Some very basic questions that are eluding me:

1. How does one define a gage's tolerance (i.e arbitrary or based on a spec)
1.1 From the Manufacturer's documents (they don't seem to like to specify this)?
1.2 From a specification (i.e. GGG-C-105C is obsolete, is there a replacement or other document that is more comprehensive?)
1.3 What are the best resources for learning more about gage tolerance
1.4 Is it acceptable to apply the 1:10 rule based on the tolerances of the parts I will be inspecting - can anyone provide a citation that invokes the 1:10 rule?

2.0 The basis for my question include 0-1" Digimatic Micrometers with .00005" resolution (blade, 1/4" barrel, and .080" step barrel); 0-1" Digimatic Height Gage; SJ-400 Surface Finish Gage; and even a Nikon vision system. The manuals for these instruments speak to "accuracy" but are silent about their uncertainty (at least not in numeric terms - inches or mm).

Thank You,

Jim
 
Elsmar Forum Sponsor
A

Antonio53

#3
Usually if you need a gage, you will choose the one that meets your need (specs), so the tolerance will be defined by the manufacturer of the gage chosen that you will accept. Sometimes the tolerance can be redefined by the end user, but, in this case, it (the new one) must be fully validated.
 
Last edited by a moderator:
S

samsung

#4
Usually if you need a gage, you will choose the one that meets your need (specs), so the tolerance will be defined by the manufacturer of the gage chosen that you will accept. Sometimes the tolerance can be redefined by the end user, but, in this case, it (the new one) must be fully validated.
Though I am not good at handling the measuring devices, yet, IMO, what you call 'tolerence' here is the 'least count' that can be measured by the instrument and certainly it is different from the term 'tolerance', also known as 'acceptance' to which the rights are (should be) mostly reserved with the customer and not the otherwise.

The 'tolerance' (which I suppose to mean 'LC' here) even though defined by the manufacturer should be validated (confirmed) prior to putting the instrument in use.
 
A

Antonio53

#5
Of course, even if you will accept the specs of the manufacturer, you should, in any case, qualify/validate the gage in order to demonstrate its Effectiveness/Efficiency in the process that you want to keep under control
 
S

samsung

#6
Of course, even if you will accept the specs of the manufacturer, you should, in any case, qualify/validate the gage in order to demonstrate its Effectiveness/Efficiency in the process that you want to keep under control
You validate or verify the gage just to ensure that it reads correctly whatever is subjected to measurement. What demonstrates the efficiency of the process is not the measuring accuracy but the decision or action that you take based upon the measurement so as to keep the process on track/ under control. Isn't it ?
 
A

Antonio53

#7
You should validate the gage in order to demonstrate that it is the tool you need to assure that the part of the process monitored with it will give product according to the specs you have defined. The decision/action you wqill take, based on the measurement, is part of the control of that specific part of the process, that's it. If the tolerance measured will be out of the specs this will mean that the process needs some correction in order to have, again, something that will be suitable for the next step.
 
S

samsung

#8
You should validate the gage in order to demonstrate that it is the tool you need to assure that the part of the process monitored with it will give product according to the specs you have defined. The decision/action you wqill take, based on the measurement, is part of the control of that specific part of the process, that's it. If the tolerance measured will be out of the specs this will mean that the process needs some correction in order to have, again, something that will be suitable for the next step.
Correct. That's the point.
 

BradM

Staff member
Admin
#9
Jim, you ask a million dollar question.:agree1:

For starters, I will use the mfg. tolerance for a device. Because, in theory, that is what the device is supposed to be able to do. However, some mfg. devices never meet the mfg. specification; no matter how short the interval. So, that leads me to #2.

#2 set the tolerance so that the device has the required accuracy for the process. Say I have a pressure gauge that the mfg. says is +/.1 PSI. I use the gauge on a process where I'm measuring something +/-5 psi. I might set the gauge tolerance to .5 or even 1 psi.

So, I would start with MFG. tolerance. Then if your process spec is known, go backwards from there and determine what the device should meet.

Hope that helps.
 

MalibuJim

Starting to get Involved
#10
Brad and Everyone else,

Thank you for the guidance on the gage meeting the minimum requirements for any given application. To me this is straightforward 10:1. I have an O.D. with a tolerance of ±.0001". So my gage must be accurate to within ±.00001" (ten millionths). Accuracy being the ability to repeat and distinguish between to different parts whose values might be .10001" and .10002". So, the "read-out" on the gage is meaningless. In my example I would probably only be able to use a super-micrometer.

Next issue. Does anyone have any suggestions for specifications or text books that define the flow of gage selection and incorporate the 10:1?

Perhaps I'm hung up on having a published document that directs this. Obviously I could write my own procedure and follows my first paragraph in this post and then put same into practice. No problem. I just like to have a body-of-knowledge reference (plus I like to read).

Take Care,

Jim ASQ-CQE; SSBB
 
Thread starter Similar threads Forum Replies Date
V Defining the criteria for equipment to be qualified or requalified Qualification and Validation (including 21 CFR Part 11) 2
R Criteria for defining FMEA & Control Plan Critical and Significant Characteristics FMEA and Control Plans 10
T Defining Criteria for Risk Acceptability - ISO 14971 Clause 3.2 ISO 14971 - Medical Device Risk Management 4
W Calibration 4.11 2c - Defining acceptance criteria for all equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D What is the acceptance criteria/requirement for Stability Study? General Measurement Device and Calibration Topics 5
L Gage R&R TMV Acceptance Criteria Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S On-time delivery CRITERIA AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
M V&V phase: Justification of acceptance criteria (statistical method ) - (Medical Device) Design and Development of Products and Processes 2
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
M ISO 14971:2019: Criteria for overall residual risk ISO 14971 - Medical Device Risk Management 6
O GR&R Acceptance Criteria Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
S Shore A hardness Tester Acceptance Criteria Standard Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 2
lanley liao Purchase Acceptance Criteria - Tensile testing Oil and Gas Industry Standards and Regulations 2
M Audit Criteria Training Materials Internal Auditing 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
B GM GP-12 Exit Criteria - Additional Safe Launch Inspection - (Also see GM1927-28 ) Customer and Company Specific Requirements 2
B AS9100 8.4.1 Supplier Selection/Evaluation criteria and reevaluations AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 13
I Mother bobbin final inspection criteria Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
R Procedure, Frequency and Acceptance Criteria for Replicate, Recalibration, Before-After and Intermediate Checks ISO 17025 related Discussions 8
P Design FMEA - Detection Rating criteria ISO 14971 - Medical Device Risk Management 3
J Criteria Matrix To Initiate An A3 or 8D ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
qualprod Best criteria to measure Corrective Action effectiveness - Poor Maintenance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
B ASTM F2924-14 /ASTM F3001 - Room temperature classification criteria Other US Medical Device Regulations 0
M Supplier selection criteria - Medical devices ISO 13485:2016 - Medical Device Quality Management Systems 3
qualprod Criteria to raise a Nonconformance based on KPI values ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 39
M Criteria and when to carry out Setup Approval Manufacturing and Related Processes 5
V What is the criteria to cite an good documentation practices observation as an data integrity related issue US Food and Drug Administration (FDA) 6
S Acceptance criteria for gauges General Measurement Device and Calibration Topics 8
M Characterization Testing - NO acceptance criteria, no minimum performance requirement Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
J AIAG PPAP: 2.2.11.3 Acceptance Criteria for Initial Study (page 9, 4th edition) APQP and PPAP 1
C FDA on changing acceptance criteria re: analytical method validation US Food and Drug Administration (FDA) 1
Z UDI assignment Criteria Software IVDR Other Medical Device Regulations World-Wide 0
Ashok sunder What are the criteria for setting a target KPI value against a quality objective? Benchmarking 7
qualprod Criteria for print shop - How do you consider the cycle time in a print shop? Lean in Manufacturing and Service Industries 21
eule del ayre IATF 16949 / ISO 9001:2015 audit criteria IATF 16949 - Automotive Quality Systems Standard 17
GStough Auditing Against Criteria Unfamiliar to Auditee - Yea or Nay? General Auditing Discussions 11
M Internal Audit Assessment Criteria - ISO 13485:2016 Internal Auditing 21
S Agreement for allowance for visual criteria after submission APQP and PPAP 1
V Process and Internal Audit Criteria matrix wanted Internal Auditing 8
I Writing a Cosmetic Criteria for Switches Document Control Systems, Procedures, Forms and Templates 1
M Example ISO 14971 policy and risk criteria ISO 14971 - Medical Device Risk Management 0
D How to set equipment calibration/verification criteria General Measurement Device and Calibration Topics 1
A Operating criteria for production processes - ISO 14001 Cl. 8.1 ISO 14001:2015 Specific Discussions 3
B The most important criteria of IEC 60601-1 medical devices safety IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P Supplier (Vendor) Evaluation Criteria - IATF 16949 Cl. 8.4.1.2 IATF 16949 - Automotive Quality Systems Standard 1
H CER - Equivalent Medical Devices Criteria - MDR/MEDDEV REV 4 CE Marking (Conformité Européene) / CB Scheme 9
K Internal Audit Scope and Criteria Internal Auditing 9
K ISO 9001:2015 clause 9.2.2 a. - Define the audit criteria and scope Internal Auditing 2
Q Criteria to pursue an opportunity ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7

Similar threads

Top Bottom