Deciding if a medical device shall be supplied sterile or not
RA Lille
Has anyone got a good rationale, a decision tree or a list of questions that could be used to help DnD engineers to decide if a medical device shall be supplied sterile or not ?
I'm aware of the Spaulding classification, but it's more applied for retreating/reuse of MD.
Here the point is, as a manufacturer how could we decide that a MD shall be sterilized or not ?
lots of factors play an essential role : impact of the method of sterilization (discoloration, temperature,...) impact of the manufacturing flow (environment or not), compatbility of materials…
For example, a gastroscope should be disinfected before being used on a patient. But shall a single used device used in the digestive tract be sterile ? it will be in contact with food and so on… so, physicians say, give it non sterile… the point is, is there any constructive justification of this ? how could I incorporate such rationale in a technical file ? in a sterilization procedure (let's say "choose of sterilization" procedure) that would be defended in a microbiological audit ?
I'm sure I'm clear,
any idea/info is welcome,
thanks !
I wouldn't ask a physician and I wouldn't want a non-sterile gastroscope anywhere near me, would you?
This will be determined in your risk assessment.
You can market it non-sterile, but you'll still have to validate the sterilisation/cleaning method so that you can tell the user how to sterilise/clean it safely before use and to ensure the sterilisation/cleaning method doesn't impact the device's safety and efficacy.
When you say it's single use, is it going to be used once and discarded?
Or will it be reused?
Reuse and resterilisation/cleaning also has to be validated to determine how many times it can be reused safety and effectively.
RA Lille
dear J0anne,
thanks for your answer.
You know, most of gastroscopes used in endoscopy are not sterile, they're disinfected… not sterilized… anyway, it's not the point.
I was initially wondering whether there is any simple decision tree or questions (as part of a risk assessment) that could be used to discuss and debate with RnD engineers. Something to help them to decide if the device should be sterile or not, if it should be sterilized prior use or simply disinfected… .
single use is single use, i.e. used then discarded.
reuse means reprocessing.
Even if the debate about reuse/reprocessing (means cleaning + disinfection or cleaning+disinfection + sterilization) is really interesting, the point of my wondering came earlier : how could an engineer decide or choose if the device he is designing shall be single used, sterile or not, reusable or not. I'm microbiologist and QA/RA, I'm not the engineer… i'm looking for Tools to help risk assessment during RnD process.
somashekar
Just throwing in my hat .... If the device is going to come in contact with blood (Vascular system), then the risk of infection leads you to think sterilization. Otherwise as you mentioned about gastroscopes, they are cleaned and disinfected following a laid out method. (Clinically clean)
You must also look at the pyrogen levels apart from sterilization in your risk assessment.
RA Lille
Dear Somashekar,
i agree with you, when you're QA or microbiologist or both, you could easily think about infection risks.
Indeed, pyrogen levels should also looked at.
I'm looking for something more systematic, as a list of questions or a list of imputs that mechanical engineers could use to check if they'd thought at all risks to define the status of the device they're designing, and also the choice of the method of sterilization (if needed) they could use…
a bit as
which tissues will your device be in contact with ? blood, mucuous tissue, bone, skin….
is your device entering into the body ?
Is your device resistant to high temperature (>120°C) Yes/ No. if Yes, is it resistant to pressure ?
Could your device be discolored ?
is your device containing lots of tubes, inner parts…
does your device contain ambient air ?
I'm looking to draw up a kind of systematic review of risks and sterilization issues to help engineers to manage the risks at the beginning, from the conception.
I understand. My personal preference would be for it to be supplied for use as sterile.
As for tools, the best tool you have is your risk management procedure.
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5
J Deciding between Professional Use or Lay Person IVD Test - Human fecal stool specimens Other Medical Device and Orthopedic Related Topics 1
A Trimming the Inspection Report - Deciding which are the CTF dimensions Manufacturing and Related Processes 3
P Deciding whether or not a new 510k is needed Other US Medical Device Regulations 2
J Deciding what Raw Materials we should require CofCs for ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
shimonv Deciding When to Submit a 510(k) for a Change to an Existing Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
B Deciding when to submit a Special 510(k) US Food and Drug Administration (FDA) 10
A Deciding on when to go for Cp & Cpk or Pp & Ppk Study? Statistical Analysis Tools, Techniques and SPC 3
R Deciding what Processes need Control and Monitoring - 820.70 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
V Is permission from Registrar needed for deciding on in-house calibration? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Deciding if Calibration is needed for Production Instruments General Measurement Device and Calibration Topics 8
A Process FMEA - Help in deciding "detection ranking" of a failure mode FMEA and Control Plans 5
Q Deciding HOw to Measure Alloy Hardness after Casting Various Other Specifications, Standards, and related Requirements 2
A Canadian Safety Labels for medical electrical equipment Canada Medical Device Regulations 0
A Applicability of Photobiological Safety Evaluation for LED used in medical devices Reliability Analysis - Predictions, Testing and Standards 1
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
T Google Fit Health datas and Medical devices Application Medical Information Technology, Medical Software and Health Informatics 0
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Procedure packs with non-medical devices EU Medical Device Regulations 1
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
A Customer Approval (Medical Devices) Document Control Systems, Procedures, Forms and Templates 4
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
D Electrical Medical Devices class I EU Medical Device Regulations 0
J Are complaints applicable to development of medical devices? Customer Complaints 2
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
M Do I need separation in my circuit with a medical charger? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
P ISO 8 classified medical manufacturing room Qualification and Validation (including 21 CFR Part 11) 1
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
Brian Benzarti Medical Devices CLP Regulation Quality Assurance and Compliance Software Tools and Solutions 2
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
J Medical Device Regulations in Lebanon? Other Medical Device Regulations World-Wide 2