Defining a lower ESD test level in IEC 60601 safety test
We are facing a major ESD test challenge with integrating a high-quality image sensor into a class-2 product. We spent 5 months changing our optical and mechanical designs without getting to the required test levels thus far. We can not mechanically block the optical path due to the loss of video quality, and we can only get about +/-6KV with thin transparent windows. If the product remains safe, and it does not cause any harm to the patient or operators, can the manufacture of a class-2 medical device define a lower ESD test level for a sensitive imaging sensor that can not pass +/-8KV or +/-15KV static discharge? Is this managed in the risk management document? Should there be a specific instruction for the certified test labs to exclude a specific area during the 60601 ESD test? If this is permitted, does this need special labeling?
Peter Selvey
It's not unusual to need reduced levels for immunity due to the medical device technology. It's well covered in the standard. Do you have a copy?
Thank you Peter! I have an old copy of 60601-1:2005. Is there a particular chapter and section that I should be looking into? Thanks, Alex
Jaydub
There is a clause in the ESD standard (IEC 61000-4-2:2008, 8.3.2e) which says that connector contacts or other accessible parts that are ESD sensitive because of functional reasons and are provided with an ESD warning label, may be excluded. You would also need something in your documentation about any special ESD mitigation procedures.
Peter Selvey
FYI The detail is IEC 60601-1-2 (just for EMI/EMC), not IEC 60601-1.
It's a bit annoying as the EMC standards are expensive, and most of the detail is there for the test labs. But medical device manufacturer still needs to have the standard as there are requirements for the instructions for use, test planning, risk management and so on that feed into/from the tests. And of course in this case, the structure on how to justify, arrange and disclose any modified test levels.
Thank you for your inputs! This is very useful. I will look into the option of adding a label to the product and writing a warning in the instructions.
J Defining staff competence - Small mechanical workshop Occupational Health & Safety Management Standards 20
D Question regarding ECO process, specifically for Life Science products and defining form fit and function ISO 13485:2016 - Medical Device Quality Management Systems 1
M Defining frequency of measurement tools callibration Calibration and Metrology Software and Hardware 3
M Defining and Documenting Record Retention CE Marking (Conformité Européene) / CB Scheme 5
G Defining performance metrics for DFMA implementation Design and Development of Products and Processes 2
S Defining a Quality System from scratch - Preferred system and documentation names Document Control Systems, Procedures, Forms and Templates 4
A Defining Expected Service Life in Risk Management File Reliability Analysis - Predictions, Testing and Standards 5
C Defining Approvals Required for Design Control Documents ISO 13485:2016 - Medical Device Quality Management Systems 6
K Defining Acceptance Quality Level, I need clarity on AQL 1.5, 2.5, 4.0 AQL - Acceptable Quality Level 5
M Defining the lifetime of orthopedic implants joints Other Medical Device and Orthopedic Related Topics 2
C AS9100 rev D 8.5.1 c 2 - Defining the Machine in-process frequency per ANSI/ASQ Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
V Defining Safety Precautions for Category 4,5 Molecules Occupational Health & Safety Management Standards 2
E European Regulations defining the terms Repair and Refurbish EU Medical Device Regulations 5
T Defining Major vs. Minor Changes to Procedures ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
GStough Adequately Defining Which Suppliers to Audit and Frequency Supplier Quality Assurance and other Supplier Issues 8
E Quality Techucuan (Technician) in Electronics - Defining Postion Requirements Career and Occupation Discussions 4
moritz Defining a good Scope for Critical SOPs ISO 13485:2016 - Medical Device Quality Management Systems 7
T Standards for defining audible alarms/warnings for OR instruments IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Defining Critical Vs. Non-Critical Suppliers/Service Providers (API Q1, 9th. Ed.) Oil and Gas Industry Standards and Regulations 2
B IEC 60601-2-24 - Defining Storage Volume IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Defining Sub-Disciplines for Chemical Testing Laboratory Employee Proficiency Testing General Measurement Device and Calibration Topics 1
T Defining Nonconformances in a Service Organization ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
M Defining Reliability and Confidence Levels Reliability Analysis - Predictions, Testing and Standards 3
J Defining Martial Arts and Gymnastics Statistical Techniques Statistical Analysis Tools, Techniques and SPC 4
V Defining the criteria for equipment to be qualified or requalified Qualification and Validation (including 21 CFR Part 11) 2
R Need help on defining scope for Design Verification File for Class III IVD 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J Defining CCP (Critical Control Points) in a Rice Mill Plant Food Safety - ISO 22000, HACCP (21 CFR 120) 9
S Process Map and defining KPIs Misc. Quality Assurance and Business Systems Related Topics 5
5 Major Nonconformance for not "clearly" defining the "device lifetime" ISO 13485:2016 - Medical Device Quality Management Systems 2
E Defining the lifetime of an Implantable Medical Device Other Medical Device and Orthopedic Related Topics 5
B Defining Expected Oxygen Leakage for Safety Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G Defining Post Mold Cure Ramp-Down Temperature Manufacturing and Related Processes 2
K Audit Nonconformity on Defining 'Outsourced' Infrastructure Maintenance Quality Manager and Management Related Issues 21
G Points to consider while defining the Quality Policy AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
G Defining Quality Objectives for Product Realization and Design and Development AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
S Developing Documentation and Defining Processes as Subcontractor IATF 16949 - Automotive Quality Systems Standard 6
C Defining ISO 9001:2008 Scope for a Sterilization Company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Defining Skilled vs. Semi-Skilled vs. Unskilled Labor Manufacturing and Related Processes 1
I Defining the scope for ISO 9001 Registration - Software, Hardware and Customer Care ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Product Specification vs. Information Defining Product - The differences? 7.3.3.1 IATF 16949 - Automotive Quality Systems Standard 6
R Defining Interaction of Processes in a Software Company Software Quality Assurance 3
N Where to begin defining and monitoring Quality Metrics in a Machine Shop Manufacturing and Related Processes 9
A Testing Process Audit - Defining a Process Compliance Mechanism Software Quality Assurance 2
R Defining the type of Applied Part - Metal Probe (Applied Part) employs Water Cooling IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Defining Security Interfaces for Scope for ISMS - Need help IEC 27001 - Information Security Management Systems (ISMS) 10
A Defining the differences between Prototype vs. Production Guidelines? Contract Review Process 5
R Help defining Eyewear Customer Complaint Categories Customer Complaints 5
G Quality Objectives - Where to start defining Quality Objectives? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
R Criteria for defining FMEA & Control Plan Critical and Significant Characteristics FMEA and Control Plans 10