Define "Rework" Under Medical Device Standards
Beth Workman
We extrude component tubing for medical devices. With extrusion a true re-work of our standard process is never done because we cannot reprocess the raw material for medical devices. However, in one occasion this past year we received a returned product that was cut to the incorrect length and had to trim the product to its required specification.
I would, and did, consider this a re-work, as we were altering the parts from their originally produced state, however, there are some that disagree with my assessment on this.
Looking for some input on this one.
Do you agree or disagree with my assignment of this term? Why or why not?
Thanks!
Rework: A process that adds, removes, or replaces parts to create an item with a changed bill of materials. The configuration of the inbound device or sub-assembly changes to a different configuration when it leaves the process. Configuration in, does not equal configuration out at some level.
Repair: A process that replaces one or more parts with conforming parts of the same part number(s) (including dash number) to bring the product into compliance.
This is how we operate. Successful Manufacturer, Re-worker, repairer of Class III medical devices and being audited by FDA.
Hope it helps.
QE.
P
Pudge 72
Your assessment is spot on.
However, also being in the Medical Device / Component manufacturing industry there are more than just a few who disagree with me and refuse to even acknowledge the term re-work and its associated meaning.
Rework is not an approved process, period end if story. If you didn't have a non-conformance you wouldn't have rework.:notme: But, you cannot have a non-conformance because there is all kinds of bad things associated with that. :lmao:
Al Rosen
Hope it helps.
QE.
You have it backwards. Rework brings it into conformance to drawing & spec requirements while a repair does not meet all drawing requirements but makes it functional. Look at From MIL-STD-1520C, Corrective Action and Disposition System for Nonconforming Material
Marcelo
On medical device standards (Basically ISo 13485), rework is applied only to the finished non-conforming product and means bringing the prodict back into full conformity so it can be sent to the customer as new.
MIREGMGR
We recently had a periodic re-assessment during which rework became a discussion topic and eventually an OFI.
Among our many other processes, we apply flexomeric coatings to fabricated-complex-shaped flexible polymeric foam materials so as to create multiple-use, practically cleanable patient positioners for medical imaging applications. The general first step in application of the coating is a spray process. This step often, but not always, results in pinholes, which are identified by visual inspection of each part after the first drying step. A subsequent step involves various microtechniques for pinhole patching using small quantities of the same coating. Visual inspection and patching continues until each part is determined to meet specs, i.e. no detectable pinholes.
We don't have the option of engineering out the creation of pinholes. That's been studied in the past. They're inherent to the materials and micromorphologies in question. We don't know of any other materials that would result in end products that would be as safe and effective.
Our work process is formally procedurized, but in a soft way. There isn't a quantification of how many pinholes is too many, or what percentage of total area can be patched, or any other characteristic other than that a part with unpatched pinholes does not meet specs but is eligible for further work, and a part with no remaining pinholes (and other quality characteristics, able to be consistently discriminated) does meet specs.
We've been doing this for approximately 32 years with hundreds of thousands of multiple-use devices sold, and essentially complete user satisfaction. The products are simple, relatively inexpensive and clinically safe and effective.
The involved auditor has asked us to define when the existing production process ends, and rework begins. His argument is that at some point, this patching or touchup process must be rework. He wants the existing procedure to be quantified to the extent possible.
Because of the subjective nature of the process and the vague meaning of rework in the context of such subjective processes, we haven't decided how to proceed. We take OFIs seriously, but we are leaning toward this one being too difficult to do in the way he would like.
Roland Cooke
Off the cuff, I would make the case that MIREGMGR's example isn't rework...it's a (vaguely) circular manufacturing process that doesn't finish until the part meets spec.
To my mind rework would apply if a part was deemed to be free of pinholes...and only subsequently found to not be, i.e. it's non-conforming product.
L
Laura Halper
I agree with Roland. I do not think there is anything wrong about a process in which some of the steps may be repeated as necessary.
I'm assuming that MIREGMGR's plant does not write up a non-conformance each time they have to patch some pinholes. This is because they have determined (with 32 years of experience!) that some frequency of pinholes is inherent in their process. So I would consider that the entire process includes initial spray, inspection, and subsequent patching as needed.
Since the auditor wants a definition of when the process ends and rework begins, I would suggest defining the process as above (spray, inspect, patch as needed). Rework could be defined as re-processing that is done after the part has been released for the next operation, or re-processing done on product returned from the customer.
I think you could draw an analogy with a craftsman's work. For example, a carpenter may sand a piece, inspect it for smoothness and shape, and then resand and reinspect several times. The inspection and resanding is part of the process. But if a customer returned a piece because they got splinters, then the carpenter would be doing re-work when he resanded the returned part.
A Define timeline for Major and Miner Audit finding General Auditing Discussions 4
T How do you define your Hazards? <a Risk Management discussion> ISO 14971 - Medical Device Risk Management 16
M Define voltage and frequency to perform tests 61010-1 and 61326-1 for CE certification CE Marking (Conformité Européene) / CB Scheme 4
I Sampling processes - Who must define the AQL level? AQL - Acceptable Quality Level 9
V Who should define and own the Design and Development Plan and how to maintain the updates and revisions. ISO 13485:2016 - Medical Device Quality Management Systems 2
S API Spec Q1 - How to define Management Representative competency for QMS Oil and Gas Industry Standards and Regulations 12
M How To Define ISMS (information Security Management System) Scope IEC 27001 - Information Security Management Systems (ISMS) 18
K How to define Expected life service life of medical device Other Medical Device Related Standards 4
S How to Define Importers under EU MDR / Brexit EU Medical Device Regulations 3
M Should Potential Customer Complaint Outcome Define Registrar NC Rating? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
K ISO 9001:2015 clause 9.2.2 a. - Define the audit criteria and scope Internal Auditing 2
Q QI Macro Histogram - Can someone define *sorted data*? Capability, Accuracy and Stability - Processes, Machines, etc. 7
H How to define Root Cause when some points are out of control chart Statistical Analysis Tools, Techniques and SPC 6
I How do you define Risk (Medical Device)? ISO 14971 - Medical Device Risk Management 30
M SOP or template for a study to Define Storage Conditions of Orthopaedic Implants EU Medical Device Regulations 3
D Definition Client - How does the government define their clients? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 1
alonFAI How to define a Risk Based Approach for Supplier Management per ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
G How to define the scope of QMS as per ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 21
R How to define QMS certification scope statement? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
R Review of "Key Data" for contract labs, but SOP doesn't define "key data". Problem? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
G Procedure to define Signing Authority for Procurement Limits ISO 13485:2016 - Medical Device Quality Management Systems 2
P Can a company define new quality standards for special industry ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
L Are there any requirements to define barcode requirements ? Misc. Quality Assurance and Business Systems Related Topics 2
X How to define Calibration Acceptance Criteria General Measurement Device and Calibration Topics 3
H ISO 17025 - How to define a "Test Equipment" ? ISO 17025 related Discussions 2
J Where do you define Internal Auditor qualifications? Internal Auditing 9
V Is there an approach to define the "must 'or' should" in supplier audits? US Food and Drug Administration (FDA) 2
T Internal Audit - How to define the Importance of Departments and Processes Internal Auditing 8
T Help me understand how to define Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 40
L How to define R & D Receiving (Incoming) Inspection Plan Design and Development of Products and Processes 18
B How to define and implement Configuration Management Document Control Systems, Procedures, Forms and Templates 5
C How to Define and Document Controls of Outsourced Processes Food Safety - ISO 22000, HACCP (21 CFR 120) 5
S Please help me define training requirements for a Career in Regulatory Affairs 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
K How to define the Acceptances Criteria for all equipment? Manufacturing and Related Processes 7
L Definition Program - How do you define Program with regard to ISO 9001? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
A Supplier Evaluation SOP - How do I Define Major and Minor Suppliers? Supplier Quality Assurance and other Supplier Issues 14
Q Where to define Authorities and Responsibilities in Documentation? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
J Define Energy Used/Delivered - Applicable to Electrical or Mechanical Power or both? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R 820.50 (A) (2)Define the Type and Extent of Control to be exercised over Vendor Misc. Quality Assurance and Business Systems Related Topics 5
A Where to define Process Tailoring Form used in CMMI in the ISO 9001 Quality Manual? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Q Criteria to define QMS processes in ISO/TS 16949:2009 IATF 16949 - Automotive Quality Systems Standard 23
S How to define New Equipment? Device is Returned, Refurbished or Repaired Misc. Quality Assurance and Business Systems Related Topics 3
C Controlling Documents: Beyond the standard, how do we truly define what to control ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
C How to define Process Special Characteristics (SC) FMEA and Control Plans 4
kedarg6500 What is the meaning of "define/defined" in ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
Crusader Local Control Document procedure....define it or not? Document Control Systems, Procedures, Forms and Templates 24
M Excel Templates for Plan & Define Phase in NPI Process for Tire Manufacturer Excel .xls Spreadsheet Templates and Tools 1
R Define Data from Taguchi to Response Surface Methodology in Minitab Using Minitab Software 2
K How to define PVC Pellet Quality Manufacturing and Related Processes 8
J How to define a Product's Realization Process and Scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8