Deciding When to Submit a 510(k) for a Change to an Existing Device
shimonv
Hi Everyone,
Some of you may know this draft guidance:
"Guidance for Industry and FDA Staff - 510(k) Device Modifications: Deciding When to Submit a 510(k) for a Change to an Existing Device"
It is interned to replace the current version (from 1997).
Does anyone know when will it become effective? and which parts of it?
If implemented "as is" it will make our lives more difficult and, in may opinion, won't make medical devices more safe & effective to use...
Thanks!
Shimon
Ronen E
Problem Solver
Moderator
Hi Everyone,
Thanks!
Shimon
As stated in: http://www.fda.gov/MedicalDevices/DeviceRegulationandGuidance/GuidanceDocuments/ucm265274.htm :
Although the notice of availability for this Draft Guidance stated that the comment period would close on October 25, 2011, FDA will continue to accept comments until the deadline specified in the forthcoming notice extending the comment period. FDA will post the notice announcing the extension of the comment period when it becomes available.
Obviously the guidance will not take effect while it's open for comment. I understand the above as "this draft is open for comment until further notice", and thus will not take effect until further notice...
Cheers,
Ronen.
MIREGMGR
I've had direct experience with US FDA to the effect that guidances are applied and effectively enforced as soon as they are published, even though they are in draft status. FDA has a number of guidances that have been in draft status for years, and for which there is no active program for further development due to resource prioritization.
Citations of course are always in regard to the US Code, not to a guidance, so FDA's application of a guidance in shaping inspection judgements and filing reviews may only be apparent by inquiry or discussion.
My understanding of CDRH's operational approach is that inspectors and examiners are expected to regard FDA's latest published version of every guidance as definitive unless there has been internal notice that a revision is planned. A majority of guidances that have comment periods upon initial publication aren't revised.
shimonv
MIREGMGR,
Thank you for the valuable input.
My followup question is: what are we to do with regards to this guidance document? Can I simply ignore it and and to internal filing based on the current guidance document. If a field auditor comes we can explain that we are using the current guidance to interpret the meaning of "significant change". I'd be surprised if we get a citation based on a principles from the new draft document which is not effective!
But than again, we are not in a position to argue for federal agent...
Thanks,
Shimon
MIREGMGR
Sometimes when I have a question about what guidance is effective, or about an apparent conflict between an older final guidance and a newer draft guidance, I call FDA DSMICA and ask.
The other option is just to assume that the latest published information is definitive, even if it's stated to be a draft. Usually that's what I do.
Ronen E
Problem Solver
Moderator
OK, perhaps my comment was a bit naive, so I take it back. Thanks for putting things in the right perspective.
A question then: What is the meaning, in this context, of the term "draft"? If they issue guidance and mostly start operating upon it right away, then this term is almost meaningless to me. Every published guidance should be regarded as effective and final until further notice, regardless of its "titled" status, shouldn't it?
shimonv
Hello Ronen,
To answer your question - the best thing would be to call FDA and ask.
If I try to round up all the information available in this thread I'd say this:
A guidance document is a all that it is: a guidance document.
It's not likely to get a citation just because we didn't consider the content of a new draft guidance document; with an effective document we'd get one for sure...
If a field inspector comes, he will challenge your decision based on all the information available to him, including a draft guidance.
If you didn't address the guiding questions laid-out in this draft version as part of your rational for letter-to-file, you could get a citation if he concludes that the change could significantly affect safety and effectiveness. If you don't prepare in advance you'd be more vulnerable. The citation would be for none-compliance with CFR code.
This is taken from the new draft document:
"If a manufacturer determines that its device modification(s) could not
significantly affect safety or effectiveness and therefore decides not to submit a new 510(k), it should document the basis for concluding that it does not require a 510(k). Manufacturers should scientifically justify their conclusions that modifications, individually and collectively, could not affect safety or effectiveness.
A copy of this documentation should be maintained. It is recommended that manufacturers answer each question below to satisfy basic
Quality System requirements for documenting device modifications.
See 21 CFR 820.30 and 820.70(b). "
Therefore I conclude that a letter to file must answer the guiding questions laid out in this document, otherwise he could get in trouble.
Once the document becomes effective he will for sure get in trouble.
Does anyone agree or think otherwise?
Shimon
MIREGMGR
My perhaps-incomplete historical understanding is that US FDA in earlier times intended to have a draft-not-yet-effective, period-for-comments, consensus-final-version guidance development system. In theory that's how it works now.
The US governmental system for bureaucratic review and approval of rules and regulations being such that it is, though, as I understand it the operational people at FDA who are tasked with actually accomplishing a safe and effective medical device environment and who create the first drafts of guidances began to advise inspectors and examiners to regard drafts as representing where the rules were expected to go, during the lengthy period of review.
That then evolved over time, as the review process proved to be practically endless, into an effective abandonment of getting official approval of drafts, and a hardening of the "advice" to inspectors and examiners into "direction".
It's a much more effective system, albeit dictatorial, i.e. it's totally dependent on the FDA's operational people being on exactly the right wavelength to demand just the right amount of rigor.
Wes Bucey
FDA and FAA are similar in that they require notice of changes in products under their purview. The decision point is what is the extent of change which triggers resubmission, perhaps even retesting for safety and efficacy. It is no joke when one of these regulatory bodies comes after a manufacturer becausethey think the manufacturer was trying to "put one over on" the agency, possibly jeopardizing end users, just to avoid the time, trouble, and cost of dealing with a little bureaucratic red tape.
My risk assessment when dealing with either FDA or FAA was always the downside that the agency could arbitrarily shut me down (cutting off income, but not expenses) while I dealt with a glacial bureaucracy.
We have an icon for just this situation: :ca:
Look at it this way - a phone call today is a lot cheaper than most of the alternates. Try to get written verification of the verbal opinion. Don't forget that it isn't only a random FDA employee who may get a bee in his bonnet about your documentation - competitors, disaffected employees, and others have been known to "drop a dime" to make anonymous complaints (sometimes not so anonymous) in hopes of getting a share in any penalties levied against your organization. I'm aware of an organization in Florida that has made a large fortune in such fees in a number of cases with the FDA (not strictly on medical devices, of course.)
B Deciding when to submit a Special 510(k) US Food and Drug Administration (FDA) 10
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5
J Deciding between Professional Use or Lay Person IVD Test - Human fecal stool specimens Other Medical Device and Orthopedic Related Topics 1
R Deciding if a medical device shall be supplied sterile or not Other Medical Device and Orthopedic Related Topics 5
A Trimming the Inspection Report - Deciding which are the CTF dimensions Manufacturing and Related Processes 3
P Deciding whether or not a new 510k is needed Other US Medical Device Regulations 2
J Deciding what Raw Materials we should require CofCs for ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
A Deciding on when to go for Cp & Cpk or Pp & Ppk Study? Statistical Analysis Tools, Techniques and SPC 3
R Deciding what Processes need Control and Monitoring - 820.70 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
V Is permission from Registrar needed for deciding on in-house calibration? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Deciding if Calibration is needed for Production Instruments General Measurement Device and Calibration Topics 8
A Process FMEA - Help in deciding "detection ranking" of a failure mode FMEA and Control Plans 5
Q Deciding HOw to Measure Alloy Hardness after Casting Various Other Specifications, Standards, and related Requirements 2
supadrai Possible to Submit a 510(k) if you have never, do not and cannot produce the device for which you're submitting the application? US Food and Drug Administration (FDA) 4
F How to submit a PPAP when a subcontractor is the sole manufacturer IATF 16949 - Automotive Quality Systems Standard 2
gunnyshore Adding a new facility - do I need to submit an amendment to the MDL or MDEL, or both? Canada Medical Device Regulations 3
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
F How to submit PPAP without IMDS APQP and PPAP 10
M Informational TGA presentation: How to submit an effective good manufacturing practice clearance application Medical Device and FDA Regulations and Standards News 0
Watchcat EU Certification - What do you submit to an NB EU Medical Device Regulations 5
B Submit a 510(k) with a Predicate cleared but not Listed US Food and Drug Administration (FDA) 9
L Reasons to submit a new PPAP or notify the customer APQP and PPAP 3
L 510k exemption process - Do you still need to submit technical dossier? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
N How to deal with catalog parts suppliers who refuse to submit PPAP documents? APQP and PPAP 0
I Where to submit Adverse Events Reports in Europe EU Medical Device Regulations 7
B Should I Request/Submit a PPAP from/to a Supplier/Customer for only Repackaging Parts APQP and PPAP 7
Ronen E FDA releases new draft guidance on when to submit a 510(k) for device modification Other US Medical Device Regulations 10
B Submit a Special 510(k) or "Documentation by our own" 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
L 20 days to submit PPAP APQP and PPAP 9
J Can we submit data with Special 510(k)? US Food and Drug Administration (FDA) 8
A Using Regulations.gov to Submit Comments on FDA Guidance Documents Other US Medical Device Regulations 5
E When to submit a Technical File for a Class IIb Medical Device EU Medical Device Regulations 3
P Do we need to submit Special 510K for a Blood Glucose System? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Customer will not submit Purchase Orders - System for ISO 9001 Compliance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
L Can we submit an AS9102 if we are not AS9100 certified? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
S Guidance to prepare and submit a Level 3 PPAP to a Customer APQP and PPAP 6
I Where do we submit the FAA Preapplication forms? Federal Aviation Administration (FAA) Standards and Requirements 1
O Medical Device Registration in EU - Which Ministry to submit Technical File? EU Medical Device Regulations 9
S How to Submit ISO/TS 16949 & ISO14001 Certificates to HMMA IATF 16949 - Automotive Quality Systems Standard 4
W Submit MDR for a device cleared by FDA but not sold in US? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
P Part Submission Form - Need to submit PSW only once? APQP and PPAP 5
M If I update a DFMEA do I need to submit to my Customer? FMEA and Control Plans 7
C When to submit a 510(k)? After the DMR (Device Master Record) is finished? Other US Medical Device Regulations 4
A Should I submit the entire copy of my software in a 510(k) submission?? US Food and Drug Administration (FDA) 7
Howard Atkins Is there a time frame for CBs to submit the TS 16949 Audit Report? Registrars and Notified Bodies 2
J What is the time limit for an auditor to respond to NCR's and submit audit report? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
sagai New Draft Guidance - When to Submit a 510(k) for a Change to an Existing Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S How to locate a Sponsor in Australia to submit Product Approval Applications by TGA Other Medical Device Regulations World-Wide 4
R In 510k application do we submit documents showing compliance with 21CFRPart820 Other US Medical Device Regulations 12
P Does our company need to be certified in order to submit a 510k Other US Medical Device Regulations 4