I hope this can be useful to others too:

According to 60601-1 applied parts are considered defibrillator protected if, upon testing with a 5KV defib-like pulse on the part, no transient above 1V appears on the enclosure and other accessible parts of the instrument. Obviously the instrument should also withstand this pulse, i.e. should work properly and safely after the pulse, to be considered defib protected.

BUT, 60601-1 also states in 8.9.1.15 that defib protected applied parts require >4mm clearance/creepage.

Here come the 2 questions:

1. >4mm is needed according to 60601-1, but between the applied part and what? which parts/elements of the front-end circuit should be separated by at least 4mm to achieve defib protection? what is the rational for this in the context of defib protection?

2. I am assuming that often a front-end (for example for an ECG) is designed with components such as GDTs, resistors, capacitors, TVS, etc. so that the defib transient is shunted very close to the input port where the electrode is connected. why 4mm then? and again where? only if physical separation is deliverately/by-design used as a MOPP/MOOP in the defib circuit?

Thanks in advance
 
Elsmar Forum Sponsor

MedMartin

Involved In Discussions
#2
There is a rationale for impulse test voltage in one of the appendices of 60601-1. The following is from Appendix A in an older version of UL60601-1: "When a defibrillation voltage is applied to the thorax of a PATIENT, via externally applied paddles (or defibrillation electrodes), the body tissue of the PATIENT in the vicinity of the paddles and between the paddles becomes a voltage dividing system.
The voltage distribution can be gauged roughly using three-dimensional field theory but is modified by local tissue conductivity which is far from uniform.
If the electrode of another item of MEDICAL ELECTRICAL EQUIPMENT is applied to the PATIENT, roughly within the compass of the defibrillator paddles, the voltage to which such an electrode is subjected depends on its position but will generally be less than the on-load defibrillation voltage.
Unfortunately it is not possible to say how much less as the electrode in question may be placed anywhere in this area, including immediately adjacent to one of the defibrillator paddles. In the absence of a relevant
Particular Standard, it must therefore be required that such an electrode and the EQUIPMENT to which it is connected will be able to withstand the full defibrillation voltage, and this must be the no-load voltage as one of the defibrillator paddles may not be making good contact with the PATIENT.
This amendment to the General Standard therefore specifies 5 kV as the appropriate value in the absence of a relevant Particular Standard."

It also clarifies the requirement for the 4 mm:
"From IEC 664, table II, a distance of 4 mm is adequate for pulses of 5 kV having a short duration of less than 10 ms, such voltages arising typically from the use of a defibrillator, with a reasonable safety margin.
The validity of this margin, which has been retained to ensure that the EQUIPMENT pass the defibrillator test, and not only remain safe afterwards but also function normally, comes from three factors:
– The values in IEC 664 already have an inherent safety margin;
– in practice the applied voltage even on the PATIENT’S thorax will be much less than the assumed open-circuit voltage of 5 kV, as the defibrillator will be on load, and it has an appreciable internal impedance and a series inductor which adds to this impedance;
– IEC 664 allows for heavily contaminated surfaces, whereas in MEDICAL ELECTRICAL EQUIPMENT internal surfaces are clean."

And for your question 2 please consider also that there is a common mode test, where the defibrillator protection devices typically do not come into action but the voltage is on the isolation distance to primary.
Does this help?

Best regards,
Martin
 
#3
Thank you Martin. I really appreciate your message and that you went out of your way to find the exact text from the standard.

The rational behind the need for defib protection of up to 5Kv (e.g. gas discharge tubes or semiconductor alternatives like MAX30034) as in 60601-1 was clear to me. No question about that.

The question was more about the why and how of the 4mm clearance. I assume that 60601-1 wants to minimise the risk of those 5Kv coupling to a nearby metal part before a protection device at the input can shunt the high voltage and bring it down to a low risk value. Then common sense seems to dictate that the front-end should be designed with the incoming wires from the electrodes and their PCB traces, up to the first safety energy-rated resistor, remaining at 4 mm from any metal part/or nearby traces.

Is my interpretation consistent with the standard? 4mm between any point that would be at 5Kv during defib test and any other trace or metal part of the enclosure?

As for the second part of the question, you say "consider also that there is a common mode test, where the defibrillator protection devices typically do not come into action".

But wouldn't you always expect to find a protection from each electrode to ground to protect the Instrumentation Amplifier IC? Otherwise the Applied Part would not be defib protected. And then, if each electrode is protected with respect to ground, a common mode test would trigger the protection on all electrodes simultaneously (only the differential protections will not be activated). Is this what you had in mind?

Again thank you for taking the time to answer.
 

MedMartin

Involved In Discussions
#4
Hi HenryC,

I agree with your interpretation and that is also the way we handle it in our designs. A distance of 4 mm (or more) from patient inputs to other parts
before the defibrillation protection resistors.

And the 4 mm is also required to the primary side of the circuit, e.g. the mains power supply or a USB connection. This is for the test with all patient leads connected together during common mode test. In this test the defibrillator protection does not draw current but the voltage is applied across the isolation barrier. The secondary ground from e.g. the ECG amplifier is floating together with the patient leads. Is this concept clear?

Best regards,
Martin
 
#5
Martin,

Yes, it is clear that during common mode ESD or defib test, the electrodes (e.g. ECG) will be connected together and driven by the pulse generator.

In this configuration several Kilovolts will fall on the input protection Resistor, which is in series with a MAXIM300034 device to ground, which is also in series with the stray capacitance between the (isolated) secondary and primary circuits.

Here I think the defib protection (MAX300034 between electrode and floating ground) should draw current, correct? (whereas in your message you were probably thinking of a GDT placed differentially between the input electrodes, in which case the common mode test will not trigger the GDT, and therefore it will not draw current during the common mode test).

Did I understand your comments correctly?

In any case, you note that 4mm clearance is needed also for USB lines and mains wires. I suppose 4mm will be needed for any trace/wire from its entry point into the enclosure to the first element of protection that drops the possible KVolts voltage to low values (the protection resistor or ESD protection semiconductor).

Henry
 
#6
Just a quick additional comment to correct myself:

I said the defib protection between electrode and ground would draw current during a common-mode test. But only transiently I should have added, to charge the stray capacitance between the primary-seconday isolation. After that all the common-mode test (kilovolts) will fall on the primary-secondary isolation. And given that the stray CAP between primary and secondary circuits is small, this transient flow might not be very relevant for safety purposes.
 
#8
Hi HenryC and MedMartin,

I followed your discussion with great interest, and I have an additional question.

Is the 4 mm air clearance also required within one applied part with several patient connections? ECG will have 3-10 patient connections within the same applied part, so do we need 4 mm clearance between all pins in the connector, or just between all pins and any other conductive part.

Thanks!
 

MedMartin

Involved In Discussions
#9
Hi,

as I understand it, the 4 mm is also required within one applied part.
To use your example between ECG leads are also requirements for differential mode testing and energy reduction testing acc. to 60601-1.
Often there are resistors included in the ECG cable with adequate clearance and creepage distances and pulse handling capability. Then there is no requirement on the connector.

Best regards
Martin
 

Peter Selvey

Staff member
Moderator
#10
Would have been good to jump in early here.

The limit of 4mm in 8.9.1.15 is intended to keep the isolation barrier at the normal 4.0mm limit for BF-CF isolation, and ignore the 5kV pulse in cr/cl assessments. Without this clause, manufacturers could be asked for distances up to 126mm (5kV, table 12).

So it is a relaxation for most devices. There are two scenarios where it is not a relaxation: where a BF/CF device intended only for 100-120V regions would normally have a limit of 3.0mm, and for Type B equipment which has no BF/CF isolation, in this case 8.9.1.15 would represent a increased limit.

The distance only applies to the 1.5kV.4mm barrier between the applied part (ECG circuit) and other parts as is normally required for BF/CF classified applied parts. It does not apply between parts of the circuit, for example between RA and LA electrodes. The reason is a little complicated but it comes down to understanding the whole purpose of Clause 8 and means of protection. Any requirement inside of Clause 8 (8.7, 8.8, 8.9 etc) has to be tied back to a identified MOP used to meet 8.1 and 8.4.

That does not mean the distances between RA and LA (for example) before the voltage limiting device are not important. I have seen cases of connectors with 0.3mm clearance being used at this point which makes no sense. Generally, for a rounded pads, 1mm is enough to pass 5kV pulse without breakdown. But anyway, it is an risk/engineering decision, not a limit from Clause 8.
 
Thread starter Similar threads Forum Replies Date
R Defib-proof applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A IEC 60601-2-2 - Defib-proof test amendment to base standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Is Defib Protection mandatory for wearable ECG? IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
A Bookmarking my security protected IEC 60601-1 .pdf file IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
M Touch Protected Plugs used for Peripheral Nerve Stimulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Ajit Basrur How to allow users to enter values in a protected MS Excel worksheet ? Excel .xls Spreadsheet Templates and Tools 4
S Read Only & Password Protected - Sufficient to Control Documents under 4.2.3? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
E Photovoltaic Systems Investments Protected by CB Certification! CE Marking (Conformité Européene) / CB Scheme 1
V Control of Forms - Do the blank forms need to be Password Protected? Document Control Systems, Procedures, Forms and Templates 19
M Identifying Applied Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Q Do you classify applied parts? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B ISO 14971 Applied to Software ISO 14971 - Medical Device Risk Management 2
M 2xMOPP insulation for Applied Part B. IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Technical File - Updating Applied Standards EU Medical Device Regulations 4
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J ISO 14971 applied to ISO 13485? Low risk class 1 devices ISO 13485:2016 - Medical Device Quality Management Systems 3
T Overvoltages consideration in Applied Part for RMS Calculation. IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
B F-type applied part - Separation from ALL(?) other parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
G ISO 9001 8.4 applied to 7.1.3 - Suppliers of infrastructure requirements - IT and Print solutions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
E Functional Gage for TOP applied at RFS Design and Development of Products and Processes 5
K F-TYPE APPLIED PART IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J ISO 9001 8.4.1 - Determining controls applied to externally provided processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
T IEC60601-1: Isolation for an applied part with working voltage of 1500Vp. IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
L PC in patient environment - BF applied part that uses USB to connect to PC IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
N Applied Parts Classification Clarification - Breast Pump IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Are EEG electrodes considered as one applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Number of MOPP for Type BF Applied Part (Medical Device Electronics) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Tagin Can/should SPC be applied to Computer Assembly and Software Imaging? Statistical Analysis Tools, Techniques and SPC 8
A EFQM applied on the ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
L No entanglement hazard of applied parts cables? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D NQA-1 Standard Applied to Services rather than Items for Nuclear Power Various Other Specifications, Standards, and related Requirements 3
S Internally powered ME with non-conductive applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Can a Surface of a profile be applied to a feature of size? GD&T questions Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
S Oxygen mask and nebilzer mask Applied standards EU Medical Device Regulations 2
E Requirements for primary battery powered ME with BF-applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Identifying Medical Device Class and Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T Electrical Surgical Tool with Big Metal Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E What does "other APPLIED PARTS" mean in IEC 60601-1 ed 3.0 8.7.4.7 b) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
kedarg6500 How much test force to be applied for Vickers hardness Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
R Classification of Applied Parts - Blood Measurement Sensors IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
R Is protective earth connection regarded as a part of the applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R Is insulation between F type applied part and other parts a must? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K Definition of ME Equipment/Applied part according to IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
L The definition of B, BF (Body Floated) and CF - BF applied to parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Reason a plane would have a runout applied and not a flatness callout Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
optomist1 Feature Control Frame is applied to the end of a Cylinder - Datum Dilemma Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
W Insulation requirements between F-type applied part and enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Lean Management applied to Rework Zones Lean in Manufacturing and Service Industries 8
P Internet Medical Device Distribution applied to Internet/Catalog Shops Japan Medical Device Regulations 1

Similar threads

Top Bottom