DC-AC Power inverter with two MOPPs
SebastienBoisvert
I'm working on a medical product that has an applied part type B, and I'm struggling to find a power inverter that has 2 MOPP.
I've been wondering if someone has experienced the same issue? I wondered if this kind of component has the same requirement as an AC-DC, two means of protection. Indeed, since the source is not 'dangerous' (the DC level will be at 24V), an isolation fault would not keep a high voltage on the DC bus.
You can see the diagram below as a reference.
The power inverter is here to provide energy from the battery to a power inlet if the product is not connected to the main supply. Before the inlet, we plan to use a toroidal transformer with 2 MOPP of protection.
Anyone can help?
Benjamin Weber
According to your diagram, the applied part is supplied directly with the 24 VDC? What kind of applied part is this? Is it fully metal? Is it conductive? If yes: Is it necessary to have this conductive connection to the patient? If yes? Maybe it should be BF?
If not, I think it should be possible to apply some kind of solid insulation between the applied part / patient connection(s) and the patient? Unfortunately you cannot apply a PE connection as one MOPP, since PE connection is not always there (when it is not connected to the supply mains)
Kind regards,
Benjamin
SebastienBoisvert
Hello Benjamin,
Thank you for your reply!
The 'applied part' would be a touch panel and probably other control and display interfaces. In my opinion, it would stay a B type.
Indeed, maybe it would be easier to applied two MOPPs at each interface instead of at the inverter.
Thus, the AC-DC and the toroidal would have 2 MOPPs, and each interface would have its own, but the inverter would be a 'generic' part.
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Judy Abbott The impact of laser power on biocompatibility of the products Manufacturing and Related Processes 6
C Power Supply EN60601-1 vs EN62368-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Surge Protectors and other power protection Manufacturing and Related Processes 7
Sidney Vianna Informational APQP4Wind - Advanced Product Quality Planning for the Wind Power Supply Chain APQP and PPAP 3
U Power supply - Change From 160 to 300 Watts - Significant change? EU Medical Device Regulations 2
M MS Power Automate/Sharepoint Files Document Control Systems, Procedures, Forms and Templates 1
O Are power distribution units Multiple Socket Outlets (MSOs)? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L Medical-grade power adapter labeling Other US Medical Device Regulations 2
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Proud Liberal Power Query to combine multiple tables of unequal number of columns Statistical Analysis Tools, Techniques and SPC 1
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
L IEC 62368-1 / Electrical Energy Source vs. Power Source? CE Marking (Conformité Européene) / CB Scheme 2
B Need For BIS Standard Mark? Imported OEM's Power supply,Li-Ion Battery Other Medical Device Regulations World-Wide 0
I Containment procedure for parts after a power outage. IATF 16949 - Automotive Quality Systems Standard 1
P UDI labeling for power module IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
E IEC 61010-1 Dielectric strength test of inner power supply Other ISO and International Standards and European Regulations 0
S Power rating for medical devices in the US - Radiotherapy devices US Food and Drug Administration (FDA) 0
T Is anyone working with N299.1 (Supply/service to nuclear power plants)? Various Other Specifications, Standards, and related Requirements 0
E IEC 60335-1 power input deviation - Table 1 Values Other ISO and International Standards and European Regulations 0
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K Power point training slides on DOE (Design of Experiments) Wanted Training - Internal, External, Online and Distance Learning 2
Marc Renewables, led by wind, provided more power than coal in Germany in 2018 Sustainability, Green Initiatives and Ecology 2
S Responsibilities between government owner and private subcontractor of a power plant seeking ISO 9001 certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
C Electrical Power Plant Supply failure in PFMEAs FMEA and Control Plans 10
H IEC 60601-1 Main Power Cord Requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Infusion Pump Power Failure Alarm IEC 60601-2-24 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Power Cord Length Issue - Medical Device Electrical Safety IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ISO 9001:2015 - Non Applicable Design and Development (Power Plant Operator) Design and Development of Products and Processes 2
H Enclosed vs. Open-Frame Power Supplies - Overdoing it? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Pneumatic & Electrical Torque Tool - Calibration/Verification of 'Power Tools' General Measurement Device and Calibration Topics 2
G Medical Device Power Inlet/Outlet Voltage Labeling IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
x-files [Thermal Power plant KOSTOLAC, SERBIA] Sucessfully certified to ISO 50001 Sustainability, Green Initiatives and Ecology 1
M Is IEC 60601 compliance possible when only generic power-supply is specified? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
U IEC 60601-1, Class 1 or Class 2 on emergency power system IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Power Traceability certificate for values for voltage, current and phase General Measurement Device and Calibration Topics 1
D NQA-1 Standard Applied to Services rather than Items for Nuclear Power Various Other Specifications, Standards, and related Requirements 3
R USB power between Medical Device Data Collection Device Component IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
F Power and sample size for factorial design Using Minitab Software 4
P Medical Device Marking - Power Input Ports IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Reaction plan in case of power outage Misc. Quality Assurance and Business Systems Related Topics 7
B Detachable Power Supply Cable Connection ESD Clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
K How to determine rated input power for the markings? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Selling of Medical Power Supply in Australia/New Zealand and Japan Other Medical Device Regulations World-Wide 5
M Multiple versions of device due to power connectors? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Connection to a generic separate power supply IEC 60601 - Medical Electrical Equipment Safety Standards Series 5