Current standard references IEC 60601:1988 instead of current version
milagre
I am trying to go through each subclause of EN ISO 8185 right now to ensure we are compliant with it. It continually references IEC 60601-1:1988, which I don't have a copy of (we only keep a copy of the newest 60601-1).
I am finding it hard reading 8185 because I don't always understand what sections of 60601-1 it is referring to and modifying. Anyone else dealt with this problem before? Any advice? I'd really rather not buy a 1988 version of 60601-1.
SteveK
I sympathise with your problem. I have to comply with ISO 8185 also; however our device is a simple (passive) bubble humidifier so IEC 60601-1 clauses do not come into play, only a limited number of clauses in ISO 8185 do in fact apply for compliance purposes. This problem with the 1988 version of IEC 60601-1 comes into play with another standard which is an issue for me i.e. ISO 10079-1 (for Electrical Suction Devices), which references its clauses similarly throughout. We are developing a new device which needs to comply with this (equally messy) standard. Whoever developed and approved these standards referencing IEC 60601-1:1988 in the past were totally lazy b*%$£?ds in my opinion. I my case I do have a copy of IEC 60601-1:1988 (IEC 601-1:1988) but it does not really help as we would not be able to conduct the tests etc. referenced in-house (by the way I archive copies of all old standards). I am hoping that the new draft (DIS) ISO 10079-1 will be published as a full standard soon, as this removes any reference to IEC 60601-1:1988 and we would be able to conduct all the tests required in-house. Similarly ISO 8185 is being replaced by - ISO 80601-2-74 “Medical Electrical Equipment Part 2-74: Particular requirements for basic safety and essential performance of respiratory humidifying equipment”. Again this removes any reference to 60601-1:1988 but is also only at the DIS stage. Incidentally this DIS standard does not include bubble humidifiers! There is now a new separate standard for passive humidifiers but this is only at the WD stage at present. Ultimately in the past when we have had to show compliance with a standard which references IEC 60601-1:1988 clauses, we have relied on a test house and their certification. Sorry about not being too helpful.
Steve
SteveK
Correction!
Just checked my Bubble Humidifier Product Dossier and a number of clauses (referenced in ISO 8185) from IEC 60601-1:1988 do in fact apply!
Steve
milagre
Thanks for all the info Steve. At least I'm not the only one frustrated by this.
Marcelo
I see the following options:
- use the harmonized standard and related second edition of IEC 60601 - this is the most straightforward and you keep the presumption of conformity
- use the harmonized standard, but instead of using the second edition of IEC 60601, use the third edition. This would require analysis of which related clauses of third edition applies. But although possible, it?s generally not expected that editions are mixed.
- use the revision of ISO 8185, ISO 80601-2-74 Ed. 1.0
Medical Electrical Equipment - Part 2-74: Particular requirements for basic safety and essential performance of respiratory humidifying equipment. This would be the better solution, however ISO 80601-2-74 is still in the CDV stage of the drafting process. If it was at least in DIDS, its use could be readily accepted in principle, but on CDV, it?s more difficult.
So it does seem that the first option is really the best at the moment.
S Current Inspection Standard for AS9100 Certified Job Shop AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
drgnrider Certify to new ISO 14001 standard or stay with current standard? ISO 14001:2015 Specific Discussions 8
S ISO 9001, Fourth edition, 2008-11-15 - Do I have the most current standard? Document Control Systems, Procedures, Forms and Templates 7
Project Man What is the most current ISO 9000 standard? I heard there is an ISO 9100 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
E Is there a summary of changes from a withdrawn ISO standard to the current version Other ISO and International Standards and European Regulations 5
K Leakage current tests for battery powered device with IEC 62368 compliant charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H 60601 - Patient leakage current via the patient connection test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Using current gauge to verify masters General Measurement Device and Calibration Topics 1
K Alternatives to Metal Foil for touch current measurement of enclosures made of insulating material. IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Help Understanding Current Role; Distributor, Manufacturer, or Importer EU Medical Device Regulations 4
C IEC60601-1: Impedance and current-carrying capability (8.6.4) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Leakage current measurement without isolating transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Precision Current Loop for Current Probe Calibration Reliability Analysis - Predictions, Testing and Standards 2
N Auxiliary current limit with coil cell powered device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B Leakage Current AC-DC discrimination per cl. 8.7.3 b) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Earth (ground) leakage current routine test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T How can I determine the current level of exposure to talc in the workplace? Occupational Health & Safety Management Standards 6
H Prodigit DC Electronic Load - Drawing current but showed 0 on the screeen Manufacturing and Related Processes 0
D Add new device to current license Canada Medical Device Regulations 3
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W FMEA - Current control and occurrence rating FMEA and Control Plans 3
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 5
W Need for current design or process control FMEA and Control Plans 2
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E Best way to keep up with pharmaceutical quality knowledge and current affairs Service Industry Specific Topics 1
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Informational TGA – Current status of breast implant products in Australia Medical Device and FDA Regulations and Standards News 0
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
A Touch current in single fault conditions test and earth leakage current in normal conditions test, are they really different tests? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
T Tests IEC 60601-1 - Grounding Resistance, Leakage Current and Dielectric Rigidity in an automated way IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
S Calibration - Isolated current meter - Measuring patient auxiliary leakage current Calibration and Metrology Software and Hardware 4
Q Raising A CAPA on Current CAPA system US Food and Drug Administration (FDA) 3
A SPC for Eddy Current Testing Statistical Analysis Tools, Techniques and SPC 28
A Adding Remote Site to Current IATF 16949 Certification IATF 16949 - Automotive Quality Systems Standard 2
S Leakage Current test and Dielectric strength for Battery operated ECG device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Leakage current measurement with non-frequency-weighted device IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
A QMS - Medical Device - New company under a current one ISO 13485:2016 - Medical Device Quality Management Systems 3
E IEC 60335-1 leakage current measurement Other ISO and International Standards and European Regulations 3
K Questions on Technical File to show it is current EU Medical Device Regulations 5
H IS there any database for list of CE marked Medical devices and there current status EU Medical Device Regulations 8
B Mains fuses and OVER-CURRENT RELEASES IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Adding an additional distributor to a current Mexico Medical Device Registration Other Medical Device Regulations World-Wide 0
M IEC 60601-1 and Active Current Protection Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Testing a Low Current (Battery) DC Medical Device Other Medical Device Related Standards 4
D Referencing Standards in Procedures - Current version or actual? Document Control Systems, Procedures, Forms and Templates 6
I Curious if a List of Current Standards Revisions Exsists Other Medical Device Related Standards 3
C Touch Current for Exposed Charging Contacts on Bodyworn Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Current version of EN62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5