Defibrillator Requirement for ME Equipment connected to HF Electrosurgical Generator

  • Thread starter Seek and Comply
  • Start date
S

Seek and Comply

#1
We manufacture ME Equipment with a handpiece which provides mechanical means for surgical cutting.
We will be testing an HF accessory which attaches to the handpiece, and connects to an HF Electrosurgical Generator (ESU). The HF accessory will provide a coagulation effect using the handpiece.
The ESU has a Defibrillator-Proof Patient Return receptacle.
Is anyone aware if there a requirement that our ME Equipment must be classified as Defibrillator-Proof, due to the HF accessory connection to the ESU?
I have been unable to find anything in IEC 60601-1 or IEC 60601-2-2 which clearly states this.
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Super Moderator
#2
Re: Defibrillator Requirement for ME Equipment connected to HF Electrosurgical Genera

There is no requirement in IEC 60601-2-2 to be defibrillator proof. However, it would be popular and reasonable choice as the neutral electrode would normally remain on the patient during defibrillation.

Technically, this should also apply to accessories. For example if the generator is rated defib proof, then the accessories should also be defib proof.

But in practice it would not be a concern. IEC 60601-2-2 already reduces the voltage to 2kVp defib pulse. As this is just a short pulse (~3ms) it is actually a very weak test. Usually the other tests like HF & mains frequency dielectric strength would be tougher especially for active electrodes. Finally, active electrodes would not normally be used during defibrillation. So, it would not normally be a real consideration for active electrodes.
 
S

Seek and Comply

#3
Re: Defibrillator Requirement for ME Equipment connected to HF Electrosurgical Genera

Thank you for your input, Peter.
The applicable 60601-2-2 standard in this case does not clearly state "APPLIED PARTS shall be classified as DEFIBRILLATION-PROOF APPLIED PARTS", which is certainly the case for patient monitors (60601-2-49), electrocardiographs (60601-2-27), and pacemakers (60601-2-31).

Our compliance lab and certified body agreed with our rationale that the defibrillator-proof classification is not mandatory for our ME equipment.
 
Thread starter Similar threads Forum Replies Date
M Applicability of Means of Protection, working voltage in an Automated External Defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Are Defibrillator Protection and Pacemaker pulse detection features mandatory as per IEC 60601-2-25 or 2-27...? IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
S Defibrillator protection test IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Does anyone have example test report of defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Does OHSAS 18001 require organization to have AED (Automatic External Defibrillator) Occupational Health & Safety Management Standards 4
E Applicability of Defibrillator Energy Test to Implantable Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Marc FDA begins inquiry on Guidant defibrillator flaw World News 1
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
L Water requirement for Non-sterile topical OTCs Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
W Pfmea function requirement and failure mode FMEA and Control Plans 6
Y 510k requirement for bringing in gloves Medical Device and FDA Regulations and Standards News 0
S PSUR - MDR First Submission Requirement? EU Medical Device Regulations 2
J When PPAPs are not a contractual requirement APQP and PPAP 9
I Overwhelmed with attribute MSA requirement for visual inspection IATF 16949 - Automotive Quality Systems Standard 8
Brizilla Employee Data Privacy Policy - ISO 9001:2015 requirement(s)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P Clinical Trial Requirement - No equivalent devices in the market CE Marking (Conformité Européene) / CB Scheme 3
J Requirement for Retention of Records of Withdrawn Documents of External Origin Document Control Systems, Procedures, Forms and Templates 3
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
A The requirement of the PMS plan--suitable indicators and threshold values EU Medical Device Regulations 0
S Battery powered device - electrical protection requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 18
K Regulatory requirement of SaMD with machine learning component IEC 62304 - Medical Device Software Life Cycle Processes 3
A ASL requirement when the supplier is certified for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
P Regulatory or registration requirement for disinfectant in ASEAN Registrars and Notified Bodies 1
P IATF 16949 requirement - error-proofing in control plan IATF 16949 - Automotive Quality Systems Standard 3
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
B IATF16949 audit requirement - Auditor request UCL and LCL must be show Xbar-R, IATF 16949 - Automotive Quality Systems Standard 7
J 21 CFR 821 Medical Device Tracking Requirement 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
W Misinterpretation of requirement acceptable as root cause? Problem Solving, Root Cause Fault and Failure Analysis 19
S ECG Cable Banana to Snap or Tab electrode Adapters -- ISO 10993 Requirement Other Medical Device Related Standards 0
D IATF 16949 Requirement for CMMI in a Global Company Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
L Stage 2 audit - Requirement for 3 months of records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M Requirement to manufacture under a quality management system EU Medical Device Regulations 4
A Risk Number for each software requirement IEC 62304 - Medical Device Software Life Cycle Processes 7
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
I Laboratory Fridge / Freezer Inventory - Requirement(s) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
A Several sections of traditional 510(K) have the same or similar requirement of content. US Food and Drug Administration (FDA) 4
S Requirement to Conduct New Shelf-life Testing? (re-do testing for design change) EU Medical Device Regulations 3
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
Q User Requirement Specification for HR (Human Resource Management System) Manufacturing and Related Processes 1
P Obstruction alarm requirement ISO 80601-2-12 Other Medical Device Related Standards 1
P Electrosurgical Device User Need: Cord Flexibility -> Requirement Other Medical Device and Orthopedic Related Topics 4
C Industrial scales and MSA (IATF 16949 requirement 7.1.5.1.1) IATF 16949 - Automotive Quality Systems Standard 30
C ISO 13485 Requirement in Australia/NZ for class 1? ISO 13485:2016 - Medical Device Quality Management Systems 1
N Label Placement Requirement in MDD/MDR EU Medical Device Regulations 1
adir88 MDR requirement: Risk Management Plan for "each device" ISO 14971 - Medical Device Risk Management 5

Similar threads

Top Bottom