Defibrillator Requirement for ME Equipment connected to HF Electrosurgical Generator

  • Thread starter Seek and Comply
  • Start date
S

Seek and Comply

We manufacture ME Equipment with a handpiece which provides mechanical means for surgical cutting.
We will be testing an HF accessory which attaches to the handpiece, and connects to an HF Electrosurgical Generator (ESU). The HF accessory will provide a coagulation effect using the handpiece.
The ESU has a Defibrillator-Proof Patient Return receptacle.
Is anyone aware if there a requirement that our ME Equipment must be classified as Defibrillator-Proof, due to the HF accessory connection to the ESU?
I have been unable to find anything in IEC 60601-1 or IEC 60601-2-2 which clearly states this.
 

Peter Selvey

Leader
Super Moderator
Re: Defibrillator Requirement for ME Equipment connected to HF Electrosurgical Genera

There is no requirement in IEC 60601-2-2 to be defibrillator proof. However, it would be popular and reasonable choice as the neutral electrode would normally remain on the patient during defibrillation.

Technically, this should also apply to accessories. For example if the generator is rated defib proof, then the accessories should also be defib proof.

But in practice it would not be a concern. IEC 60601-2-2 already reduces the voltage to 2kVp defib pulse. As this is just a short pulse (~3ms) it is actually a very weak test. Usually the other tests like HF & mains frequency dielectric strength would be tougher especially for active electrodes. Finally, active electrodes would not normally be used during defibrillation. So, it would not normally be a real consideration for active electrodes.
 
S

Seek and Comply

Re: Defibrillator Requirement for ME Equipment connected to HF Electrosurgical Genera

Thank you for your input, Peter.
The applicable 60601-2-2 standard in this case does not clearly state "APPLIED PARTS shall be classified as DEFIBRILLATION-PROOF APPLIED PARTS", which is certainly the case for patient monitors (60601-2-49), electrocardiographs (60601-2-27), and pacemakers (60601-2-31).

Our compliance lab and certified body agreed with our rationale that the defibrillator-proof classification is not mandatory for our ME equipment.
 
Thread starter Similar threads Forum Replies Date
S Defibrillator Protection applicability When using Multiple applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Applicability of Means of Protection, working voltage in an Automated External Defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S Are Defibrillator Protection and Pacemaker pulse detection features mandatory as per IEC 60601-2-25 or 2-27...? IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
S Defibrillator protection test IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Does anyone have example test report of defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Does OHSAS 18001 require organization to have AED (Automatic External Defibrillator) Occupational Health & Safety Management Standards 4
E Applicability of Defibrillator Energy Test to Implantable Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Marc FDA begins inquiry on Guidant defibrillator flaw World News 1
P On-going stability study requirement for medical device US Food and Drug Administration (FDA) 2
F CQI-22 - CQI-12 Requirement for Mills IATF 16949 - Automotive Quality Systems Standard 5
R Regarding UL Certification requirement CE Marking (Conformité Européene) / CB Scheme 6
U Does Medical Device training video falls under labeling requirement ? Other US Medical Device Regulations 4
M Citing Against Requirement 7.1.4 Environment for the operation of processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 59
M Is there a mandatory requirement to notify customers of the complaint resolution in ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
D New Device Post Market Requirement for MDR EU Medical Device Regulations 3
S AS9120 - New Internal Auditor Training requirement? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
H Is it a requirement for run charts to have inspection data or can it have just a pass/fail check mark Records and Data - Quality, Legal and Other Evidence 4
I Brazil clinical data/trial requirement Other Medical Device Regulations World-Wide 1
K ISO 13485 Design Requirement with respect to "component" manufacturers ISO 13485:2016 - Medical Device Quality Management Systems 11
D Packaging requirement ? EU Medical Device Regulations 0
C Isolation requirement for USB connection IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C Creepage requirement Wearable ECG IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
V IATF 16949 GM CSR 10.2.4 Error-proofing requirement IATF 16949 - Automotive Quality Systems Standard 2
D IS0 13485 - Design perspective to regulatory requirement. ISO 13485:2016 - Medical Device Quality Management Systems 3
CaliperJim Uncertainty and Accuracy Requirement Deviation Provision General Measurement Device and Calibration Topics 2
Y Software to Manage IEC 62304 Traceability Requirement IEC 62304 - Medical Device Software Life Cycle Processes 3
Ajit Basrur Where to find a pagination requirement as "Page x of y"? ISO 13485:2016 - Medical Device Quality Management Systems 7
T Logistics and 3PL regulatory requirement in India for Pharmaceuticals and Med devices Other Medical Device Regulations World-Wide 0
G ISO 9001 - when implementing customer requirement ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
H Regarding external surface temperature limit requirement for lab incubators. CE Marking (Conformité Européene) / CB Scheme 2
H OTC DRUG Private Labelling Distributor: GMP REQUIREMENT ISO 13485:2016 - Medical Device Quality Management Systems 0
R UDI Requirement at home patient use EU Medical Device Regulations 2
L Definitions Requirement ISO 14971 - Medical Device Risk Management 10
D What distributor requirement should we (manufacturer) require from the Distributor? EU Medical Device Regulations 4
Y Brazilian lab requirement for medical device manufacture Other Medical Device Regulations World-Wide 1
J QTC Requirement as per API 6A-21st Edition Oil and Gas Industry Standards and Regulations 1
J NADCAP Root Cause - Misinterpretation of Checklist Requirement Various Other Specifications, Standards, and related Requirements 5
eldercare Multi-Site ISO/AS Certification Requirement for some sites ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 22
H FAI Requirement After Moving CNC Machine? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
S Biocompatibility requirement for electrode adapters Other Medical Device Related Standards 1
briteme4 AS9102 FAIR Requirement AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
B Put on escalation by customer? is there a requirement to notify registrar? IATF 16949 - Automotive Quality Systems Standard 6
E Translation requirement labels EU Medical Device Regulations 4
Ed Panek Does this FDA Requirement Apply to international (not USA) distributors for USA based manufacturing companies? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
C Requirement for Revision of Standard in all Documentation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K ROHs compliance requirement REACH and RoHS Conversations 10
E Language requirement DoC EU Medical Device Regulations 6
I QMS monitoring, measurement, analysis and evaluation requirement - Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
J Requirement for Signature on Document Document Control Systems, Procedures, Forms and Templates 14

Similar threads

Top Bottom