearance IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H Defib Protected Applied Parts: 4mm clearance/creepage? IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
A IEC 60601 Creepage and Clearance Requirements on PCB IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
F IEC 61010 Creepage and Clearence Requirements Other ISO and International Standards and European Regulations 2
A Secondary Circuits, Voltage Test and Clearance/Creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F MOP (Means Of Protection) - Creepage and Clearance questions IEC 60601 - Medical Electrical Equipment Safety Standards Series 19
R Creepage and air clearance under vacuum conditions IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N PCB Creepage and Air Clearance on 8 Layer Boards IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Creepage and Clearance for high voltage low energy circuits IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C Creepage and Clearance for CR2032 powered fob for a Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
T IEC 60601-1 Conformal Coating not an acceptable mitigation for creepage and clearance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M 4G LTE Cat4 operator distance restriction IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
P Brazil - ANVISA - distance selling Other Medical Device Regulations World-Wide 0
P Minimum Distance from Floor to Underneath of Machine Human Factors and Ergonomics in Engineering 5
K IPC-610 Section 10.4.2.2 clarification - Distance to be measured Various Other Specifications, Standards, and related Requirements 0
F Holes position inspection - Calculation of centre distance and plug sizes of a gauge Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
E Internal clearance distance according to 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Bachelor of Engineering Course - Distance learning Training - Internal, External, Online and Distance Learning 2
L Patient Monitor Viewing Distance Usability Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
JoCam Equipment Guarding - Safe Distance - Is there a specified minimum distance? Occupational Health & Safety Management Standards 6
V How to design a gauge (fixture) for measurement of distance between two hole Design and Development of Products and Processes 23
C Storage distance between flammable & non-flammable goods - guidelines/standards? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
Marc TV Viewing Distance Factor After Work and Weekend Discussion Topics 9
S Advice on Distance TS 16949 Auditor Training General Auditing Discussions 1
Stijloor Google Voice - Free long-distance? Oh yeah, Google's going there After Work and Weekend Discussion Topics 9
hogheavenfarm Reference Distance - Print which shows "1.656 REF" as a dimension Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
J Distance between two holes tolerance Inspection, Prints (Drawings), Testing, Sampling and Related Topics 18
F Maximum walking distance for assembly line worker Human Factors and Ergonomics in Engineering 28
P What is Distance learning course? Training - Internal, External, Online and Distance Learning 8
Q Aerospace Quality related Distance Training/Learning/Exams Training - Internal, External, Online and Distance Learning 2
D Distance Education Question. (Online Degree) Training - Internal, External, Online and Distance Learning 5
Marc Distance Learning - From ivory tower to academic sweatshop Training - Internal, External, Online and Distance Learning 2
Wesley Richardson Internet distance training voice over IP (VOIP) for internet synchronous training Training - Internal, External, Online and Distance Learning 2
N Distance Learning - Quality Management and TQM - Seeking Recommendations Training - Internal, External, Online and Distance Learning 1
A Distance Learning where I would be able to obtain a 4 year degree in Quality? Training - Internal, External, Online and Distance Learning 7