Definition Critical Component Definition

RoxaneB

Change Agent and Data Storyteller
Super Moderator
Re: Critical Component Defined

regarding the point that all the components are critical, I am not with you on that.

I did not say that every component is critical, but I asked HOW to determine what is critical above other aspects. I wanted you to provide more of a definition that what you originally said and your response above provided that...hopefully to the Original Poster's use. :applause:
 

somashekar

Leader
Admin
Re: Critical Component Defined

We mfg a medical device and currently going through CE mark for a new product. I guess I am looking for a general definition of critical component. We test the product 100%. The component is an oscillator. There is no UL/Safety marks on the component.

No UL/Safety mark is fine.
Is your medical device subjected to safety tests and approvals by a recognized test house ?
When this is done, the test house along with your designers will decide which of the components are critical and list them in the report. You may then have to respect this in your purchasing and inspection controls.
This decision is taken based on safety analysis of your device. Safety concerned to the patient / user.
Usually termed as safety critical components
You can further add components to this list based on your design analysis for performance as well, and get your bunch or critical components.
.
..
Further my interpretation ...
When a component is critical to the medical device effectiveness in the sense that the lack of effectiveness could lead to death or major irreversible damage, such a component is into safety critical component.
 
Last edited:

Sullen-gent

Involved In Discussions
Hoping some one responds but what Regulatory requirements is there to define parts as 'critical'? I see Critical comes up one the Electrical side and is a requirement of 60601-1. However for non-active device, what clauses would we not meet by not defining 'critical' parts components etc. Background I'm working on older devices where DFMEA were never created as part of D+D.
 

ChrisM

Quite Involved in Discussions
A little late but....
generally recognised definition is "a critical component is one whose failure could result in a hazardous situation". (I know that this is a bit vague). Examples quoted are failure of insulation on a mains power lead (as someone could get an electric shock if they were to touch it), fuse (may not blow when it should and therefore allow higher current than designed to pass), batteries, relays etc
 

Ed Panek

QA RA Small Med Dev Company
Leader
Super Moderator
It can be tied to risk or what the device's "essential performance" is tied into. For example, a car would have a differential, brakes, and engine as part of its essential performance. Lighting, seat belts, airbags as part of its risk mitigation.

I would think cup holders, radio, and speakers would be not critical.
 

RoxaneB

Change Agent and Data Storyteller
Super Moderator
It can be tied to risk or what the device's "essential performance" is tied into. For example, a car would have a differential, brakes, and engine as part of its essential performance. Lighting, seat belts, airbags as part of its risk mitigation.

I would think cup holders, radio, and speakers would be not critical.

There is the risk to the "essential performance" and there is also the risk of not meeting customer requirements. If a part is not necessary to essential performance but (a) is tied to the customer requirements and (b) has a long lead time and low inventory, it could be deemed critical.

If, sticking with the car analogy, the customer wanted the cup holders to be made from titanium or some fancy marble (silly, but work with me here), and these were not common requirements so there is low inventory (with a long lead time for delivery), a delay in the ability to install could result in a failure to provide the customer's product on time (i.e., risk).

This is why, as part of the risk analysis, it's important to look beyond pure performance specs and use a numbers-based (i.e., math) approach to calculating the criticality level of each component. We may say that a part is critical from a customer requirement perspective but that carries less weight than a part that is vital to the performance. So be it...but at least we've done the exercise.
 

ChrisM

Quite Involved in Discussions
So what happens with a small device with very limited electronics? The failure of any single component would affect the performance/operation of the device, so are all components "critical" or only those where safety becomes a risk upon failure ?
 

Ed Panek

QA RA Small Med Dev Company
Leader
Super Moderator
So what happens with a small device with very limited electronics? The failure of any single component would affect the performance/operation of the device, so are all components "critical" or only those where safety becomes a risk upon failure ?

Possibly. A car remote is very simple. I suspect every item is critical in the electrical layout. Is the color of the plastic enclosure? Maybe but I could see production with the slightly off color plastic under deviation being ok

it’s also important though for specific customer requirements. If the car is a Bentley that color might be really important to the Bentley image or brand as Roxane stated.
 

Miner

Forum Moderator
Leader
Admin
Part of your answer will depend on your particular industry. These terms are not consistently applied across industries.

In the automotive industry, the term Critical is reserved solely for safety characteristics governed by FMVSS (Federal Motor Vehicle Safety Standards) or the local country's equivalent standards. The terms Significant (Ford) and Key (GM) are used for characteristics that are important to the function of the product.
 

AndreB

Registered
For my point of view, after conducting a criticality evaluation of components which is based on severity, occurence (RPN2) and also detectability (RPN3), this latest is normally conducted during developpement process, the final evaluation after implementation of apprpriate mitigationswill normally reveals components elected to be critical components. It becomes important to keep such evlauation traceable in the DHF of the medical devices.
 
Thread starter Similar threads Forum Replies Date
L Automatically considered Critical Component when within a fire ENCLOSURE? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
S Who compiles the Critical Component List, Insulation Diagram etc.? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
MDD_QNA What is the function of the Critical Component List? ISO 14971 - Medical Device Risk Management 2
J Clarification in UL Mark in Critical Component - IEC 60601 testing for CE mark in UL IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G Changing Non-Critical Component Suppliers - Convenience kits (FDA) Procedure packs Other Medical Device and Orthopedic Related Topics 2
L Clause related to Failure to fulfill delivery date of critical item ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
P Vendor for shipment critical or non-critical ISO 13485:2016 - Medical Device Quality Management Systems 6
G MDR critical supplier audit checklist EU Medical Device Regulations 0
P Product label supplier, critical or non-critical vendor ISO 13485:2016 - Medical Device Quality Management Systems 6
R 5.6.1.2 Initial Supplier Evaluation—Critical Purchases Supplier Quality Assurance and other Supplier Issues 4
J API Q1 - 5.6.1.2 (b) - Initial Supplier Evaluation—Critical Purchases Oil and Gas Industry Standards and Regulations 1
Bev D Development Process: FMEA, Critical Characteristics, Spec Setting and Controls Misc. Quality Assurance and Business Systems Related Topics 0
K Critical Material Vendors: Distributors Vs Manufacturers and Supply Agreements ISO 13485:2016 - Medical Device Quality Management Systems 1
J Critical Supplier Initial evaluation for single purchases Oil and Gas Industry Standards and Regulations 16
Pmarszal Technical File - Critical Supplier List Questions CE Marking (Conformité Européene) / CB Scheme 2
D Critical Supplier will not allow us to audit Plant floor US Food and Drug Administration (FDA) 12
M IEC 62366 Hazard-related Use Scenario vs FDA Critical Task EU Medical Device Regulations 3
F Change in address for critical supplier - does NB need to be informed? CE Marking (Conformité Européene) / CB Scheme 20
S Critical spare management guideline IATF 16949 - Automotive Quality Systems Standard 1
G Critical Characteristics/Dimensions of PCBAs Inspection, Prints (Drawings), Testing, Sampling and Related Topics 16
H Critical subcontractors & crucial suppliers EU Medical Device Regulations 5
A Critical Characteristics - Sheet metal stamping APQP and PPAP 9
D What are the Activity deemed critical for the API 6D Ball Valve ? Oil and Gas Industry Standards and Regulations 4
lanley liao Questions regarding the change of critical suppliers and key personnel Oil and Gas Industry Standards and Regulations 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
S Distinction between a critical supplier and a Virtual manufacturer EU Medical Device Regulations 3
R Critical Characteristics & Special Characteristics FMEA and Control Plans 2
J Where is the definition of a critical supplier? ISO 13485:2016 - Medical Device Quality Management Systems 5
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
B Software service provider as critical supplier ISO 13485:2016 - Medical Device Quality Management Systems 5
M Critical Characteristics APQP and PPAP 2
B FCA US Customer Specific IATF 16949- Critical Characteristics 8.6.2 Customer and Company Specific Requirements 2
B ISO 11607-2 "Critical Parameter" vs. "Process Parameter" Other Medical Device Related Standards 6
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
Jane's Like-for-like critical raw material change qualification - type of testing/ number of lots required ISO 13485:2016 - Medical Device Quality Management Systems 4
F Is there such a thing as 'Critical supplier' status under the Dept. of Health (UK)? Medical Device and FDA Regulations and Standards News 2
D Definition of "Critical Laboratory Equipment"? General Measurement Device and Calibration Topics 1
M Changing supplier of critical raw material (III class device) Other Medical Device and Orthopedic Related Topics 1
C Critical Suppliers for EU medical device approvals May 2020 EU Medical Device Regulations 1
M Informational Critical Thinking and the Process of Evidence-Based Practice Medical Device and FDA Regulations and Standards News 0
J How to measure a Critical Characteristic that is a Basic Dimension Misc. Quality Assurance and Business Systems Related Topics 6
O Examples of Critical process parameter (CPP) and Critical quality attribute (CQA) Manufacturing and Related Processes 2
R Critical suppliers (Definition of) and MDSAP (Medical Device Single Audit Program) ISO 13485:2016 - Medical Device Quality Management Systems 22
C Suppliers re-classification from Critical to Significant or from Significant to Non-critical Supply Chain Security Management Systems 0
H Critical Supplier Agreement acc. to NBOG 2010-1 Annex II EU Medical Device Regulations 10
andika_untoro What are a good Critical to Quality (CTQ) metrics for a biotech manufacturing production? Lean in Manufacturing and Service Industries 0
A API Q1 9th ed Addendum 2 - Vendor Assessment for critical products Oil and Gas Industry Standards and Regulations 21
M Does a CDR (Critical Design Review) have to have a moderator that is unconnected to the team? Design and Development of Products and Processes 4
S Critical supplier - Obligated to have an ISO-certified QMS? ISO 13485:2016 - Medical Device Quality Management Systems 8
Al Rosen Critical Supplier/Quality Agreement Procedure ISO 13485:2016 - Medical Device Quality Management Systems 8

Similar threads

Top Bottom