Current version of EN62304/A1

R

regcos

#1
Hi all,

I am studying the EN62304/A1 and I have a question:

"The first edition of IEC 62304 was published in 2006. This amendment is intended to add requirements to deal with LEGACY SOFTWARE, where the software design is prior to the existence of the current version, to assist manufacturers who must show compliance to the standard to meet European Directives"

As "current version" we can consider the EN62304:2006 or the whole EN62304:2006/amd1:2015?
 
Elsmar Forum Sponsor
R

regcos

#3
So do you confirm that a medical device software marketed in 2010, when complying with EN62304 today can be handled as a Legacy Software, thus not requiring the whole documentation but only what is stated in Section 4.4 of EN62304/A1?
 

Pads38

Trusted Information Resource
#4
The amendment has not, yet, been published in the Official Journal (latest is May 2016) - so the current EN version is EN 62304:2006/AC:2008 (corrigendum published 2008).

But the latest version of IEC62366 has not yet been listed, and I have successfully used that standard (the Annex C for legacy usability). I highlighted the fact taht it was not yet harmonised and our NB did not object to it's use.
 

Marcelo

Inactive Registered Visitor
#5
So do you confirm that a medical device software marketed in 2010, when complying with EN62304 today can be handled as a Legacy Software, thus not requiring the whole documentation but only what is stated in Section 4.4 of EN62304/A1?
No, I only mentioned that the "current version" in the text "The first edition of IEC 62304 was published in 2006. This amendment is intended to add requirements to deal with LEGACY SOFTWARE, where the software design is prior to the existence of the current version, to assist manufacturers who must show compliance to the standard to meet European Directives" is the version + amendment. This is always true when standards in general say "current version".
 

Marcelo

Inactive Registered Visitor
#6
Anyway, although the current version as read does mean the current full version due to some technicalities, the idea of the requirement was to give a way to comply to devices that were created to 2006, not 2015.
 
Thread starter Similar threads Forum Replies Date
D Referencing Standards in Procedures - Current version or actual? Document Control Systems, Procedures, Forms and Templates 6
M Current standard references IEC 60601:1988 instead of current version EU Medical Device Regulations 5
E Is there a summary of changes from a withdrawn ISO standard to the current version Other ISO and International Standards and European Regulations 5
Z How can an employee know that he is using the most current version of documentation? Document Control Systems, Procedures, Forms and Templates 38
T Ford Q1 2002 revision the most current version? Customer and Company Specific Requirements 5
Q The most current version or draft of the Q10 - Pharmaceutical Quality Systems Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
A Electronic Forms control - Making sure the most current version of a form was used Document Control Systems, Procedures, Forms and Templates 13
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 6
W Need for current design or process control FMEA and Control Plans 2
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E Best way to keep up with pharmaceutical quality knowledge and current affairs Service Industry Specific Topics 1
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Informational TGA – Current status of breast implant products in Australia Medical Device and FDA Regulations and Standards News 0
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
A Touch current in single fault conditions test and earth leakage current in normal conditions test, are they really different tests? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
T Tests IEC 60601-1 - Grounding Resistance, Leakage Current and Dielectric Rigidity in an automated way IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Calibration - Isolated current meter - Measuring patient auxiliary leakage current Calibration and Metrology Software and Hardware 4
Q Raising A CAPA on Current CAPA system US Food and Drug Administration (FDA) 3
A SPC for Eddy Current Testing Statistical Analysis Tools, Techniques and SPC 28
A Adding Remote Site to Current IATF 16949 Certification IATF 16949 - Automotive Quality Systems Standard 2
S Leakage Current test and Dielectric strength for Battery operated ECG device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Leakage current measurement with non-frequency-weighted device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A QMS - Medical Device - New company under a current one ISO 13485:2016 - Medical Device Quality Management Systems 3
E IEC 60335-1 leakage current measurement Other ISO and International Standards and European Regulations 3
K Questions on Technical File to show it is current EU Medical Device Regulations 5
H IS there any database for list of CE marked Medical devices and there current status EU Medical Device Regulations 8
B Mains fuses and OVER-CURRENT RELEASES IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Adding an additional distributor to a current Mexico Medical Device Registration Other Medical Device Regulations World-Wide 0
M IEC 60601-1 and Active Current Protection Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Testing a Low Current (Battery) DC Medical Device Other Medical Device Related Standards 4
I Curious if a List of Current Standards Revisions Exsists Other Medical Device Related Standards 3
C Touch Current for Exposed Charging Contacts on Bodyworn Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Power Traceability certificate for values for voltage, current and phase General Measurement Device and Calibration Topics 1
Q What constitutes current technical data? Federal Aviation Administration (FAA) Standards and Requirements 2
K What are the current ISO/ TS 16949 Semiconductor Manufacturer Requirements IATF 16949 - Automotive Quality Systems Standard 12
G How to cover another manufacturing "site" or warehouse into current ISO 13485 cert ISO 13485:2016 - Medical Device Quality Management Systems 7
P Document references need to be updated to correspond to the current year? Document Control Systems, Procedures, Forms and Templates 5
P Humidity Pre-Conditioning before leakage current and dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Gman2 The current state of our Calibration System General Measurement Device and Calibration Topics 28
R Is EN/IEC 60417 still current? Other Medical Device Related Standards 4
R New Draft Guidance Document for FDA's current thinking on De Novo submission Other US Medical Device Regulations 1
W How can I test the Patient Leakage Current of an Infusion Pump? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
J Looking for feedback from current/former Intelex users Quality Assurance and Compliance Software Tools and Solutions 4
A Using your current QMS to plan for ISO 9001 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
A Accuracy Calculation of Leakage Current for an Electrical Safety Analyzer General Measurement Device and Calibration Topics 5
E How do I stay current on Regulatory Affairs? ISO 13485:2016 - Medical Device Quality Management Systems 8
J Exceeding the DC Leakage Current Limit IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
Q Seeking Examples of current ISO/TS 16949 Training Procedures IATF 16949 - Automotive Quality Systems Standard 2
L Building your own Leakage Current Tester IEC60990 Other US Medical Device Regulations 4

Similar threads