Customer Complaint - FDA Definition of Complaint

H

hitesh

#1
We are class II and Class III medical device manufacturers (OEM). Few of the medical devices we sell are re-sterilizable and non-sterile. These devices are used multiple number of times by the end-users. We provide 3 month warranty for some devices and 1 year warranty on the others.

Currently, we consider any concerns related to the identity, quality, durability, reliability, safety or performance of a medical device from the customers as a customer complaint eventhough the product is out of warranty.

What is industry standard practice? Is there any Medical Device OEM to help me clarify whether out of warranty products should be treated as complaint as per the FDA defination below or service call?

FDA Defination of Complaint: written, electronic or oral communication that alleges deficiencies related to the identity, quality, durability, reliability, safety or performance of a medical device that has been placed on the market.

Sincerely,
Hitesh
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
Currently, we consider any concerns related to the identity, quality, durability, reliability, safety or performance of a medical device from the customers as a customer complaint even though the product is out of warranty.
For the majority of our Class II disposable/single-use accessory products, the "warranty" period is equal to the stated sterile-barrier-packaging stability-lifetime, i.e. typically five years from date of manufacture.

We don't have a specific statement of policy in regard to the question you ask. If there's something wrong with one of our products defined above, it's likely to be discovered long before five years have elapsed. But, I think we would agree with your statement above, subject to the conditions you applied to it.
 
C

Chris Ford

#3
We are class II and Class III medical device manufacturers (OEM). Few of the medical devices we sell are re-sterilizable and non-sterile. These devices are used multiple number of times by the end-users. We provide 3 month warranty for some devices and 1 year warranty on the others.

Currently, we consider any concerns related to the identity, quality, durability, reliability, safety or performance of a medical device from the customers as a customer complaint eventhough the product is out of warranty.

What is industry standard practice? Is there any Medical Device OEM to help me clarify whether out of warranty products should be treated as complaint as per the FDA defination below or service call?

FDA Defination of Complaint: written, electronic or oral communication that alleges deficiencies related to the identity, quality, durability, reliability, safety or performance of a medical device that has been placed on the market.

Sincerely,
Hitesh
Hi Hitesh,

I worked at a company that was classifying customer service calls the same way. The company manufactured Class II hearing screening analyzers and breath analyzers. When they'd receive a call on a device, they'd record the problem, determine whether it was reportable, then check its warranty status. If out of warranty, it was automatically placed in the "service call" queue and considered "normal wear and tear".

FDA did not like that, and for good reason.

That's not to say that the issues weren't addressed; this company actually had great customer service and responded very quickly to customer calls. However, there is no specified time frame for complaint classification. A complaint is a complaint regardless of whether you receive it during the warranty period or after the defined life of the product.

The first line of the regulation for servicing is very clear, "(a) Where servicing is a specified requirement, each manufacturer shall establish and maintain instructions and procedures for performing and verifying that the servicing meets the specified requirements."

Let's say you make a device that clearly states on its labeling, "results not valid after 10 uses."
Your company defines a service requirement for the device to be returned and "recharged" after 10 uses.
The customer uses the device 11 times and receives a false negative result the last time it is used. The customer calls to file a complaint with the company.

This is NOT a service call, because while it may relate to a specified service requirement, the customer is unaware of the requirement, and feels that the device has malfunctioned. Your complaint investigation should include a review of your labeling, and associated customer training material. The investigation may even continue internally, if your procedures require you to notify the customers when their service is due.

At any rate, this is one of the most common misconceptions I've found regarding the QSR. Both the complaint records and servicing regulations appear lacking in detail, until you start to uncover the nuances. Servicing is a planned activity, not to be confused with "my device is broken and I need it fixed"... that would be considered a complaint.

Remember... the definition for a complaint is: "any written, electronic or oral communication that alleges deficiencies related to the identity, quality, durability, reliability, safety or performance of a medical device that has been placed on the market."

So if it's a call that alleges a deficiency (including reliability, hence your warranty period is of no consequence), it's considered a complaint. If you receive a complaint for a device that's exceeded its intended life, it's still a complaint, but you probably won't be doing much of an investigation on it.
 

howste

Thaumaturge
Super Moderator
#4
Currently, we consider any concerns related to the identity, quality, durability, reliability, safety or performance of a medical device from the customers as a customer complaint even though the product is out of warranty.

What is industry standard practice? Is there any Medical Device OEM to help me clarify whether out of warranty products should be treated as complaint as per the FDA defination below or service call?

FDA Defination of Complaint: written, electronic or oral communication that alleges deficiencies related to the identity, quality, durability, reliability, safety or performance of a medical device that has been placed on the market.

Sincerely,
Hitesh
The definition of complaint is pretty clear. If the customer alleges deficiencies, then it's a complaint. How you handle it will be different if it's out of warranty. Out-of-warranty product may not need to be investigated though. 21 CFR 820 requires us to
FDA said:
review and evaluate all complaints to determine whether an investigation is necessary. When no investigation is made, the manufacturer shall maintain a record that includes the reason no investigation was made and the name of the individual responsible for the decision not to investigate."
So if it's not in warranty that may be justification to not investigate and follow up.
 
R

rlbate

#6
There was some very helpful information in this thread. We have been performing investigations on many of our class II devices long after the warranty period had expired. This has ultimately led to some pretty thin investigations and wasted time as the manufacturing techniques may have been improved many times since the device was manufactured.

Thanks for the insight.
 
M

MIREGMGR

#7
Just my two cents, but I don't think the FDA is going to accept a no-investigation decision based solely on an "out of warranty" status, if the complaint pertained in some manner to safety or effectiveness.

Whether a complaint has any safety or effectiveness aspects is the first question to be decided in the investigation-needed decision process. If it does, our understanding is that an investigation must be done and documented.

I think Chris Ford discussed this effectively above.
 
B

brandomenon

#8
Two more cents to add:2cents:...

Even when a device is out of warranty, it's a good idea to consider risk for each complaint. You can often use a discussion of risk as an explanation for why you did not perform a thorough investigation. It will also help you to prioritize where you spend your time and effort.

If a complaint is more likely to lead to a serious health problem or a recall, then you would invest more time and effort on the investigation. If the complaint, in your expert opinion, has little opportunity to cause a negative health effect (or a limited health effect), then that is reason enough not to go overboard trying to investigate the problem. Of course if you end up with a bunch of complaints related to the same problem, the FDA would expect more follow-up. Frequency also factors into assessing risk!
 
G

gar4guv

#9
Just my two cents, but I don't think the FDA is going to accept a no-investigation decision based solely on an "out of warranty" status, if the complaint pertained in some manner to safety or effectiveness.

Whether a complaint has any safety or effectiveness aspects is the first question to be decided in the investigation-needed decision process. If it does, our understanding is that an investigation must be done and documented.
+1 on this

at the very least, one would have to show that a complaint on an "out of warranty" product will not apply to other batches that are still in-warranty.
 
R

rlbate

#10
I agree, It's the rational for the performance of a minimal investigation that is important. Our device complaints normally revolve around parts that have either broken off (a caster) or worn down (buttons on a control panel) to timers with electronics that expire.

Thanks
 
Thread starter Similar threads Forum Replies Date
J FDA definition of a complaint - Customer orders spare parts 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R Customer Complaint - Samples vs. Removal - FDA Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
K Customer Complaint Trigger Point for FDA Customer Complaints 2
M Reduce occurrence rating based on the PMS data and customer complaint data ISO 14971 - Medical Device Risk Management 2
J Customer Complaint & SCAR, false data Nonconformance and Corrective Action 14
J Customer Complaint Response 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S How to treat a customer complaint ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
PhilM Nonconformance report, Customer complaint investigations and RMAs ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
G Problem Resolution Report Monitoring - Customer complaint or PRR as general motors use Customer Complaints 12
R Does an RMA = a Customer Complaint Customer Complaints 4
G Customer complaint over a Reference Dimension Manufacturing and Related Processes 9
E Is there an Automotive Customer Complaint Index? IATF 16949 - Automotive Quality Systems Standard 3
R In a software development company: Is every bug reported by the customer a complaint? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 27
E Corrective Action or Customer Complaint ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
T Decontamination responsibility on customer complaint investigation Customer Complaints 4
N Possibly worlds first customer service complaint Supply Chain Security Management Systems 0
A What is a Customer Complaint? (ISO 9001:2015) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
M Should Potential Customer Complaint Outcome Define Registrar NC Rating? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
P Customer Complaint Definition - Food Distributor Customer Complaints 5
Moncia Customer Complaint Procedure example needed Document Control Systems, Procedures, Forms and Templates 1
C Medical Device Malfunction - Customer Complaint - Response vs. Correction Other US Medical Device Regulations 3
T PPAP - Complaint from a Tier One customer concerning Product Dimension IATF 16949 - Automotive Quality Systems Standard 26
Q Customer Complaint Vs. Nonconformance - CAPA SYSTEM Customer Complaints 2
L Creating a xlsx Customer Complaint file to track Complaints Excel .xls Spreadsheet Templates and Tools 2
N Customer Complaint - Rapid Response Containment Procedure Customer Complaints 4
J Customer Complaint - PPAP - Issues not called out on the print APQP and PPAP 3
R Customer Complaint/Nonconforming Product Inquiry ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
D Can anyone share a Customer Complaint procedure not requiring logging complaints ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
P Customer Complaint Preventive Action Assistance Preventive Action and Continuous Improvement 3
Q When to issue a Customer Complaint related to efficacy of a Medical Device? ISO 13485:2016 - Medical Device Quality Management Systems 3
S Do I need to initiate a Customer Complaint for Engineering Samples? Customer Complaints 4
A Customer Complaint - Incident Reporting Hardcopy Signoff Customer Complaints 8
B Analysis & Closure of a Customer Complaint Customer Complaints 6
C Customer Feedback, Satisfaction, Complaint Procedure and Measurement Customer Complaints 8
N Closed Customer Complaint Response Form Examples? Customer Complaints 3
J Damaged Parts in Customer Returns - Complaint Investigation? ISO 13485:2016 - Medical Device Quality Management Systems 9
D Is an NCR required in this Customer Complaint scenario? Nonconformance and Corrective Action 6
B Categorization of Customer Complaint: Making Customers feel they are Top Priority Customer Complaints 4
B Understanding the attached sample Pareto Chart for Customer Complaint Quality Tools, Improvement and Analysis 20
J Should I file a Customer Complaint as a Nonconforming Material Report or a CAP? Nonconformance and Corrective Action 5
R Help defining Eyewear Customer Complaint Categories Customer Complaints 5
K Audit NC - No evidence of a full 5 Why Analysis for the Customer Complaint Problem Solving, Root Cause Fault and Failure Analysis 8
somashekar What if a non EU customer makes a complaint to NB ? EU Medical Device Regulations 2
R Everest Customer Complaint System by Lynk Software US Food and Drug Administration (FDA) 1
T Is a Nonconformance Report required for every Customer Complaint? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
J Who should accept the samples of Customer Complaint products ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Data Flow diagram for a Customer Complaint System - Help wanted Customer Complaints 6
B Small Company Customer Complaint System ISO 13485:2016 - Medical Device Quality Management Systems 2
L Internal Audit to Focus on Customer Complaint Re-occurrences Internal Auditing 4
W Re-Evaluating Corrective Actions - Recurrence of a Customer Complaint Nonconformance and Corrective Action 7

Similar threads

Top Bottom