Damaged Load Cell After Calibration - Who Should Pay?
NikkiQSM
Quite Involved in Discussions
#1
Recently we had our mechanical tester calibrated. This mechanical tester has two grips - we place a plastic test part in each of the grips and it tears it apart. We measure the force it takes to break the part and the elongation.
We have used an outside company for the past 5 years to have the unit calibrated.
A few times, the internal measurements were screwed up after calibration and we have had to contact the company to come back and fix it.
This time, the measurement set up for the space between the grips was changed so the tech could conduct the calibration.
When the calibration was complete, we went to conduct a test as we usually do. We clicked on the "Start Position" button and the grips smashed together, possibly damaging and ruining the load cell.
Who should have to pay for this load cell which can be very expensive? Us or the calibration company.
ALSO - When we called to tell the company something was wrong, they basically told us to call the manufacturer of the mechanical company instead. We are not very pleased and definitely will be looking elsewhere for a different calibration company. :mad:
Elsmar Forum Sponsor
Ninja
Looking for Reality
Trusted Information Resource
#2
Two answers to consider:
1. Details aside, the company who performed the action which resulted in the damage was yours. Worded another way, your company broke it, not theirs. 'Nuff said.
2. If you are looking for another calibration company and refuse to use this previous company anymore...see what you can negotiate out of the situation...what have you got to lose?
A wise man once said to me: "The answer is always no if you don't ask."
Of course he was talking about girls at the time...but it applies all over.
Ron Rompen
#3
Unfortunately I have to agree with Ninja on this. The company performing the calibration did not damage the equipment - YOUR company did. Yes, the measurement space was changed, however it would be considered 'best practice' to verify that equipment is working properly after calibration and service, rather than just assuming everything was the way you left it.
You may want to review your service contract with your supplier, in case there is something noted there.
T
thnksnow247
#4
Did the vendor say the equipment was put back to it's original condition?
I think you are on the hook here as it was you that didn't verify the correct setup before operating the equipment.
What machine do you have? Does it have over limit protection enabled? Most load cells can survive well over their rated limit.
Thread starter Similar threads Forum Replies Date
R Receiving Damaged Paint AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
K Packaging and Damaged Parts returned from a Service Provider (Anodize) - Who is Responsible? Manufacturing and Related Processes 20
S Returned but Customer damaged parts. Is it a NC? Misc. Quality Assurance and Business Systems Related Topics 27
M Damaged Customer Property - ISO 13485:2016 Clause 7.5.10 ISO 13485:2016 - Medical Device Quality Management Systems 7
dubrizo How do you handle damaged IM&TE in your organization? General Measurement Device and Calibration Topics 3
K NonConformance for parts damaged during production? Nonconformance and Corrective Action 10
S ISO 9001 Clause 7.5.4 - Damaged or Lost Customer Property Record ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
J Damaged Parts in Customer Returns - Complaint Investigation? ISO 13485:2016 - Medical Device Quality Management Systems 9
Q Recover a Damaged CAPA (Corrective and Preventive Action) Project Preventive Action and Continuous Improvement 6
T J-STD-001E - Nicked or Severed (damaged) Stranding Other ISO and International Standards and European Regulations 4
I Outsourced Process Quality Problem - Reimbursement of damaged parts Supplier Quality Assurance and other Supplier Issues 1
T Notification of lost/damaged Customer Property - Customer Requirement ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Stijloor Jet engines found with damaged, missing parts World News 3
D Damaged Products by Freight Company Nonconformance and Corrective Action 10
A Method of Disposition on Damaged Inspection,Measuring Test Equipment Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
R Load for checking Plating rectifier Manufacturing and Related Processes 1
R Dedicated Chamber / Dedicated load EO sterilization US Food and Drug Administration (FDA) 4
P API Q1 - Monogram Product & API 4F - Proof Load Test Oil and Gas Industry Standards and Regulations 3
K Why does load cell supplier requires force verification General Measurement Device and Calibration Topics 3
H Prodigit DC Electronic Load - Drawing current but showed 0 on the screeen Manufacturing and Related Processes 0
H California Instrument AC Load - 222 Error Message Manufacturing and Related Processes 0
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
F Plastics cans shelf life determination under static load Manufacturing and Related Processes 1
T GDT Callout - How to call out a note that requires some clamp load Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
D Gage R&R Study on Load Cells - Large chemical blending tanks IATF 16949 - Automotive Quality Systems Standard 1
N Heavy tare to simulate high load in a routine sensitivity test of a floor scale? General Measurement Device and Calibration Topics 3
H EO Sterilization Validation - Sterility Testing and Load Configuration Other Medical Device Related Standards 3
C AMS2750E - Noble metal as a Load sensor AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Ethylene Oxide Sterilization Validation-Reference Load Other Medical Device Related Standards 8
H Max and Min Load Determination for Steam Sterilization of Liquds EU Medical Device Regulations 1
E IEC 60601-1 Tensile Load Safety Factor IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Calibration Standard of Load Indicating Device General Measurement Device and Calibration Topics 4
D What is your experience with FDA using "mixed load" i EO sterilization Other Medical Device Related Standards 1
N Load Thermocouple requirement for TUS (Temperature Uniformity Survey) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
H NADCAP Audit Nonconformances - Vacuum Furnace Load Sensors AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
A Work Load Evaluation for Project Management Personnel VDA Standards - Germany's Automotive Standards 4
K Autoclave Log - Do we have to monitor the usage for every load? General Measurement Device and Calibration Topics 3
M MCOSMOS - Monitor your CMM load time Quality Assurance and Compliance Software Tools and Solutions 3
D Gage R&R on a Load Cell fixture that tests the resistance of a Headrest Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
B Fluorescent Lamp Load Test for Switch in IEC 60669-1 Standard General Measurement Device and Calibration Topics 5
C Tolerance Estimation of a Load Cell I want to Calibrate General Measurement Device and Calibration Topics 9
W Force Calibration - In-House Load Cell Calibration General Measurement Device and Calibration Topics 7
R Tip Load Testing Information Reliability Analysis - Predictions, Testing and Standards 2
S Calibration of Load Cell using Compression Machine as Comparator General Measurement Device and Calibration Topics 5
S Load Label Missing from Ladders Occupational Health & Safety Management Standards 7
D Calibration of Load Cells - ISO 17025 & MSA (Measurement Systems Analysis) ISO 17025 related Discussions 5
D Mixed Level Pull Simulation - Stock Evolution and Machine Load Lean in Manufacturing and Service Industries 3
T Weight Load Testing Methods for Surgical Operating Tables ISO 13485:2016 - Medical Device Quality Management Systems 8
somashekar Water and Air Pollution Load Calculations help needed Miscellaneous Environmental Standards and EMS Related Discussions 6
S Load Testing Benchmarking Benchmarking 3
Similar threads
Top Bottom